Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
-
- Posts: 6
- Joined: Thu Mar 11, 2010 4:58 pm
- Location: Ontario, Canada
Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
So I've been on CPAP therapy for about 3 years. Started out at a pressure of 10 which was then increased to 13 about a year ago. These pressures were determined by an overnight sleep study. I have other health problems as well (awaiting kidney transplant) so I really want to be getting the best sleep possible.
Had another sleep study done recently and the recommended pressure is now 17. I had already been playing around with increasing the pressure on my own as I felt I needed more pressure so had tried as high as 18 when I had a stuffed up nose and then settled on 16. I am now going to increase to 17 as per my dr.
I want to get a new machine but I'm not sure what I should be getting. I actually have 2 different sleep doctors concurrenty - a regular sleep dr and my shrink who specializes in sleep research and ordered the latest sleep study. I had a conversation with my original dr last year about getting an auto CPAP - he said there 2 schools of thought about auto machines - those that believe it is good to have variable pressure adapting to your needs and those that think you should find the minimum pressure that eliminates the apnea and stick to it. He ended up giving me a script for an auto CPAP and said it wouldn't hurt to try.
My shrink had originally mentioned the possibility of a bi-pap (before the sleep study) but now says I should hold off and try the 17 pressure for a while and see how that goes. Part of her hesitation is because we are in Canada and the machines are much more expensive, especially the bi-pap which can be around $5,000 and there are all kinds of hoops to jump through to get the gov't to pay for part of them. I'm looking at just purchasing something on CPAP.com which is some cases is literally 1/3 the cost of the same machine in Canada...it's ridiculous.
I have this HSA (Health spending account) that I can to use to get a new machine but I'm not sure what to do. I currently have a Resmed S8 and am kind of partial to Resmed for some reason. They just seem to be the "coolest" CPAP's around I really like the look of the new S9 and I would get the AUTO version because an auto CPAP seems to make sense to me in theory. I'm just not sure what to do about a bi-pap - I guess I don't really have a prescription for a bi-pap right now but if I really wanted one I'm sure I could get my shrink to write me one. I'm really confused between the s8 VPAP 25 and S8 VPAP ST which is right about double the price. What is the difference between the 2 in practical terms. I understand the ST is "spontantenous timed" but what is the benefit to that? Also, I really like the look of the S9 - are S9 bi-paps coming out soon?
Finally, I've been reading about the Resscan software and really want to start taking charge of my therapy and using that but I can't find anywhere to buy it! I've read in all the S9 threads about people getting and using the newest version 3.10- where are these people obtaining it??
Sorry for the long first post! I really love that this forum is out there. My DME is useless and just try and sell me whatever machine THEY want me to have - I really had to fight to get my S8 Elite originally and when I asked about the S8 II they had no idea what I was talking about and said it must only be available in the US - it came out a month later after I bought my original S8
Had another sleep study done recently and the recommended pressure is now 17. I had already been playing around with increasing the pressure on my own as I felt I needed more pressure so had tried as high as 18 when I had a stuffed up nose and then settled on 16. I am now going to increase to 17 as per my dr.
I want to get a new machine but I'm not sure what I should be getting. I actually have 2 different sleep doctors concurrenty - a regular sleep dr and my shrink who specializes in sleep research and ordered the latest sleep study. I had a conversation with my original dr last year about getting an auto CPAP - he said there 2 schools of thought about auto machines - those that believe it is good to have variable pressure adapting to your needs and those that think you should find the minimum pressure that eliminates the apnea and stick to it. He ended up giving me a script for an auto CPAP and said it wouldn't hurt to try.
My shrink had originally mentioned the possibility of a bi-pap (before the sleep study) but now says I should hold off and try the 17 pressure for a while and see how that goes. Part of her hesitation is because we are in Canada and the machines are much more expensive, especially the bi-pap which can be around $5,000 and there are all kinds of hoops to jump through to get the gov't to pay for part of them. I'm looking at just purchasing something on CPAP.com which is some cases is literally 1/3 the cost of the same machine in Canada...it's ridiculous.
I have this HSA (Health spending account) that I can to use to get a new machine but I'm not sure what to do. I currently have a Resmed S8 and am kind of partial to Resmed for some reason. They just seem to be the "coolest" CPAP's around I really like the look of the new S9 and I would get the AUTO version because an auto CPAP seems to make sense to me in theory. I'm just not sure what to do about a bi-pap - I guess I don't really have a prescription for a bi-pap right now but if I really wanted one I'm sure I could get my shrink to write me one. I'm really confused between the s8 VPAP 25 and S8 VPAP ST which is right about double the price. What is the difference between the 2 in practical terms. I understand the ST is "spontantenous timed" but what is the benefit to that? Also, I really like the look of the S9 - are S9 bi-paps coming out soon?
Finally, I've been reading about the Resscan software and really want to start taking charge of my therapy and using that but I can't find anywhere to buy it! I've read in all the S9 threads about people getting and using the newest version 3.10- where are these people obtaining it??
Sorry for the long first post! I really love that this forum is out there. My DME is useless and just try and sell me whatever machine THEY want me to have - I really had to fight to get my S8 Elite originally and when I asked about the S8 II they had no idea what I was talking about and said it must only be available in the US - it came out a month later after I bought my original S8
Re: Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
cpap.com can't ship Resmed to Canada
to get a VPAP you need a bilevel script
to get the ST machine you need a specific script for that
if you check craigslist they don't have script requirements but of course buyer beware and who knows about using your HSA for such a purchase
to get a VPAP you need a bilevel script
to get the ST machine you need a specific script for that
if you check craigslist they don't have script requirements but of course buyer beware and who knows about using your HSA for such a purchase
-
- Posts: 6
- Joined: Thu Mar 11, 2010 4:58 pm
- Location: Ontario, Canada
Re: Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
Yeah, I found out recently they won't ship Resmed to Canada - that just means I'll have to get it shipped to Buffalo and drive over and pick it up - not a big deal.jules wrote:cpap.com can't ship Resmed to Canada
to get a VPAP you need a bilevel script
to get the ST machine you need a specific script for that
if you check craigslist they don't have script requirements but of course buyer beware and who knows about using your HSA for such a purchase
I think I could get her to write the script for the bilevel but she didn't seem to know what the ST was??? I guess I don't really need it then?
My understanding of a bi-level is that it allows a different pressure for inhalation and exhalation - is this just so people can tolearate a higher pressure which might be more difficult for exhalation?? Is Resmed's EPR kind of a form of basic bi-level where it reduces the pressure by up to 3? If this is the case I don't seem to have an issue with the exhalation pressure even when I had my S8 set to 18 - granted I do have the EPR on at 3.
Re: Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
If the sleep specialist doesn't know what ST is, maybe you need to find a sleep specialist who does. I am not saying you need it (nor will I say I know exactly what it is either as I am not a sleep specialist), but I would expect it to be "basic knowledge" for a sleep specialist.
Re: Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
cpapguy1977 wrote:So I've been on CPAP therapy for about 3 years. Started out at a pressure of 10 which was then increased to 13 about a year ago. These pressures were determined by an overnight sleep study. I have other health problems as well (awaiting kidney transplant) so I really want to be getting the best sleep possible.
Had another sleep study done recently and the recommended pressure is now 17. I had already been playing around with increasing the pressure on my own as I felt I needed more pressure so had tried as high as 18 when I had a stuffed up nose and then settled on 16. I am now going to increase to 17 as per my dr.
DSM: cpapguy1977 - the issue of your pressure going up so fast in a 'relatively' short time (2-3 years) suggests a notable decline in health. I see the mention of kidney transplant, have you gone thru a substantial weight gain in this period ? - is there any other notable medical situation ?
I want to get a new machine but I'm not sure what I should be getting. I actually have 2 different sleep doctors concurrenty - a regular sleep dr and my shrink who specializes in sleep research and ordered the latest sleep study. I had a conversation with my original dr last year about getting an auto CPAP - he said there 2 schools of thought about auto machines - those that believe it is good to have variable pressure adapting to your needs and those that think you should find the minimum pressure that eliminates the apnea and stick to it. He ended up giving me a script for an auto CPAP and said it wouldn't hurt to try.
DSM: Your Doc is pretty well right about the 2 schools of thought. I am biased towards bilevel with epap (exhale pressure) remaining at a preset pressure & a higher ipap (inhale pressure) being used to clear hypopneas etc: BUT, I am aware that some (limited numbers) of people when put on bilevel therapy, can exhibit more centrals and also for some others, bilevel can cause some PaCO2 imbalance (carbon dioxide in the blood). So, for those of us not diagnosed as CompSA sufferers, bilevel can be a very satisfying type of cpap therapy. I would always use bilevel over autocpap.
My shrink had originally mentioned the possibility of a bi-pap (before the sleep study) but now says I should hold off and try the 17 pressure for a while and see how that goes. Part of her hesitation is because we are in Canada and the machines are much more expensive, especially the bi-pap which can be around $5,000 and there are all kinds of hoops to jump through to get the gov't to pay for part of them. I'm looking at just purchasing something on CPAP.com which is some cases is literally 1/3 the cost of the same machine in Canada...it's ridiculous.
DSM: I would be *very* surprised if the current Bipaps (standard bilevel & auto bilevel) were around those prices ($5000). Both Respironics and Resmed have dropped their pricing markedly in the past couple f years. Also cpapauction.com has a variety of these at bargain prices.
