Intellipap Menu - HELLP!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Intellipap Menu - HELLP!

Post by Julie » Sun Oct 03, 2010 12:11 pm

Hi - didn't know what to title this, but... I went to go into the clinical menu mode and accidentally went into Swedish language from English (the default)! I have pushed and pulled every button now to try to go back, but cannot. Does anyone else out here have the brain I am obviously missing? I do have the instructions, but they don't address getting back from Swedish to the "default" English !

User avatar
Emilia
Posts: 1873
Joined: Mon Jul 12, 2010 8:56 am
Location: Florida, USA

Re: Intellipap Menu - HELLP!

Post by Emilia » Sun Oct 03, 2010 12:33 pm

Go into Clinician Mode..... hold down the Down and Delay buttons at the same time and then hit the On button. Now.... using either left or right arrow, the FIRST item in the list is Language. You should be able to go back to English there. To scroll through language selections use the up/down arrows.
Yes, that blue eyed beauty is my cat! He is a seal point, bi-color Ragdoll. I adopted him in '08 from folks who could no longer care for him. He is a joy and makes me smile each and every day.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Intellipap Menu - HELLP!

Post by Julie » Sun Oct 03, 2010 1:03 pm

Oh thank you! I HAD done that, but obviously freaking out does not help, and when I went back just now to try again, everything was fine. I really should remember to take my pill every day !

User avatar
Junebug999
Posts: 231
Joined: Tue Aug 24, 2010 11:02 pm

Re: Intellipap Menu - HELLP!

Post by Junebug999 » Sun Oct 03, 2010 1:18 pm

Swedish language from English (the default)! HAHAHAHAHA....well you're lucky we have Emilia. If she ever installs an 800 number I wanna put her first on that list AND I'll put her on SPEED DIAL!!!