POLL: Machine Pressure is not accurate?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.

Have you ever tested your Machines pressure accuracy?

No - don't want to. I trust that the pressure is correct.
13
16%
No - But I would like to.
34
43%
Yes - And it was working the way it should - EVERYTHING WAS OKAY
30
38%
Yes - And the pressure was higher than it should be
2
3%
Yes - And the pressure was lower than it should be
0
No votes
 
Total votes: 79

User avatar
blakepro
Posts: 507
Joined: Thu Oct 08, 2009 1:26 pm
Location: Utah

POLL: Machine Pressure is not accurate?

Post by blakepro » Tue Apr 20, 2010 1:45 pm

I've noticed some discussion about building your own manometer to test your machine to make sure its putting out the correct pressure. I know you can also purchase them.

I'm just curious.

Has anyone here gone to the trouble to test their machine and found out that it actually was putting out the wrong pressure? (example: Its set to output a pressure of 10 but instead its pumping out a 12)

If so, what did you do about it?

I've attached a poll to help figure this out.
Last edited by blakepro on Tue Apr 20, 2010 5:24 pm, edited 1 time in total.

User avatar
LSAT
Posts: 13392
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Machine Pressure is not accurate?

Post by LSAT » Tue Apr 20, 2010 2:53 pm

I can't believe that ANY DME wouldn't do this as a free service....it take about 2 minutes.

User avatar
LinkC
Posts: 3154
Joined: Wed Oct 29, 2008 1:06 pm
Location: Amelia Island, FL

Re: Machine Pressure is not accurate?

Post by LinkC » Tue Apr 20, 2010 4:00 pm

You need a choice: I don't think it matters if my machine is accurate.

Your titrated pressure is just a starting point. You should be watching your numbers and homing in on the optimal setting...regardless of the actual empirical pressure that setting represents. Once you've found the setting that works for you, who cares what the number is?

_________________
MachineMaskHumidifier
Additional Comments: 11-14 cmH2O
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...

User avatar
blakepro
Posts: 507
Joined: Thu Oct 08, 2009 1:26 pm
Location: Utah

Re: Machine Pressure is not accurate?

Post by blakepro » Tue Apr 20, 2010 4:33 pm

LinkC wrote:You need a choice: I don't think it matters if my machine is accurate.

Your titrated pressure is just a starting point. You should be watching your numbers and homing in on the optimal setting...regardless of the actual empirical pressure that setting represents. Once you've found the setting that works for you, who cares what the number is?
Thats a good point, although the number is helpful for people that use more than one machine, or people that get their optimal pressures figured out on a loaner auto and then have to use their c-pap on the 90(95) % pressure.

Its just helpful to have some sort of standard for measurement so that you can track and discuss your therapy with medical professionals and just keep track of things your self in some sort of standardized measurement system.

I do admit, if you have an auto, it becomes less important since it will ramp up the pressure as needed.

I think the other concern people sometimes have is that perhaps over time, the machine becomes worn out and is either not accurate, OR looses accuracy during a session. (slowly loosing pressure over the course of several hours but returns to proper pressure after it has rested during the day) I'm not sure how likely that is, but I think I've read a few posts of people wondering about this.

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: POLL: Machine Pressure is not accurate?

Post by Wulfman » Tue Apr 20, 2010 5:46 pm

blakepro wrote:I've noticed some discussion about building your own manometer to test your machine to make sure its putting out the correct pressure. I know you can also purchase them.

I'm just curious.

Has anyone here gone to the trouble to test their machine and found out that it actually was putting out the wrong pressure? (example: Its set to output a pressure of 10 but instead its pumping out a 12)

If so, what did you do about it?

I've attached a poll to help figure this out.
I tested one machine one time using the (cheapest) method I linked to in this thread.

viewtopic.php?f=1&t=49152&p=450105&hili ... er#p450105

And, it indicated the pressure in my machine was correct.


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
6PtStar
Posts: 2659
Joined: Mon Jun 18, 2007 1:58 pm
Location: Texas, Is there any other place?

Re: POLL: Machine Pressure is not accurate?

Post by 6PtStar » Tue Apr 20, 2010 10:06 pm

I bought the Fisher & Paykel Water Manometer and have checked 4 machines. Two of them were older machines and two were fairly new. They were all right on the correct pressure.

Jerry

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: 11cm/H2O, Encore Pro 1.8i, Pro Analyzer, Encore Viewer1.0 - 3 Remstar Pro2's, 1 Remstar Auto
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS

User avatar
blakepro
Posts: 507
Joined: Thu Oct 08, 2009 1:26 pm
Location: Utah

Re: POLL: Machine Pressure is not accurate?

Post by blakepro » Wed Apr 21, 2010 12:48 am

6PtStar wrote:I bought the Fisher & Paykel Water Manometer and have checked 4 machines. Two of them were older machines and two were fairly new. They were all right on the correct pressure.

