Does EPR reduce therapy effectiveness?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
joeattardi
Posts: 109
Joined: Wed Jul 07, 2010 5:35 pm
Location: Billerica, MA
Contact:

Does EPR reduce therapy effectiveness?

Post by joeattardi » Thu Aug 05, 2010 10:55 am

Hi everyone! I'm at work right now but can't wait to download ResScan tonight and take a look at my first night's data!
In the meantime, I was wondering today - does the changing pressure of EPR have the potential to reduce the effectiveness of CPAP therapy?
Diagnosed with OSA 2/19/10
Started CPAP therapy 7/6/10 @ 9 cm H2O

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: Does EPR reduce therapy effectiveness?

Post by Slinky » Thu Aug 05, 2010 11:03 am

No. If it did PAP manufacturers would not be allowed to offer it. Most of us end up turning it off, and turning Ramp off too eventually as we get acclimated to and comfortable w/our PAP therapy. The important issue is COMPLIANCE, using that PAP every time we sleep, even for naps. And if EPR, Ramp, C- or A-Flex or Bi-Flex encourages and improves compliance those options are viable and important.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

User avatar
Amigo
Posts: 415
Joined: Sun Oct 23, 2005 4:05 am
Location: NH

Re: Does EPR reduce therapy effectiveness?

Post by Amigo » Thu Aug 05, 2010 11:05 am

Since EPR stands for expiratory pressure relief, I cannot see how it would have a deleterious effect on therapy since it's the pressure on inhalation that keeps your airway open. On the other hand, I think the beneficial effect goes without saying...especially for those whose tolerance for exhaling against a "wind storm" is low.

User avatar
Wulfman
Posts: 12316
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Does EPR reduce therapy effectiveness?

Post by Wulfman » Thu Aug 05, 2010 11:06 am

Yes......potentially.
As has been discussed many times before, EPR drops the pressure on exhale in the number of centimeters according to which setting you select......1 drops 1 cm......2 drops 2 cm......3 drops 3 cm.......for the full exhale period. If the user is prone to having apneas at the point of "end-of-exhale"/"beginning-of-inhale", then you either need to reduce the EPR number or compensate for that with a higher set pressure if you're using CPAP mode or higher minimum pressures if you're using APAP mode.


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
joeattardi
Posts: 109
Joined: Wed Jul 07, 2010 5:35 pm
Location: Billerica, MA
Contact:

Re: Does EPR reduce therapy effectiveness?

Post by joeattardi » Thu Aug 05, 2010 11:18 am

Amigo wrote:Since EPR stands for expiratory pressure relief, I cannot see how it would have a deleterious effect on therapy since it's the pressure on inhalation that keeps your airway open.
I wasn't sure how intelligent the machines are at detecting just when you start exhaling, so I guess I was worried about it dropping the pressure when you aren't really exhaling. Because yeah, like you said, it's inhalation that counts...
Diagnosed with OSA 2/19/10
Started CPAP therapy 7/6/10 @ 9 cm H2O

User avatar
GumbyCT
Posts: 5780
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: Does EPR reduce therapy effectiveness?

Post by GumbyCT » Thu Aug 05, 2010 11:47 am

joeattardi wrote:
Amigo wrote:Since EPR stands for expiratory pressure relief, I cannot see how it would have a deleterious effect on therapy since it's the pressure on inhalation that keeps your airway open.
I wasn't sure how intelligent the machines are at detecting just when you start exhaling, so I guess I was worried about it dropping the pressure when you aren't really exhaling. Because yeah, like you said, it's inhalation that counts...
Not really so. The airway could collapse if/when the exhalation pressure falls below what the individual needs to keep their airway open.

So at a set pressure of say 10 cm an EPR of 1 may have no effect while EPR of 2 or 3 could increase the number of events. This is more pronounced with a BiPap but has the same effect with a CPAP with Cflex/Afelx or EPR.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
joeattardi
Posts: 109
Joined: Wed Jul 07, 2010 5:35 pm
Location: Billerica, MA
Contact:

Re: Does EPR reduce therapy effectiveness?

