Question about humidifier

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
newname
Posts: 391
Joined: Thu Jan 29, 2009 5:00 pm

Question about humidifier

Post by newname » Wed May 19, 2010 8:32 am

When I turn on the humidifier, the hose going in to it blows the water around like waves. As soon as I put the mask on, the water stops blowing around. I just noticed this and was wondering if this means the humidifier is not working properly. I called my dme and they could not give me an answer.

_________________
MaskHumidifier
Additional Comments: sleepyhead software

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: Question about humidifier

Post by DreamStalker » Wed May 19, 2010 8:47 am

With mask off, the blower unit is blowing air at maximum capacity cuz it cannot maintain set pressure ... makes waves on water surface.

With mask on, blower unit is only blowing air sufficient to maintain set pressure (basically, air velocity andvolume is same as your normal breathing) ... water stops blowing around.