CompSA, Hypoventilation, PLMs, Bradycardia and Hypotension?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
blizzardboy
Posts: 368
Joined: Mon Feb 15, 2010 12:13 am
Location: South Australia

Re: Or Maybe Not...

Post by blizzardboy » Fri Jun 11, 2010 5:53 am

Muffy from Public Library wrote:
Muffy wrote:...I think the command was format c: or something like that...
Well that didn't work out quite as well as I had hoped...
Machine: Resmed VPAP Adapt SV Enhanced
Mask: Resmed Quattro/Mirage Liberty
Humidifier: Resmed H2i
Other comments: Sleepzone heated tube; CMS-50E SPO2 & CMS-60C ABPM

User avatar
blizzardboy
Posts: 368
Joined: Mon Feb 15, 2010 12:13 am
Location: South Australia

Re: More Waveform Junk

Post by blizzardboy » Fri Jun 11, 2010 5:57 am

Muffy wrote:However, the attack proves ineffective as not only the flow limitation gets progressively worse, but totally fails as bb goes to arousal (EMG increases and the EEG amplitude decreases as sleep lightens from N3 to N2).
Hi Muffy, Thank you for this figure. So does this example constitute a RERA? Or is it possible that I am being aroused by the marauding pirate armed with the pneumatic dillon?
Last edited by blizzardboy on Fri Jun 11, 2010 6:50 am, edited 2 times in total.
Machine: Resmed VPAP Adapt SV Enhanced
Mask: Resmed Quattro/Mirage Liberty
Humidifier: Resmed H2i
Other comments: Sleepzone heated tube; CMS-50E SPO2 & CMS-60C ABPM

User avatar
blizzardboy
Posts: 368
Joined: Mon Feb 15, 2010 12:13 am
Location: South Australia

Re: More Waveform Junk

Post by blizzardboy » Fri Jun 11, 2010 6:02 am

-SWS wrote:Drives home the point of making sure ASV base pressure adequately addresses the obstructive component...
Hi -SWS, Gulp...not sure if the paint on my bedroom wall could handle too much more EEP, certainly have to turn the pilot light off in my gas heater before I hosed up.
Machine: Resmed VPAP Adapt SV Enhanced
Mask: Resmed Quattro/Mirage Liberty
Humidifier: Resmed H2i
Other comments: Sleepzone heated tube; CMS-50E SPO2 & CMS-60C ABPM

User avatar
Muffy
Posts: 960
Joined: Tue Apr 21, 2009 5:28 am
Location: Schenectady, New York

Re: More Waveform Junk

Post by Muffy » Fri Jun 11, 2010 6:51 am

blizzardboy wrote:So does this example constitute a RERA?
With only a 1% desaturation, no significant drop in waveform amplitude and a little increase in thoracic effort there, yep, that a RERA.

One RERA.
blizzardboy wrote:Or is it possible that I am being aroused by the marauding pirate armed with the pneumatic dillon?
That's disgusting.

Muffy
________________________________

Machine: Dell Dimension 8100
Mask: 3M N-95 (during flu season)
Humidifier: Avoided, tends to make me moldy
Software: XP Pro
Additional Comments: You can't find a solution when you don't know the problem

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: More Waveform Junk

Post by SleepingUgly » Fri Jun 11, 2010 9:31 am

blizzardboy wrote:Or is it possible that I am being aroused by the marauding pirate armed with the pneumatic dillon?
That's disgusting.

Muffy
Did he say a pneumatic dildo?!
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Re: More Waveform Junk

Post by -SWS » Fri Jun 11, 2010 11:48 am

blizzardboy wrote:
-SWS wrote:Drives home the point of making sure ASV base pressure adequately addresses the obstructive component...
Hi -SWS, Gulp...not sure if the paint on my bedroom wall could handle too much more EEP, certainly have to turn the pilot light off in my gas heater before I hosed up.
Not to worry, BB. I only meant that as a general principle.
SleepingUgly wrote:
blizzardboy wrote:Or is it possible that I am being aroused by the marauding pirate armed with the pneumatic dillon?
That's disgusting.

Muffy
Did he say a pneumatic dildo?!
No, that was merely the risque image Muffy conjured up---on purpose, no doubt! The viewer ratings on this thread are now bound to get even better...

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: More Waveform Junk

Post by SleepingUgly » Fri Jun 11, 2010 3:14 pm

-SWS wrote:Did he say a pneumatic dildo?!
No, that was merely the risque image Muffy conjured up---on purpose, no doubt! The viewer ratings on this thread are now bound to get even better...
Well, I looked up "dillon" on the web dictionary, and there is no such word, so I'm pretty sure he meant dildo. But a "pneumatic dildo"?!!?! I have heard of sword swallowing, but this is the first I've heard of dildo swallowing. I wonder if it's an Australian thing. It's probably much safer than sword swallowing.
Never put your fate entirely in the hands of someone who cares less about it than you do. --Sleeping Ugly

User avatar
blizzardboy
Posts: 368
Joined: Mon Feb 15, 2010 12:13 am
Location: South Australia

