Smart Start/Stop

Post by glen0901 » Sat Apr 03, 2010 5:48 pm

Hello, I had been using my Resmed CPAP MAchine S9 series for 2 weeks and its great since its so quiet. My only problem is that the smartstart/stop was disable for some reasons I do not know. Before once I wear my mask the machine will start immediately but now I am doing manual by pressing the start/stop button. Is there anybody who can help me to enable the smartstop/start.

Thank you.

bigk
Posts: 608
Joined: Tue Jan 22, 2008 5:35 pm
Location: Central Coast near Sydney Australia

Re: Smart Start/Stop

Post by bigk » Sat Apr 03, 2010 10:10 pm

Access the clinical menu (Press the setup button and the rotating button for 3 seconds) In there you can set Autostart to either Patient or On.

_________________
MaskHumidifier
Additional Comments: S9 Autoset, H5i, Resmed Swift FX, Resscan 3.10
Resmed AutoSet Spirit II flow generator (Backup)
Resmed H3i Humidifier
Swift FX
Mirage Swift LT Nasal Pillows
Activa Nasal Mask
Resscan Software with USB + Data Card
SPO 7500 Pulse-Ox
MedicMon Blood Pressure Monitor
Aussie Heated Hose

User avatar
Muse-Inc
Posts: 4382
Joined: Fri Jun 12, 2009 8:44 pm
Location: Atlanta, GA

Re: Smart Start/Stop

Post by Muse-Inc » Sat Apr 03, 2010 10:20 pm

The MASK ALERT will not alarm with the auto start/stop enabled. Personally, I'd rather the mask alert was working and pushing on/off is not a big deal for me as I use the mask fit feature every night.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.

bigk
Posts: 608
Joined: Tue Jan 22, 2008 5:35 pm
Location: Central Coast near Sydney Australia

Re: Smart Start/Stop

Post by bigk » Sat Apr 03, 2010 10:29 pm

There is no such thing as mask alert on the S9.

_________________
MaskHumidifier
Additional Comments: S9 Autoset, H5i, Resmed Swift FX, Resscan 3.10
Resmed AutoSet Spirit II flow generator (Backup)
Resmed H3i Humidifier
Swift FX
Mirage Swift LT Nasal Pillows
Activa Nasal Mask
Resscan Software with USB + Data Card
SPO 7500 Pulse-Ox
MedicMon Blood Pressure Monitor
Aussie Heated Hose

User avatar
Wulfman
Posts: 12316
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Smart Start/Stop

Post by Wulfman » Sat Apr 03, 2010 10:29 pm

glen0901 wrote:Hello, I had been using my Resmed CPAP MAchine S9 series for 2 weeks and its great since its so quiet. My only problem is that the smartstart/stop was disable for some reasons I do not know. Before once I wear my mask the machine will start immediately but now I am doing manual by pressing the start/stop button. Is there anybody who can help me to enable the smartstop/start.

Thank you.
Are you using a full face mask?
I don't know about the new S9 machines, but in the previous models, if the mask selector was set to MIR FULL, it disabled the smartstart feature.

From the troubleshooting section of the S8 Auto manual:

Use of a ResMed Full Facemask.

SmartStart does not work with a full-face mask as the
anti-asphyxia valve will not allow sufficiently high
pressure on exhalation.


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

bigk
Posts: 608
Joined: Tue Jan 22, 2008 5:35 pm
Location: Central Coast near Sydney Australia

Re: Smart Start/Stop

Post by bigk » Sat Apr 03, 2010 10:30 pm

That's right Den - S9 is the same as the S8.

_________________
MaskHumidifier
Additional Comments: S9 Autoset, H5i, Resmed Swift FX, Resscan 3.10
Resmed AutoSet Spirit II flow generator (Backup)
Resmed H3i Humidifier
Swift FX
Mirage Swift LT Nasal Pillows
Activa Nasal Mask
Resscan Software with USB + Data Card
SPO 7500 Pulse-Ox
MedicMon Blood Pressure Monitor
Aussie Heated Hose

User avatar
Muse-Inc
Posts: 4382
Joined: Fri Jun 12, 2009 8:44 pm
Location: Atlanta, GA

Re: Smart Start/Stop

Post by Muse-Inc » Sat Apr 03, 2010 10:40 pm

bigk wrote:There is no such thing as mask alert on the S9.
NO? Wow, mine's gone off several times when something was wrong -- that's a miss in my mind.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.

jojoejos

Re: Smart Start/Stop

Post by jojoejos » Fri Aug 19, 2011 10:25 am

I found all the answers regarding my ResMed S9 Elite very helpful. Thank you to all who posted answers here.

BrunoKnows

Re: Smart Start/Stop

Post by BrunoKnows » Tue Dec 17, 2013 8:08 pm

Thank you sooooo much! I pressed setting and held button down for 3 sec and it unlocked the specs. I found autostart and now all good. Thanks again.