Post by joeattardi » Thu Aug 05, 2010 11:50 am

GumbyCT wrote: So at a set pressure of say 10 cm an EPR of 1 may have no effect while EPR of 2 or 3 could increase the number of events. This is more pronounced with a BiPap but has the same effect with a CPAP with Cflex/Afelx or EPR.
Hmm, interesting. My current pressure is 9 cm and I believe my EPR is set to 3.
Diagnosed with OSA 2/19/10
Started CPAP therapy 7/6/10 @ 9 cm H2O

User avatar
Amigo
Posts: 415
Joined: Sun Oct 23, 2005 4:05 am
Location: NH

Re: Does EPR reduce therapy effectiveness?

Post by Amigo » Thu Aug 05, 2010 11:52 am

Den is correct, but only careful monitoring of your statistics will give you a clue if changing your settings might make a difference in your therapy.

My advice is to not go looking for problems just yet.

User avatar
Wulfman
Posts: 12316
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Does EPR reduce therapy effectiveness?

Post by Wulfman » Thu Aug 05, 2010 11:54 am

joeattardi wrote:
GumbyCT wrote: So at a set pressure of say 10 cm an EPR of 1 may have no effect while EPR of 2 or 3 could increase the number of events. This is more pronounced with a BiPap but has the same effect with a CPAP with Cflex/Afelx or EPR.
Hmm, interesting. My current pressure is 9 cm and I believe my EPR is set to 3.
Then your "effective" pressure at the point of exhale/inhale is 6 cm. If that isn't enough to keep your airway open, then you need to compensate for it in one of the two ways I mentioned before. It's been noted before that ResMed's EPR is sort of like a mini Bi-Level......but, with Bi-Level therapy, you need to set the EPAP pressure to the point where you eliminate the Apneas.


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

jnk
Posts: 5781
Joined: Mon Jun 30, 2008 3:03 pm

Re: Does EPR reduce therapy effectiveness?

Post by jnk » Thu Aug 05, 2010 2:32 pm

Nestled inside a thread about C-flex was this work of art from RG about EPR . . .

viewtopic.php?p=436272#p436272

User avatar
joeattardi
Posts: 109
Joined: Wed Jul 07, 2010 5:35 pm
Location: Billerica, MA
Contact:

Re: Does EPR reduce therapy effectiveness?

Post by joeattardi » Thu Aug 05, 2010 2:45 pm

jnk wrote:Nestled inside a thread about C-flex was this work of art from RG about EPR . . .

viewtopic.php?p=436272#p436272
Thanks for the link! That was a really good read.
Diagnosed with OSA 2/19/10
Started CPAP therapy 7/6/10 @ 9 cm H2O

jnk
Posts: 5781
Joined: Mon Jun 30, 2008 3:03 pm

Re: Does EPR reduce therapy effectiveness?

Post by jnk » Thu Aug 05, 2010 2:57 pm

She's a gem.

If you want to go deeper, you may enjoy the discussion between two other gems in this forum, -SWS and jdm2857, which encompasses Easy-Breathe and its interaction with EPR . . .

viewtopic.php?f=1&t=44898&p=402793#p402785

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Does EPR reduce therapy effectiveness?

Post by Janknitz » Thu Aug 05, 2010 3:48 pm

There was a doctor who spoke to the NY AWAKE group recently (on YouTube, but I am too lazy to figure out his name or the site right now) who was explaining EPR and one thing I seem to remember is that the effective pressure drop from EPR is not as much as it would seem because there is also the pressure from the exhalation itself.

Inhalation requires muscle action--contraction of the diaphragm and other accessory muscles, but in most of us (some lung pathologies interfere with exhalation) exhalation is mostly passive (as the diaphragm relaxes, it pushes the air out) and exerts its own pressure on the airway--sort of like when you let go of the balloon, the opening will not collapse until all or most of the air is expelled. So even though the EPR may reduce the pressure 1, 2, or 3 cm respectively going in, the actual pressure in the airway does not drop that much since there is also pressure coming out from the exhalation itself.

I may be all wet about this, but I do recall a rather lengthy explanation about that.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
billbolton
Posts: 2264
Joined: Wed Jun 07, 2006 7:46 pm
Location: Sydney, Australia

Re: Does EPR reduce therapy effectiveness?

Post by billbolton » Thu Aug 05, 2010 4:02 pm

Slinky wrote:Most of us end up turning it off, and turning Ramp off too eventually as we get acclimated to and comfortable w/our PAP therapy.
That would be some, not most.

Cheers,

Bill

_________________
MachineMask
Additional Comments: Airmini, Medistrom Pilot 24, CMS 60C Pulse Oximeter, ResScan 6

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: Does EPR reduce therapy effectiveness?

Post by dsm » Thu Aug 05, 2010 4:33 pm

Janknitz wrote:There was a doctor who spoke to the NY AWAKE group recently (on YouTube, but I am too lazy to figure out his name or the site right now) who was explaining EPR and one thing I seem to remember is that the effective pressure drop from EPR is not as much as it would seem because there is also the pressure from the exhalation itself.

Inhalation requires muscle action--contraction of the diaphragm and other accessory muscles, but in most of us (some lung pathologies interfere with exhalation) exhalation is mostly passive (as the diaphragm relaxes, it pushes the air out) and exerts its own pressure on the airway--sort of like when you let go of the balloon, the opening will not collapse until all or most of the air is expelled. So even though the EPR may reduce the pressure 1, 2, or 3 cm respectively going in, the actual pressure in the airway does not drop that much since there is also pressure coming out from the exhalation itself.

I may be all wet about this, but I do recall a rather lengthy explanation about that.
Janknitz

Your logic seems ok to me. In fact, if we place a dial manometer at the machine end of the airhose & another at the mask end, the pressure data becomes quite interesting esp when observing EPR and then EPR turned off.

*************************************************

EPR OFF

With EPR turned off, the machine is basically in cpap mode (lets ignore Auto mode for this discussion). But as you breathe in and out the dial manometer at the mask end will vary around 1-2 even 3 CMs (depending on how fast & forcefully the person breathes) but the dial manometer at the machine end will vary very little.

To go deeper, if the machine is delivering say 10 CMs then as we breathe out the back pressure down the air hose can be seen at the machine end as a very minor pressure increase & if you could hear the blower (they are so quiet these days it is hard to) you would hear it slow a bit as the machine drops speed in order to maintain the 10 CMs pressure delivered at the machine end of the hose.

As we breathe in, the blower will speed up again to make up for any 'potential' drop in pressure created by the suction of breathing in. Put simply the machine will always adjust its blower to keep the delivered pressure at 10 CMs and the act of breathing in an out has the natural effect of increasing & decreasing the back pressure from the mask to the machine.

*************************************************************

EPR ON

With EPR on the EasyBreathe kicks in. Put simply, easy breathe is the way the blower varies its speed as the pressure is dropped on exhale and raised on inhale for EPR. This variation is called the 'shark fin' wave form & the blower speed is adjusted in this pattern which tends to closely mimic normal breathing. Normal breathing in starts quickly & then slows towards full breath in then accelerates out quickly on breathing out. Looking at a time graph, that pattern of pressure rise & fall from the sleeping person looks just like a shark fin hence the original descriptions of it. Early Bilevels tended to drop & raise the ipap=>epap=>ipap=>epap airflow more like a square wave which felt unnatural to the sleeper. EasyBreathe became possible because the blowers used in these new generation of machines, can adjust pressure so quickly & quietly that the designers were able to adjust the speed to exactly match normal breathing patterns.

************************************************************

My way of getting better benefit from EPR.

When I used EPR myself, I found that to get the best benefit I raised my cpap pressure 1 CMs above my titration pressure & then set EPR to 3 & that worked really well. Many people have commented that if they stay with their titration pressure then set EPR = 3, they don't feel as good as when they turn EPR off or to a lower setting & I believe Wulfman explained this point pretty well.

I used EPR as I found that it reduced aerophagia for me compared to using the machine in straight cpap or auto mode (I was originally on 15 CMs and aerophagia was an issue at the pressure).


DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)