Re: More Waveform Junk

Post by blizzardboy » Fri Jun 11, 2010 5:05 pm

SleepingUgly wrote:
-SWS wrote:Did he say a pneumatic dildo?!
No, that was merely the risque image Muffy conjured up---on purpose, no doubt! The viewer ratings on this thread are now bound to get even better...
Well, I looked up "dillon" on the web dictionary, and there is no such word, so I'm pretty sure he meant dildo. But a "pneumatic dildo"?!!?! I have heard of sword swallowing, but this is the first I've heard of dildo swallowing. I wonder if it's an Australian thing. It's probably much safer than sword swallowing.
Hi SU, I think I have worked out the cause for confusion...you must be thinking of the electro-optic Phalanx(MK) (or is that Phall(US) ), the upgrade version of the pneumatic dillon. BTW, we 'Strines prefer to swallow kangaroos (whole), its more orga(s)nic.
Machine: Resmed VPAP Adapt SV Enhanced
Mask: Resmed Quattro/Mirage Liberty
Humidifier: Resmed H2i
Other comments: Sleepzone heated tube; CMS-50E SPO2 & CMS-60C ABPM

User avatar
Muffy
Posts: 960
Joined: Tue Apr 21, 2009 5:28 am
Location: Schenectady, New York

Re: More Waveform Junk

Post by Muffy » Fri Jun 11, 2010 5:55 pm

-SWS wrote:As an isolated incident above, that's not such a big deal. However, tasking traditional ASV (w/ fixed base pressure) to automatically address a significant outstanding obstructive component is not recommended by manufacturers.
And get a load of this one (again, just one):

Image

An unambiguous obstructive apnea, unaffected by the ASV attack.

Muffy
________________________________

Machine: Dell Dimension 8100
Mask: 3M N-95 (during flu season)
Humidifier: Avoided, tends to make me moldy
Software: XP Pro
Additional Comments: You can't find a solution when you don't know the problem

User avatar
Muffy
Posts: 960
Joined: Tue Apr 21, 2009 5:28 am
Location: Schenectady, New York

Even More Waveform Junk

Post by Muffy » Fri Jun 11, 2010 6:07 pm

Here's that area where the ASV (and/or patient) appears to stutter, more clearly seen in the CPAP channel:

Image

Just about every breath gets a double hit.

Muffy
________________________________

Machine: Dell Dimension 8100
Mask: 3M N-95 (during flu season)
Humidifier: Avoided, tends to make me moldy
Software: XP Pro
Additional Comments: You can't find a solution when you don't know the problem

User avatar
Muffy
Posts: 960
Joined: Tue Apr 21, 2009 5:28 am
Location: Schenectady, New York

Yikes!!!

Post by Muffy » Fri Jun 11, 2010 6:09 pm

OMIGOD!!! 973 VIEWS!!! THIS IS WORSE THAN MY OTHER PHOBIA!!!!
________________________________

Machine: Dell Dimension 8100
Mask: 3M N-95 (during flu season)
Humidifier: Avoided, tends to make me moldy
Software: XP Pro
Additional Comments: You can't find a solution when you don't know the problem

User avatar
blizzardboy
Posts: 368
Joined: Mon Feb 15, 2010 12:13 am
Location: South Australia

Re: Even More Waveform Junk

Post by blizzardboy » Fri Jun 11, 2010 7:12 pm

Muffy wrote:Just about every breath gets a double hit.
I am trying to understand what the algorithm is attempting to acheive during these periods of "double hitting". I suppose, given the flow-metering nature of CPAP, that mid-inhalation the machine senses a partial exhalation followed by the completion of the inhalation cycle i.e. stutter, thus demonstrating the rapid nature of the ASV response to respiratory variations and its charter to maintain tidal volume. Am I dreaming that I am a steam train?
Machine: Resmed VPAP Adapt SV Enhanced
Mask: Resmed Quattro/Mirage Liberty
Humidifier: Resmed H2i
Other comments: Sleepzone heated tube; CMS-50E SPO2 & CMS-60C ABPM

User avatar
blizzardboy
Posts: 368
Joined: Mon Feb 15, 2010 12:13 am
Location: South Australia

Re: Yikes!!!

Post by blizzardboy » Fri Jun 11, 2010 7:17 pm

Muffy wrote:OMIGOD!!! 973 VIEWS!!! THIS IS WORSE THAN MY OTHER PHOBIA!!!!
Maybe I have been misguided about the meaning of dillon. Maybe you believe that Dillon (78393) will arrive as le millième voir is eclipsed? Now that is scary...
Machine: Resmed VPAP Adapt SV Enhanced
Mask: Resmed Quattro/Mirage Liberty
Humidifier: Resmed H2i
Other comments: Sleepzone heated tube; CMS-50E SPO2 & CMS-60C ABPM

unadog
Posts: 320
Joined: Tue Feb 16, 2010 4:41 pm

Re: CompSA, Hypoventilation, PLMs, Bradycardia and Hypotension?

Post by unadog » Fri Jun 11, 2010 7:18 pm

blizzardboy wrote: Or is it possible that I am being aroused by the marauding pirate armed with the pneumatic dillon?
I think BB just doesn't know how to spell. But it is still **so** ambigous! We'll let SU decide:

1) Dillion guitars were founded here in Canada and is the only place you can still buy the original Dillion guitars from the original source, Dillion Guitars Canada

2) dillion: n. Another word for male genitalia (i.e. penis)

Same thing, right?
VPAP ASV: BiPaP ASV: Quattro FF: Activa LT: Swift FX

User avatar
SleepingUgly
Posts: 4690
Joined: Sat Nov 28, 2009 9:32 pm

Re: CompSA, Hypoventilation, PLMs, Bradycardia and Hypotension?

Post by SleepingUgly » Fri Jun 11, 2010 7:27 pm

unadog wrote:
blizzardboy wrote: Or is it possible that I am being aroused by the marauding pirate armed with the pneumatic dillon?
I think BB just doesn't know how to spell. But it is still **so** ambigous! We'll let SU decide: