Unable to fine tune my setting to improve AHI.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Hodr
Posts: 17
Joined: Sun Mar 28, 2010 7:56 pm

Unable to fine tune my setting to improve AHI.

Post by Hodr » Thu May 13, 2010 8:05 am

I have been on xPAP now for almost 2 months. I have a Resmed S9 Autoset w/ the Climate Control hose and a Quattro full face mask (I can not breathe very well through my nose).

I spent about 10 days with the original setting of 7cm (my sleep setting titration was 6) and found a variance of 0.9 (lowest) to 4.0 (highest) with an average of about 2.0 AHI when sleeping 7.5-8.0 hours (my normal range). Usually an equal split between Apnea's and Hypopnea's, and the Apneas are equal split between obstructive and central.

I usually experience one event during the first 4 hours of sleep, then the rest pile on during the last 2 to 3 hours.

I check the detailed graphs virtually every day and aside from the first night I have almost never have significant leakage, and when I do it is for a very short duration (and does not seem to correlate to events).

So, I have tried many things to lower my AHI, I set the CPAP settings to both 6cm and 8cm (no apparent difference). I set the machine to APAP with a wide range (6-10), a middle range (7-9), and even a short range of 7-8cm. With the APAP it typically stays at the low value until an event then rides the high value the rest of the night.

I spent several days at each setting, and I tend remember waking once or so per night to reposition the hose after turning over. I sleep on my sides, rarely on my back. Nothing I do appears to have much effect on my pattern of AHI's, and all else being equal I will probably stick with a 6cm CPAP setting as it is the easiest to deal with.

Do you guys have any suggestions for my situation? Do you think a different mask may help?

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: Unable to fine tune my setting to improve AHI.

Post by DreamStalker » Thu May 13, 2010 8:39 am

If you are savvy enough to strip actual values out of your Res Scan software, and then plot AHI vs pressure like the plot below in a spread sheet or graphing package -

Image

... You can then optimize your pressure by selecting the pressure at the lowest AHI.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
jmelby
Posts: 463
Joined: Sat Nov 14, 2009 7:52 pm
Location: Minneapolis, MN

Re: Unable to fine tune my setting to improve AHI.

Post by jmelby » Thu May 13, 2010 8:45 am

The real question is how are you feeling? An AHI under 5 is considered normal in any case, so if you are feeling fine, I don't see any reason for you to tweak at all given that your highest AHI was 4. If you are still feeling tired, focusing only on AHI may not be the best approach. So, could you provide more information on why you are concerned?

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 3.12, APAP 9 - 13, no EPR, ClimateControl 75F
(yet another Jeff)

User avatar
KatieW
Posts: 1672
Joined: Sun Sep 27, 2009 11:07 am
Location: southern AZ

Re: Unable to fine tune my setting to improve AHI.

Post by KatieW » Thu May 13, 2010 8:58 am

Actually, I think you're doing pretty well, for 2 months in. Though I certainly understand your wish to fine tune your therapy, because I was the same way. Here are my thought:

Are you using EPR? If so, this decreases your exhale pressure by 1,2 or 3 cm, so I raise my minimum pressure accordingly.

What is your 95% pressure, and Maximum pressure? I would set my minimum pressure 1 cm below the 95% pressure. If I'm hitting my maximum pressure, then raise it by 1 cm. I don't like to leave the maximum wide open because of aerohagia and/or leaks, but have gradually raised by maximum so I can tolerate it. Others are fine with leaving it wide open at 20.

In your Detailed Graphs, zoom in on the obstructive apneas, and look at the pressure immediately before the event, and then what your cpap raised it to. I don't worry about centrals, I only have 1 or 2, if I'm over-tired or stressed. You can do a search on Centrals, to alleviate fear of centrals.

I used to get apneas in the last 2 hours of sleep also, around 4-5 am. I think it's because that's the time of longest and deepest REM, and in the beginning we are REM deprived from years of sleep apnea. During your sleep study, did you have more apneas during REM?

You said you cannot breathe very well through your nose, and that is something to discuss with your doctor or a ENT. That could be affecting your therapy. I just listened to a talk by Dr. Steven Park called Un-Stuff your Stuffy Nose, and he said that breathing through your nose is better--because your sinuses make nitric oxide (which dilate blood vessels), and then you can absorb 20% more oxygen. He discussed different ways to deal with congestion: allergy meds, sinus rinses, Breathe Right strips, and surgery options. He has a very informative website.

Hope this helps.

_________________
Mask: Pico Nasal CPAP Mask with Headgear
Additional Comments: Resmed AirCurve 10 ASV and Humidifier, Oscar for Mac
Last edited by KatieW on Thu May 13, 2010 9:00 am, edited 1 time in total.
KatieW

User avatar
Tielman
Posts: 188
Joined: Wed Jan 06, 2010 11:06 am
Location: Orlando, Fl.
Contact:

Re: Unable to fine tune my setting to improve AHI.

Post by Tielman » Thu May 13, 2010 8:59 am

DreamStalker wrote:... You can then optimize your pressure by selecting the pressure at the lowest AHI.
Would the higher numbers (in your graph) at higher pressure be Centrals, or still OAs?? I've been concerned about increasing my pressure due to the risk of increased centrals (which keep showing up in my Resp. Encore reports) as they are the majority of my Apnea events. Is this just a reporting issue?

Thanks

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure 11.5 cm, Cflex off, Humidity off, or 1 to 2

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: Unable to fine tune my setting to improve AHI.

Post by DreamStalker » Thu May 13, 2010 9:02 am

While how one feels may be important, treatment efficacy is more complex than that.

I lived with untreated OSA for over 20 years and did not notice how bad I felt because my body had gotten used to the condition.

As for AHI under 5 being considered "normal" ... it is NOT. Having apneas is NOT normal anymore than sleeping while water-boarded. The AHI value of 5 is just a standard that the sleep health industry uses to determine if your "un-normal" apnea condition needs to be treated ... more specifically if insurance will cover the cost of treatment.

Having said that, it is true that "feeling" tired has multiple causes of which inadequate CPAP settings may be just one component. Remember, CPAP treatment is NOT just for alleviating a "feeling of tiredness" ... more importantly, CPAP treatment is used to mitigate if not eliminate hypoxia induced tissue damage and hormonal disregulation.
Last edited by DreamStalker on Thu May 13, 2010 9:08 am, edited 1 time in total.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: Unable to fine tune my setting to improve AHI.

Post by DreamStalker » Thu May 13, 2010 9:06 am

Tielman wrote:
DreamStalker wrote:... You can then optimize your pressure by selecting the pressure at the lowest AHI.
Would the higher numbers (in your graph) at higher pressure be Centrals, or still OAs?? I've been concerned about increasing my pressure due to the risk of increased centrals (which keep showing up in my Resp. Encore reports) as they are the majority of my Apnea events. Is this just a reporting issue?

Thanks
In my case, I think they would be OSA since centrals were not significant in my PSG. Nonetheless, centrals should show up with increased pressure on a plot.

I'm not that familiar with the S9 so I'll defer to other members with more experience w/ S9. (BTW - If you have S9 machine, your software must be ResScan ... Not Encore.)
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
echo
Posts: 2400
Joined: Wed Jul 18, 2007 4:20 pm

Re: Unable to fine tune my setting to improve AHI.

Post by echo » Thu May 13, 2010 9:08 am

Hodr wrote:I spent several days at each setting, and I tend remember waking once or so per night to reposition the hose after turning over.
You might want to look into hose management techniques so that changing sleeping position doesn't wake you up - like hanging the hose from your headboard, wall, or ceiling.
PR System One APAP, 10cm
Activa nasal mask + mouth taping w/ 3M micropore tape + Pap-cap + PADACHEEK + Pur-sleep
Hosehead since 31 July 2007, yippie!

Hodr
Posts: 17
Joined: Sun Mar 28, 2010 7:56 pm

Re: Unable to fine tune my setting to improve AHI.

Post by Hodr » Thu May 13, 2010 10:41 am

I appreciate the responses guys. I think I mentioned that while running in APAP mode, after my first event, the machine stays pretty close to pegged to the upper value. The highest I have run it is 10cm, and even at that level I start to feel as though and air bubble is stuck in my throat. I have never tried higher than 10cm because up to that point I saw no change, and I wouldn't want to run the machine open ended, end up breathing against 20cm all night, just to end up with the same numbers that I would get running 6cm.

I wouldn't say I feel refreshed when I wake up, but I do believe I am a bit more rested as I don't seem to be falling asleep as often in the early afternoon.

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: Unable to fine tune my setting to improve AHI.

Post by DreamStalker » Thu May 13, 2010 11:00 am

Hodr wrote:I appreciate the responses guys. I think I mentioned that while running in APAP mode, after my first event, the machine stays pretty close to pegged to the upper value. The highest I have run it is 10cm, and even at that level I start to feel as though and air bubble is stuck in my throat. I have never tried higher than 10cm because up to that point I saw no change, and I wouldn't want to run the machine open ended, end up breathing against 20cm all night, just to end up with the same numbers that I would get running 6cm.

I wouldn't say I feel refreshed when I wake up, but I do believe I am a bit more rested as I don't seem to be falling asleep as often in the early afternoon.
Your mention of air bubble in throat points to additional complicating factor ... have you been checked for GERD?
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
LinkC
Posts: 3154
Joined: Wed Oct 29, 2008 1:06 pm
Location: Amelia Island, FL

Re: Unable to fine tune my setting to improve AHI.

Post by LinkC » Thu May 13, 2010 11:11 am

I might suggest that if you've tried all those different setups in only 2 months, you aren't giving things a chance to stabilize after the change.

Do you even have leaks under control in just 2 mos? Something as simple as a mask leak can toss all your fine-tuning out the window.

Patience, Grasshopper...

_________________
MachineMaskHumidifier
Additional Comments: 11-14 cmH2O
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...

Hodr
Posts: 17
Joined: Sun Mar 28, 2010 7:56 pm

Re: Unable to fine tune my setting to improve AHI.

Post by Hodr » Sun May 16, 2010 5:03 pm

LinkC wrote:I might suggest that if you've tried all those different setups in only 2 months, you aren't giving things a chance to stabilize after the change.

Do you even have leaks under control in just 2 mos? Something as simple as a mask leak can toss all your fine-tuning out the window.

Patience, Grasshopper...
Thanks for the reply.

Leak doesn't appear to be an issue. As I mentioned in my initial post, other than the very first night (where there was some significant leak) I have had virtually no leak issues (as in the my running total median and 95th percentile for leak are both 0.0). I really don't mind having my mask fit extra snug. I get funny looks at work for the red spots on my nose and forehead, but that doesn't bother me (and the tight mask doesn't hurt).

I spent more than 2 weeks with the initial setting, and I believe I gave each subsequent change a week or so. Guess I will just continue to collect data points and see where things fall out.

FYI- Last night (set for CPAP at 7cm) I had one of my best nights (1.1 AHI), the last time I had it in APAP (the 11th) It was set a bit wide at 7-10 and the median was 8.9 while the 95th was 9.9.