Jerry
Thanks for the confirmation. My suspicion is that this will overwhelmingly be the case, but it will be interesting to see if there are any people that had a different story.

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: POLL: Machine Pressure is not accurate?

Post by DreamDiver » Wed Apr 21, 2010 12:06 pm

I've checked both my machines with manometers at the sleep lab and using my home-made manometer. (Yes, I got it working correctly.)
My M-Series Pro was off by just shy of half a cm lower than listed. The S9 is dead on target. I tested both at 11cm and 16cm.

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF

User avatar
reitzell
Posts: 56
Joined: Wed Dec 23, 2009 9:30 pm

Re: POLL: Machine Pressure is not accurate?

Post by reitzell » Wed Apr 21, 2010 9:15 pm

I was sure that my cpap when powered directly via a DC power source was giving me less pressure then via an AC power source. I tested it and nope the pressure was exactly the same... about 1/4cm high on both.

Image

Image

_________________
Mask
Additional Comments: smartcodes

User avatar
park_ridge_dave
Posts: 234
Joined: Sat Apr 03, 2010 5:46 am
Location: Chicago Burbs

Re: POLL: Machine Pressure is not accurate?

Post by park_ridge_dave » Thu Apr 22, 2010 12:57 pm

Your Poll Got me to thinking Maybe I should check it. So I did.

I used a "machinist's scale" Graduated in 64ths of an inch (It's what I had, OK?)

Anyway, found the procedure in my "Provider's Manual" and checked it 7.5 cm Spot on and at 12 cm it was 12.065 cm +/- my eyesight

In other words there goes my excuses

I am not surprised at the results, the manufacturers must have a stated tolerance for that reading for the FDA; and current Good Manufacturing Practices probably indicate that you have to be within +/- minus a tenth of a cm or better (just my SWAG based on my experience with the FDA inspectors).

Now I plan to repeat the checks every 6 months or so to monitor performance for "issues".

Cheers,

Dave

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: I have EncoreViewer, but, the Best software is Sleepy Head for MAC
"S/He who has but a thousand friends has not a friend to spare"
Let's be careful out there! Because no matter where you go..... There you are :lol:

User avatar
Muse-Inc
Posts: 4382
Joined: Fri Jun 12, 2009 8:44 pm
Location: Atlanta, GA

Re: POLL: Machine Pressure is not accurate?

Post by Muse-Inc » Thu Apr 22, 2010 8:47 pm

My equipment comes along with me every visit to my sleep doc (I don't go often) for checking and pressure testing. Even tho I didn't buy my APAP from him, he tested & set it up initially...well, his RT did . Seeing him in a few wks, so it'll all go in with me: Escape (backup for now), AutoSet, power cords, humidifiers, reports, hoses, mask, the whole shebang.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.

User avatar
blakepro
Posts: 507
Joined: Thu Oct 08, 2009 1:26 pm
Location: Utah

Re: POLL: Machine Pressure is not accurate?

Post by blakepro » Fri Apr 23, 2010 12:39 pm

Muse-Inc wrote:My equipment comes along with me every visit to my sleep doc (I don't go often) for checking and pressure testing. Even tho I didn't buy my APAP from him, he tested & set it up initially...well, his RT did . Seeing him in a few wks, so it'll all go in with me: Escape (backup for now), AutoSet, power cords, humidifiers, reports, hoses, mask, the whole shebang.
Thats probably a very good idea. Does your doc inspect your equipment when you bring it in?

User avatar
Hawthorne
Posts: 3972
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: POLL: Machine Pressure is not accurate?

Post by Hawthorne » Fri Apr 23, 2010 2:48 pm

I had my old Remstar Pro tested just after my last sleep study. It was a free service by the Provider.

I used that machine EVERY NIGHT and EVERY NAP from September 2002 until February 2008 and it was still right on!

After I bought my M Series Auto, I bought one of those little pressure gauges and I check the M Series one in a while. It has always been right on.

I am assuming that little pressure gauge device is fairly accurate.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap

User avatar
LinkC
Posts: 3154
Joined: Wed Oct 29, 2008 1:06 pm
Location: Amelia Island, FL

Re: POLL: Machine Pressure is not accurate?

Post by LinkC » Fri Apr 23, 2010 6:06 pm

Be sure to use distilled water in your manometers as the dissolved minerals in tap water will make it slightly heavier...

_________________
MachineMaskHumidifier
Additional Comments: 11-14 cmH2O
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...

User avatar
Muse-Inc
Posts: 4382
Joined: Fri Jun 12, 2009 8:44 pm
Location: Atlanta, GA

Re: POLL: Machine Pressure is not accurate?

Post by Muse-Inc » Fri Apr 23, 2010 7:13 pm

blakepro wrote:...Does your doc inspect your equipment when you bring it in?
Yup, tests everthing, examines/looks for leaks, holes, tears, the whole shebang. Took me several visits to remember to bring EVERYTHING for them to checkout...finally just made a checkoff list.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.