AHI all over the place lately

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Jason1975
Posts: 51
Joined: Sat Apr 24, 2010 7:37 pm
Location: West Central Wisconsin

AHI all over the place lately

Post by Jason1975 » Sat Aug 21, 2010 11:07 am

I've been struggling with trying to keep my AHI under 10 ever since I was put on cpap almost 2 years ago. I have an AutoSet II on auto mode. My pressure settings I had set from 10-20. It seems like I go in spurts. I can have ahi's of 8-10 for a few months straight and then all of a sudden they jump from 14-20. Could allergies cause this or what? I've been trying different ranges and it seems like my machine works better if I have my min. pressure close to my original prescribed pressure. But then again, other nights I think the opposite. When I have my minumum pressure set at 8, my 95th % is like 10 something with high ahi's. When I have my minumum pressure set at 10, my 95th % usually is in the 11 range. When I have my minumum set at 12, sometimes my 95th % stays at 12 and sometimes it goes up to like 13 or 14. Last night, my ahi was 21 with an ai of 4.1. My minumum pressure range was set to 11 and my 95% said 11.6. Any thoughts?? I just can't get any consistancy with anything. I just so discouraged.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan Version 3.11, ResScan Data Card Reader, Aussie Heated Hose, backup Resmed S8 Autoset II for travel.

Mary Z
Posts: 1493
Joined: Wed Oct 28, 2009 9:55 am

Re: AHI all over the place lately

Post by Mary Z » Sat Aug 21, 2010 12:25 pm

Jason1975, I have the same problem and I know how frustrating it can be when you're doing everthing you can. I have been told with the ResMed machine to pay more attention to the AI than the AHI because of the way the ResMed scores hypopneas. Still I know your AI is not in the range you want it to be. I am on Spontaneous Mode with pressures of 24/20 and I have days when I meet the target, but not consistently.
Sorry, no answers, just wanted you to know you're not the only one out there.

_________________
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Additional Comments: PR System One Remstar BiPap Auto AS Advanced.
Dog is my copilot

ozij
Posts: 10473
Joined: Fri Mar 18, 2005 11:52 pm

Re: AHI all over the place lately

Post by ozij » Sat Aug 21, 2010 12:42 pm

Jason1975 wrote:I've been struggling with trying to keep my AHI under 10 ever since I was put on cpap almost 2 years ago. I have an AutoSet II on auto mode. My pressure settings I had set from 10-20. It seems like I go in spurts. I can have ahi's of 8-10 for a few months straight and then all of a sudden they jump from 14-20. Could allergies cause this or what? I've been trying different ranges and it seems like my machine works better if I have my min. pressure close to my original prescribed pressure. But then again, other nights I think the opposite. When I have my minumum pressure set at 8, my 95th % is like 10 something with high ahi's. When I have my minumum pressure set at 10, my 95th % usually is in the 11 range. When I have my minumum set at 12, sometimes my 95th % stays at 12 and sometimes it goes up to like 13 or 14. Last night, my ahi was 21 with an ai of 4.1. My minumum pressure range was set to 11 and my 95% said 11.6. Any thoughts?? I just can't get any consistancy with anything. I just so discouraged.
I would try 2 weeks at my the PSG recommended pressure.

The S8 is programmed not to respond to apneas appearing above a pressure of 10. It doesn't respond to hypopneas at all. It does respond to flow limited breaths and to snores. Your breathing pattern may simply be of a type the Autoset doesn't understand or respond to properly.

What was your PSG pressure?

If you get nowhere with set pressure, then you may want to consider a different APAP.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

Guest

Re: AHI all over the place lately

Post by Guest » Sat Aug 21, 2010 5:37 pm

Mary Z wrote:Jason1975, I have the same problem and I know how frustrating it can be when you're doing everthing you can. I have been told with the ResMed machine to pay more attention to the AI than the AHI because of the way the ResMed scores hypopneas. Still I know your AI is not in the range you want it to be. I am on Spontaneous Mode with pressures of 24/20 and I have days when I meet the target, but not consistently.
Sorry, no answers, just wanted you to know you're not the only one out there.

I'm glad I'm not the only one struggling with this!! I am so frustrated because I can't get any consistancy to be able to determine anything. I've kept my range at 10 - 20 for a couple months and my numbers are all over the place. My leak rates are consistantly almost non-existant. That's the only thing that is consistant. My prescribed pressure was originally 10 and I have just recently had another titration study. They confirmed 10 was supposedly good enough. However, my numbers are all over the place. My 95th percentile pressure is always at least 11 but sometimes up to 13 or 14 some nights. My numbers aren't any better though regardless. Can allergies make you have an increase in pressure needs? I do have bad allergies but am on Zytec.

Jason1975
Posts: 51
Joined: Sat Apr 24, 2010 7:37 pm
Location: West Central Wisconsin

Re: AHI all over the place lately

Post by Jason1975 » Sat Aug 21, 2010 5:42 pm

Sorry, forgot to sign in before I posted.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan Version 3.11, ResScan Data Card Reader, Aussie Heated Hose, backup Resmed S8 Autoset II for travel.

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: AHI all over the place lately

Post by DoriC » Sat Aug 21, 2010 6:39 pm

What ozij said! Keep us posted.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
ifrimmel
Posts: 119
Joined: Tue Jan 26, 2010 5:00 pm
Location: Atlanta, GA

Re: AHI all over the place lately

Post by ifrimmel » Sat Aug 21, 2010 6:41 pm

HI

Leaks!!?? First place to start looking - have you looked at your detailed leak data - make sure it's all under the line?

