Hi, has anyone ever used a waeco portable fridge battery 36amp hours to power their cpap whilst campinf
If so could you use solar panels to recharge ?
my cpap uses 3.75 amps per hour so does this mean the battery woul d last for 48 hours ie 36 divided by 3.75
now i know i am a dummy, but i do need some help
jollyjoe1234
cpap and batteries
Re: cpap and batteries
That fridge battery should work if you already have but you might want to just get a marine battery for $50. These are made for long drain and have a 72 month (pro-rated) warranty. Ive had good luck running a CPAP (he Devilbiss Intellipap switches over automatically) in a power outage. If the solar panels dont work you can get one of these portable 2-cycle generators for around $75. So for $125 you can get a heavy duty rechargeable power source and start each knight with a full charge. Even if the sun doesnt come out and you cant get a solar charge you are covered.
- JohnBFisher
- Posts: 3821
- Joined: Wed Oct 14, 2009 6:33 am
Re: cpap and batteries
jollyjoe1234, you wonder if the 36amp hour batteries would work for 48 hours. It depends on the batteries. If they are deep cycle batteries, they should last for 24 hours before they need to be recharged. You should not drain a battery more than 1/2 way before you recharge it. So, the equation would be.
You can find information about Resmed and batteries at:
http://www.resmed.com/us/assets/documen ... lo_eng.pdf
You can find the Respironics information off their sleep therapy page:
http://www.healthcare.philips.com/main/ ... /index.wpd
Hope that helps!
By the way, if you do that math, it comes out as:hours = (amp-hours /amps per hour) / 2
Does your unit really draw 3.75 amps per hour. Seems high.hours = (amp-hours /amps per hour) / 2
hours = (36 / 3.75) / 2
hours = 9.6 / 2
You can find information about Resmed and batteries at:
http://www.resmed.com/us/assets/documen ... lo_eng.pdf
You can find the Respironics information off their sleep therapy page:
http://www.healthcare.philips.com/main/ ... /index.wpd
Hope that helps!
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O |
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński