- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
Looking at the data from the S9 in the EDFBrowser made me realize something interesting.
You have to convert the *EVE.edf data from one format to another in order to visualize it in the EDFBrowser.
While looking at the various format conversions, I discovered that there is a way to take ascii data and convert it to .edf format.
It should just be a case of taking the data from the contec CMS-60D (or your preferred contec recording model) and converting
the .csv file to .edf format, and then adding the time stamp in such a way that ResScan can visualize it.
I haven't made the attempt yet, but am considering trying to do so. If it could be made to work, it might end up with some amount of skew, so it wouldn't be as accurate as the ResLink method. However, the price is right. Because the edflib is avaialble, it might just be possible to create a software app that would do the manipulation for us quickly from the four file types involved: .spoR, .csv, *SAD.edf, *SAD.crc.
I'd like to get some feedback on whether others think this might be worth the effort or is just a hole in time, waiting for itself to be dug.
It shouldn't void any warranties, since we'd be working with a copy of the SD card -- not the one that we use with the S9.
If nothing else, something like this could still be useful with the EDFBrowser.
You have to convert the *EVE.edf data from one format to another in order to visualize it in the EDFBrowser.
While looking at the various format conversions, I discovered that there is a way to take ascii data and convert it to .edf format.
It should just be a case of taking the data from the contec CMS-60D (or your preferred contec recording model) and converting
the .csv file to .edf format, and then adding the time stamp in such a way that ResScan can visualize it.
I haven't made the attempt yet, but am considering trying to do so. If it could be made to work, it might end up with some amount of skew, so it wouldn't be as accurate as the ResLink method. However, the price is right. Because the edflib is avaialble, it might just be possible to create a software app that would do the manipulation for us quickly from the four file types involved: .spoR, .csv, *SAD.edf, *SAD.crc.
I'd like to get some feedback on whether others think this might be worth the effort or is just a hole in time, waiting for itself to be dug.
It shouldn't void any warranties, since we'd be working with a copy of the SD card -- not the one that we use with the S9.
If nothing else, something like this could still be useful with the EDFBrowser.
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
- alterego61
- Posts: 83
- Joined: Wed Aug 26, 2009 10:11 pm
- Location: Ontario, Canada
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
I think this is a fine idea, if someone has the time to look into it. It will be made easier once we can take a look at a .SAD (s9 oximetry) file from a machine that is hooked up through the oximetry interface to see how the data are laid out.
Coincidentally I have just started looking at the data files from a CMS-50F - there are two: the first is a CSV file that contains two series of values, one for pulse and one for SpO2. The other file is of type .spoR - I haven't figured out yet how it's structured but it looks like some kind of header file that contains the start time etc.
If someone has the programming smarts and a set of equivalent s9 files no doubt the data could be massaged so that they lined up in ResScan by manipulating the start time and either chopping or padding the data values in the CMS file.
Coincidentally I have just started looking at the data files from a CMS-50F - there are two: the first is a CSV file that contains two series of values, one for pulse and one for SpO2. The other file is of type .spoR - I haven't figured out yet how it's structured but it looks like some kind of header file that contains the start time etc.
If someone has the programming smarts and a set of equivalent s9 files no doubt the data could be massaged so that they lined up in ResScan by manipulating the start time and either chopping or padding the data values in the CMS file.
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: APAP 5-7, EPR 1, ClimateLine Hose, ClimateControl Auto 27C/80F, ResScan 3.10 / Win7 64, 16GB SanDisk Ultra Class 4 (15MB/s) SD Card |
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
I have the ResScan SpO2 units & could look at collecting the file from one.
I am sure we can format the data into a file format that ResScan recognizes.
The SpO2 units are the 3012 Nonin xPod/iPod units.
DSM
I am sure we can format the data into a file format that ResScan recognizes.
The SpO2 units are the 3012 Nonin xPod/iPod units.
DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)
- alterego61
- Posts: 83
- Joined: Wed Aug 26, 2009 10:11 pm
- Location: Ontario, Canada
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
Hi DSM, thanks for the feedback. If I read you correctly, you have the oximeter model that the s9 is likely to require (Nonin 3010), but not the s9 oximetry interface - is that right?dsm wrote:I have the ResScan SpO2 units & could look at collecting the file from one.
I am sure we can format the data into a file format that ResScan recognizes.
The SpO2 units are the 3012 Nonin xPod/iPod units.
DSM
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: APAP 5-7, EPR 1, ClimateLine Hose, ClimateControl Auto 27C/80F, ResScan 3.10 / Win7 64, 16GB SanDisk Ultra Class 4 (15MB/s) SD Card |
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
There's more than just start time data there. In Frhed, it looks like a bunch of jumbled patterned letters, but I cannot make out what anything is beyond the first identifying string for date and name. I'm not sure how useful the bulk of the data in the .spoR files will end up being if we cannot manipulate the it into something EDFBrowser can see.alterego61 wrote:...The other file is of type .spoR - I haven't figured out yet how it's structured but it looks like some kind of header file that contains the start time etc.
I wonder if the files are in the .crc and .edf format for the VPAP Adapt SV and the S8. I'll bet they are.dsm wrote:I have the ResScan SpO2 units & could look at collecting the file from one.
I am sure we can format the data into a file format that ResScan recognizes.
The SpO2 units are the 3012 Nonin xPod/iPod units.
DSM
If you're interested in working with some files from the S9 and some corresponding contec oximetry for comparison, I could send you some.
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
Here's some examples of what I was able to do with data from the .csv file and EDFBrowser:
Could be a challenge trying to get it into ResScan though...
Hey! No apneas last night! Only a single hypopnea. It will be interesting to line up hypopneas with oximetry to see how they correspond.
It looks like some of the lines below are derived from an algorithm that compiles some combination of data found here for ResScan.
Two different mask pressures, Exp pressure, RR, Vt, MV -- I wonder what some of these mean. Does anyone have a clue?
One full mask session:

A five-minute interval:

A twenty-minute interval:

Could be a challenge trying to get it into ResScan though...
Hey! No apneas last night! Only a single hypopnea. It will be interesting to line up hypopneas with oximetry to see how they correspond.
It looks like some of the lines below are derived from an algorithm that compiles some combination of data found here for ResScan.
Two different mask pressures, Exp pressure, RR, Vt, MV -- I wonder what some of these mean. Does anyone have a clue?
One full mask session:

A five-minute interval:

A twenty-minute interval:

_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
- alterego61
- Posts: 83
- Joined: Wed Aug 26, 2009 10:11 pm
- Location: Ontario, Canada
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
DD that's a pretty neat result.
Did you just upload a set of S9 files together with the CMS files? It was that easy?
Did you just upload a set of S9 files together with the CMS files? It was that easy?
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: APAP 5-7, EPR 1, ClimateLine Hose, ClimateControl Auto 27C/80F, ResScan 3.10 / Win7 64, 16GB SanDisk Ultra Class 4 (15MB/s) SD Card |
- alterego61
- Posts: 83
- Joined: Wed Aug 26, 2009 10:11 pm
- Location: Ontario, Canada
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
Have you compared them to the same charts in ResScan? It should be easy from that to figure out what they are.DreamDiver wrote: Two different mask pressures, Exp pressure, RR, Vt, MV -- I wonder what some of these mean. Does anyone have a clue?
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: APAP 5-7, EPR 1, ClimateLine Hose, ClimateControl Auto 27C/80F, ResScan 3.10 / Win7 64, 16GB SanDisk Ultra Class 4 (15MB/s) SD Card |
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
No, I had to massage the data. I created the .edf file from the oximetry .csv file (using the EDFBrowser conversion tools) for the same time frame as the mask session. The oximetry data is actually longer than one session, so I zoomed in to this particular session.alterego61 wrote:Did you just upload a set of S9 files together with the CMS files? It was that easy?
That's just it. They don't exactly match. It's why I suspect they have incorporated some sort of algorithm that takes an average, or does some other sample filtering to create the graphs we see for our use. Either that or they just don't choose to show us all the data, but make it available for those in a real lab who have an interest in seeing exactly what's involved.alterego61 wrote:Have you compared them to the same charts in ResScan? It should be easy from that to figure out what they are.DreamDiver wrote: Two different mask pressures, Exp pressure, RR, Vt, MV -- I wonder what some of these mean. Does anyone have a clue?
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
- alterego61
- Posts: 83
- Joined: Wed Aug 26, 2009 10:11 pm
- Location: Ontario, Canada
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
DreamDiver wrote: Two different mask pressures, Exp pressure, RR, Vt, MV -- I wonder what some of these mean. Does anyone have a clue?
alterego61 wrote: Have you compared them to the same charts in ResScan? It should be easy from that to figure out what they are.
Oops, sorry DD, I didn't read your post closely enough.DreamDiver wrote:That's just it. They don't exactly match. It's why I suspect they have incorporated some sort of algorithm that takes an average, or does some other sample filtering to create the graphs we see for our use. Either that or they just don't choose to show us all the data, but make it available for those in a real lab who have an interest in seeing exactly what's involved.
A quick web search yields: Vt = Tidal Volume, RR = Respiratory Rate, MV = Minute Volume, Exp Pressure = Expiratory Pressure
I'm not sure about the two mask pressures
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: APAP 5-7, EPR 1, ClimateLine Hose, ClimateControl Auto 27C/80F, ResScan 3.10 / Win7 64, 16GB SanDisk Ultra Class 4 (15MB/s) SD Card |
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
Doh! I should have looked closer. Thanks alterego61.alterego61 wrote:A quick web search yields: Vt = Tidal Volume, RR = Respiratory Rate, MV = Minute Volume, Exp Pressure = Expiratory Pressure
I'm not sure about the two mask pressures
So that means they've elected not to show Tidal Volume, Resp Rate or Expiratory Pressure for the S9 Autoset in ResScan 3.10, even though they're recording the data. I wonder why they did that?
You'll notice the S9 data in ResScan has AHI - a graph of data that does not show up here. Perhaps I missed an edf file with the data...
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
DD
Interesting info.
Will try to get back & look more into it & also see if I can extract a data file from a Nonin 3012 unit.
DSM
Interesting info.
Will try to get back & look more into it & also see if I can extract a data file from a Nonin 3012 unit.
DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
alterego61 wrote:Hi DSM, thanks for the feedback. If I read you correctly, you have the oximeter model that the s9 is likely to require (Nonin 3010), but not the s9 oximetry interface - is that right?dsm wrote:I have the ResScan SpO2 units & could look at collecting the file from one.
I am sure we can format the data into a file format that ResScan recognizes.
The SpO2 units are the 3012 Nonin xPod/iPod units.
DSM
I have Nonin models 3011, 3012 & 3212 but can't see a 3010 listed anywhere ? was wondering where that info came from ?.
I finally discovered a couple of months back that the Reslink uses the 3012 (& by default the 3212 as well). The 3010 sort of implies
a lot lower sampling rate. The 3011 does 3 bytes per second while the 3012 delivers 255 bytes of data per second. The 3010 thus
by deduction looks like an older model that samples at a lot lower rate. I would have guessed that the S9 used the 3012.
DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)
- DreamDiver
- Posts: 3082
- Joined: Thu Oct 04, 2007 11:19 am
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
Given that the Contec firmware does things at 1hz rather than 3hz in the Nonin model, as suggested by dsm, I'm not sure we would be able to rig up software to interface with ResScan 3.10. At least, I'm pretty sure this is a level of complexity that is beyond the scope of my current capabilities. The fact that there is an entire pane just for oximetry suggests the data may need to be in a very specific format in order for information to show properly in that pane. However, at least it is nice to be able to take the data from the S9 and pair it with contec oximetry using EDFBrowser.
_________________
| Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions |
| Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF |
- alterego61
- Posts: 83
- Joined: Wed Aug 26, 2009 10:11 pm
- Location: Ontario, Canada
Re: Reverse-engineering Oximetry for ResScan 3.10. Can we do it?
I don't have the files with me here, but the EVE file contains the event data - perhaps it is calculated from that?DreamDiver wrote:Doh! I should have looked closer. Thanks alterego61.alterego61 wrote:A quick web search yields: Vt = Tidal Volume, RR = Respiratory Rate, MV = Minute Volume, Exp Pressure = Expiratory Pressure
I'm not sure about the two mask pressures
So that means they've elected not to show Tidal Volume, Resp Rate or Expiratory Pressure for the S9 Autoset in ResScan 3.10, even though they're recording the data. I wonder why they did that?
You'll notice the S9 data in ResScan has AHI - a graph of data that does not show up here. Perhaps I missed an edf file with the data...
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: APAP 5-7, EPR 1, ClimateLine Hose, ClimateControl Auto 27C/80F, ResScan 3.10 / Win7 64, 16GB SanDisk Ultra Class 4 (15MB/s) SD Card |