What can cause a sudden increase in pressure on CPAP?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Steves396lt1
Posts: 28
Joined: Sun Mar 01, 2009 5:10 pm

What can cause a sudden increase in pressure on CPAP?

Post by Steves396lt1 » Fri May 07, 2010 12:23 pm

Last night I was woken by the machine putting too much air inme. I usually see about 10 (cmH2), but last night it jumped to 13.5, and this is with the machine in APAP mode, humidifier running etc. I do lay on my side now, and the seal is really good after advice fro this board..
Is there anything that can be checked? Filters, Hoses (mine is the style of hose with a thinner tube within it for pressure monitoring to adjust within APAP mode), and my mask is a RESMED Quattro.
Thanks

_________________
Mask
420E CPAP

User avatar
drj130
Posts: 565
Joined: Mon Jun 15, 2009 1:14 pm
Location: Florida

Re: What can cause a sudden increase in pressure on CPAP?

Post by drj130 » Fri May 07, 2010 12:28 pm

If your machine increased pressure from the lower to the higher, that is because it detected one of two things. Either you were having problems while you were sleeping (had an event) or that there was a leak and it increased the pressure to try and keep you breathing like you should. I would check the clinical settings and see what the pressure is set for as the Max.

An APAP when it operates in the Auto mode has 2 settings, the min and max.

If you go into the clinical settings, you should be able to turn off the APAP mode and just have it run as a CPAP to where it will stay at your setting of 10.

David

_________________
Mask: ComfortGel Blue Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Wisp Mask, Pressure is set between 8.5 and 12.5. With a M Series Auto as a backup. Also a CMS-50D+ PulseOX
Look twice, save a life. You never know when you might see a motorcyclist.

jules
Posts: 3304
Joined: Mon Nov 27, 2006 10:51 pm

Re: What can cause a sudden increase in pressure on CPAP?

Post by jules » Fri May 07, 2010 12:34 pm

usually pressure increases are not "sudden" but gradual and many machines have algorithms that will raise the pressure only 3 units in a short period of time -

do you have the software to look closer at the data? I bet there were several events that triggered this.

User avatar
jmelby
Posts: 463
Joined: Sat Nov 14, 2009 7:52 pm
Location: Minneapolis, MN

Re: What can cause a sudden increase in pressure on CPAP?

Post by jmelby » Fri May 07, 2010 1:06 pm

drj130 wrote:...or that there was a leak and it increased the pressure to try and keep you breathing like you should.
It is my understanding that the machine should not increase pressure in response to a leak, but rather it will increase the air flow to attempt to maintain the target pressure... I'm not an expert, so my understanding could be mistaken. I'm curious if some of the forum veterans could clarify... I have seen this stated before that pressure will be increased in response to leaks, but it doesn't seem right to me.

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResScan 3.12, APAP 9 - 13, no EPR, ClimateControl 75F
(yet another Jeff)

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: What can cause a sudden increase in pressure on CPAP?

Post by Wulfman » Fri May 07, 2010 1:38 pm

jmelby wrote:
drj130 wrote:...or that there was a leak and it increased the pressure to try and keep you breathing like you should.
It is my understanding that the machine should not increase pressure in response to a leak, but rather it will increase the air flow to attempt to maintain the target pressure... I'm not an expert, so my understanding could be mistaken. I'm curious if some of the forum veterans could clarify... I have seen this stated before that pressure will be increased in response to leaks, but it doesn't seem right to me.
That's true (generally speaking). I verified that with some of my own experiments once. However, if the "pressure" is not sufficient (as opposed to volume) due to leaks, and/or allows some events to occur, then those events (flow limitations, hypopneas and apneas) will cause the machine to increase pressure......which can increase the leaks and then the cycle continues. At some point, it will hit the upper pressure limit ("runaway") or the leaks will stop or the user wakes up.


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Steves396lt1
Posts: 28
Joined: Sun Mar 01, 2009 5:10 pm

Re: What can cause a sudden increase in pressure on CPAP?

Post by Steves396lt1 » Fri May 07, 2010 10:17 pm

Thanks everyone. I will CPAP it tonight, and see what happens. I dont have software yet, but compliance and hour useage.

_________________
Mask
420E CPAP

jweeks
Posts: 1474
Joined: Tue Feb 10, 2009 8:32 pm
Location: Minneapolis, MN
Contact:

Re: What can cause a sudden increase in pressure on CPAP?

Post by jweeks » Fri May 07, 2010 10:25 pm

Steves396lt1 wrote:Thanks everyone. I will CPAP it tonight, and see what happens. I dont have software yet, but compliance and hour useage.
Hi,

Until you get the software and know what happened, I think I'd consider staying on APAP. If you were having events and need the pressure, then CPAP will not be able to respond. I think you need to know if you are having events, and if you need the extra pressure to prevent them. If you do go to CPAP, then you might want to run at the higher pressure so the machine is always ready to give you treatment. The issue with doing that is that higher pressure can have side effects, too. I think you should keep things the way they are at until you have visibility into the data.

-john-

Steves396lt1
Posts: 28
Joined: Sun Mar 01, 2009 5:10 pm

Re: What can cause a sudden increase in pressure on CPAP?

Post by Steves396lt1 » Fri May 07, 2010 11:02 pm

jweeks wrote:
Steves396lt1 wrote:Thanks everyone. I will CPAP it tonight, and see what happens. I dont have software yet, but compliance and hour useage.
Hi,

Until you get the software and know what happened, I think I'd consider staying on APAP. If you were having events and need the pressure, then CPAP will not be able to respond. I think you need to know if you are having events, and if you need the extra pressure to prevent them. If you do go to CPAP, then you might want to run at the higher pressure so the machine is always ready to give you treatment. The issue with doing that is that higher pressure can have side effects, too. I think you should keep things the way they are at until you have visibility into the data.

-john-
Good point. Maybe in some due time I can get the software. 100 dollars is a lot right now.

_________________
Mask
420E CPAP

User avatar
ozij
Posts: 10453
Joined: Fri Mar 18, 2005 11:52 pm

Re: What can cause a sudden increase in pressure on CPAP?

Post by ozij » Sat May 08, 2010 8:43 am

.
Steves396lt1 wrote:Thanks everyone. I will CPAP it tonight, and see what happens. I dont have software yet, but compliance and hour useage.
The Puritan Bennett 420E which you show in your profile shows neither compliance nor usage on the its screen.
Something doesnt' fit here.

O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

Country4ever
Posts: 1373
Joined: Wed Oct 31, 2007 6:22 pm

Re: What can cause a sudden increase in pressure on CPAP?

Post by Country4ever » Sat May 08, 2010 8:58 am

I hated that aspect of apap. It woke me up all night long. Any little leak and that pressure goes up. You cough, the pressure goes up, you turn, the pressure goes up. I changed to cpap and its great.

_________________
MachineMask
Additional Comments: Resmed HumidAir 11 for humidifier

User avatar
cobra4x4
Posts: 272
Joined: Tue Mar 23, 2010 10:44 pm

Re: What can cause a sudden increase in pressure on CPAP?

Post by cobra4x4 » Sat May 08, 2010 9:00 am

Steves396lt1 wrote: Good point. Maybe in some due time I can get the software. 100 dollars is a lot right now.
Check around online ... I bet the software is available for download somewhere.

If you know the name of the software, do a search in google or whatever search engine.

_________________
Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Secondary Mask: Swift FX

Steves396lt1
Posts: 28
Joined: Sun Mar 01, 2009 5:10 pm

Re: What can cause a sudden increase in pressure on CPAP?

Post by Steves396lt1 » Sat May 08, 2010 11:32 am

cobra4x4 wrote:
Steves396lt1 wrote: Good point. Maybe in some due time I can get the software. 100 dollars is a lot right now.
Check around online ... I bet the software is available for download somewhere.

If you know the name of the software, do a search in google or whatever search engine.
Been searching, cant find it anywhere. ;(

_________________
Mask
420E CPAP

Steves396lt1
Posts: 28
Joined: Sun Mar 01, 2009 5:10 pm

Re: What can cause a sudden increase in pressure on CPAP?

Post by Steves396lt1 » Sat May 08, 2010 11:35 am

ozij wrote:.
Steves396lt1 wrote:Thanks everyone. I will CPAP it tonight, and see what happens. I dont have software yet, but compliance and hour useage.
The Puritan Bennett 420E which you show in your profile shows neither compliance nor usage on the its screen.
Something doesnt' fit here.

O.
All you do is hit the information access button on the bottom left for a few seconds then click through and it shows you all kinds of data.

_________________
Mask
420E CPAP

User avatar
ozij
Posts: 10453
Joined: Fri Mar 18, 2005 11:52 pm

Re: What can cause a sudden increase in pressure on CPAP?

Post by ozij » Sat May 08, 2010 12:38 pm

Steves396lt1 wrote:
ozij wrote:.
Steves396lt1 wrote:Thanks everyone. I will CPAP it tonight, and see what happens. I dont have software yet, but compliance and hour useage.
The Puritan Bennett 420E which you show in your profile shows neither compliance nor usage on the its screen.
Something doesnt' fit here.

O.
All you do is hit the information access button on the bottom left for a few seconds then click through and it shows you all kinds of data.
Please check the machine you put in your profile. The Puritan Bennett 420e does not have a information access button, at all. As a matter of fact, it has not buttons on the bottom left.

Please check the machine next to your bed and the data you get on the screen. If you have the Sandman Auto, there is no reason your data should be limited to leak and compliance.

O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

Steves396lt1
Posts: 28
Joined: Sun Mar 01, 2009 5:10 pm

Re: What can cause a sudden increase in pressure on CPAP?

Post by Steves396lt1 » Sat May 08, 2010 3:40 pm

Give me your email and I will send you a manual for my CPAP 420 E. It has information access buttons and I use it everyday.

_________________
Mask
420E CPAP