Well I'm new to this whole CPAP thing, about a month into 'treatment'. I was using a RESMED S8 Escape II. But when I finally received my scrip and saw my Health provider here in Canada I ended up getting an S9 Elite with the climatline tube (Based on favorable reviews I found here). Anyways my question is, how do you set the EPR rate (1,2,3)? on the S8 I was able to scroll through the menu and change it in a snap.
TIA
J
Setting S9 EPR??
Re: Setting S9 EPR??
EPR has to be set to patient in the clinical manual for you to change it. To access the clinical manual press and hold the round button and the 2 check marks button for 3 seconds. The EPR level setting should be on the first screen. When level is set to patient the patient setup menu will include EPR rate.JafaCakes wrote:Anyways my question is, how do you set the EPR rate (1,2,3)? on the S8 I was able to scroll through the menu and change it in a snap.
You can download the clinical manual from the link in this post.