In my experience (which is substantially less than the amount of time that you have on the hose - so .. YMMV ) - when the machine is "autoranging" it will try to compensate for leaks which then creates more pressure and more leaks - which creates more pressure and more leaks - etc etc - if you are sleeping and not aware of what's happening it accumulates in the data and shows up as AHI's ..

For me - a tighter range at higher pressure gave me better result - it DOES help being able to look at the data.

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Additional Comments: Started Jan '10 Diag AHI:169. BiPAP BiFlex 1 12/21

User avatar
ifrimmel
Posts: 119
Joined: Tue Jan 26, 2010 5:00 pm
Location: Atlanta, GA

Re: AHI all over the place lately

Post by ifrimmel » Sat Aug 21, 2010 6:55 pm

ozij wrote:
Jason1975 wrote:I've been struggling with trying to keep my AHI under 10 ever since I was put on cpap almost 2 years ago. I have an AutoSet II on auto mode. My pressure settings I had set from 10-20. It seems like I go in spurts. I can have ahi's of 8-10 for a few months straight and then all of a sudden they jump from 14-20. Could allergies cause this or what? I've been trying different ranges and it seems like my machine works better if I have my min. pressure close to my original prescribed pressure. But then again, other nights I think the opposite. When I have my minumum pressure set at 8, my 95th % is like 10 something with high ahi's. When I have my minumum pressure set at 10, my 95th % usually is in the 11 range. When I have my minumum set at 12, sometimes my 95th % stays at 12 and sometimes it goes up to like 13 or 14. Last night, my ahi was 21 with an ai of 4.1. My minumum pressure range was set to 11 and my 95% said 11.6. Any thoughts?? I just can't get any consistancy with anything. I just so discouraged.
I would try 2 weeks at my the PSG recommended pressure.

The S8 is programmed not to respond to apneas appearing above a pressure of 10. It doesn't respond to hypopneas at all. It does respond to flow limited breaths and to snores. Your breathing pattern may simply be of a type the Autoset doesn't understand or respond to properly.

What was your PSG pressure?

If you get nowhere with set pressure, then you may want to consider a different APAP.
Ditto that !

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Additional Comments: Started Jan '10 Diag AHI:169. BiPAP BiFlex 1 12/21

Jason1975
Posts: 51
Joined: Sat Apr 24, 2010 7:37 pm
Location: West Central Wisconsin

Re: AHI all over the place lately

Post by Jason1975 » Sat Aug 21, 2010 7:15 pm

As I said before, my leaks are under control. I also have resscan and check my data every day. My leaks are between .00 and .06 usually. My PSG is 10 and I was on that for most of my time on cpap. I just recently got an auto but used to have an Elite II. My AHI's at pressure 10 averaged between 12 and 18 a night. AI's between 1 and 4 per night. This is the main reason why I got an auto. I think when I go into rem, my pressure needs to be increased. But, because of the A10 algorithm, its not getting them every time. So, I've been bumping up my minimum set pressure, but I can't get anything consistant. Sometimes my readings are good (for me) and then the next night it will be terrible. I like consistancy and right now I'm just not seeing it.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan Version 3.11, ResScan Data Card Reader, Aussie Heated Hose, backup Resmed S8 Autoset II for travel.

User avatar
KatieW
Posts: 1671
Joined: Sun Sep 27, 2009 11:07 am
Location: southern AZ

Re: AHI all over the place lately

Post by KatieW » Sun Aug 22, 2010 9:29 am

Since the S8 does not differentiate between obstructive and central apneas, it could be that some of your AI are centrals.

When I switched to the S9 in apap mode, I found that my apneas were mostly centrals, and that raising the minimum and lowering the maximum, thereby tightening the range helped. Also, lowering the EPR to 1 helped even more.

_________________
Mask: Pico Nasal CPAP Mask with Headgear
Additional Comments: Resmed AirCurve 10 ASV and Humidifier, Oscar for Mac
KatieW

ozij
Posts: 10473
Joined: Fri Mar 18, 2005 11:52 pm

Re: AHI all over the place lately

Post by ozij » Sun Aug 22, 2010 10:03 am

I have a similar problem with the S8 APAP -- waiting for a repeat PSG to get more info about what's going on.
Additionally, I have flow limitations that apparently make the APAP zoom way up - my pressure is less than 10, and my Purtian Bennett machine used to show I got more central apnea above 8. So I'm using the S8 at fixed - looking for the pressure that will give the most consistent sleep quality -- which in my case is correlated with time spent in apnea.

I can't use the HI as an indicator of much -- it will increase when I'm tossing and turning, which has nothing to do with obstructions.

ResScan statistics has a "time in apnea" measure, reporting the amount of time spent in apnea. I use that as an indicator -- shorter is better.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

Jason1975
Posts: 51
Joined: Sat Apr 24, 2010 7:37 pm
Location: West Central Wisconsin

Re: AHI all over the place lately

Post by Jason1975 » Sun Aug 22, 2010 10:53 am

ozij wrote:
ResScan statistics has a "time in apnea" measure, reporting the amount of time spent in apnea. I use that as an indicator -- shorter is better.
That's what I use as well. My average time is between .6 and 1.5 % time in apnea.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan Version 3.11, ResScan Data Card Reader, Aussie Heated Hose, backup Resmed S8 Autoset II for travel.