I have this HSA (Health spending account) that I can to use to get a new machine but I'm not sure what to do. I currently have a Resmed S8 and am kind of partial to Resmed for some reason. They just seem to be the "coolest" CPAP's around I really like the look of the new S9 and I would get the AUTO version because an auto CPAP seems to make sense to me in theory. I'm just not sure what to do about a bi-pap - I guess I don't really have a prescription for a bi-pap right now but if I really wanted one I'm sure I could get my shrink to write me one. I'm really confused between the s8 VPAP 25 and S8 VPAP ST which is right about double the price. What is the difference between the 2 in practical terms. I understand the ST is "spontantenous timed" but what is the benefit to that? Also, I really like the look of the S9 - are S9 bi-paps coming out soon?
DSM: If you can, go for an Auto Bilevel as you can choose to activate or deactivate the auto aspect. Unless you have central apnea you don't need the S/T models. S/T is the preferred device for dealing with people who have serious central apneas (vs not serious ones such as sleep onset centrals & casual (turning over in bed / yawning etc: centrals)).
Finally, I've been reading about the Resscan software and really want to start taking charge of my therapy and using that but I can't find anywhere to buy it! I've read in all the S9 threads about people getting and using the newest version 3.10- where are these people obtaining it??
DSM: ResScan is impressive. But I will be honest & say that after the novelty of looking into your data wears off (took a few years for that to happen to me) it doesn't really matter that much BUT had it been around 5 years ago I would have gone for it without hesitation.
Sorry for the long first post! I really love that this forum is out there. My DME is useless and just try and sell me whatever machine THEY want me to have - I really had to fight to get my S8 Elite originally and when I asked about the S8 II they had no idea what I was talking about and said it must only be available in the US - it came out a month later after I bought my original S8
DSM: Yes this forum is tops
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)
- rested gal
- Posts: 12881
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Re: Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
I would not want an S/T (spontaneous/timed) machine unless I really, really, really needed it. Just for the heck of it, I tried a borrowed bilevel S/T machine for a couple of weeks; playing around with the timed mode. Since I don't actually need a "timed" backup rate, I found any timed rate other than the very lowest BPM (breaths per minute) times to be annoying...
Since you haven't been told you need "timed", I wouldn't even consider S/T machines if I were you.
Bilevel is very comfortable for many people. I don't actually even need a bilevel machine, but I use one for the comfort when exhaling.
A bilevel auto machine is handy since it can be used in just bilevel mode or in autotitrating bilevel mode. Can also be set to work like plain CPAP by setting both the EPAP and IPAP pressures the same when bilevel mode is chosen. Very versatile.
Since you haven't been told you need "timed", I wouldn't even consider S/T machines if I were you.
Bilevel is very comfortable for many people. I don't actually even need a bilevel machine, but I use one for the comfort when exhaling.
A bilevel auto machine is handy since it can be used in just bilevel mode or in autotitrating bilevel mode. Can also be set to work like plain CPAP by setting both the EPAP and IPAP pressures the same when bilevel mode is chosen. Very versatile.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Re: Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
I have gained weight...however that is mostly from my kidneys enlarging - my condition causes my kidneys to grow to several times normal. Also, along with any chronic kidney disease comes fluid retention which also causes some weight gain. I've also had other sleeping issues requiring the use of sleeping pills at timea as well.dsm wrote: DSM: cpapguy1977 - the issue of your pressure going up so fast in a 'relatively' short time (2-3 years) suggests a notable decline in health. I see the mention of kidney transplant, have you gone thru a substantial weight gain in this period ? - is there any other notable medical situation ?
I have to admit I haven't specifically priced out VPAP's but I did compare pricing on auto CPAP's. For the S8 Autoset II I was given quotes ranging from $1535 CDN to just under $2000 - last I checked CPAPdotcom's pricing was about $500. The S8 VPAP ST is about $2500 on cpapdotcom so I wouldn't be surprised if it was $5000 in Canada. There is a program where the governments pays a large portion of CPAP machines so I can see that causing the price to be higher than it needs to be. However, to get them to pay for a Bi-pap is apparently very difficult due to their high cost.dsm wrote:
DSM: I would be *very* surprised if the current Bipaps (standard bilevel & auto bilevel) were around those prices ($5000). Both Respironics and Resmed have dropped their pricing markedly in the past couple f years. Also cpapauctiondotcom has a variety of these at bargain prices.
From your posts (and some others) it sounds like a bi-pap might be a good idea and the VPAP Auto 25 is about $1375 on cpapdotcom which I could cover out of my HSA so I might have to look into that. I just wish there was a new S9 version!!
Re: Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
cpapguy1977,
That kidney issue is obviously a big worry. I wish you all the best in getting that under control quickly.
I can easily see that being a major contributor to fluid retention & weight gain which will also be
affecting your xPAP therapy.
The Vpap 25 is an excellent choice & yes I guess an S9 version would be nice for the data analysis.
Don't ignore the Respironics Bipap models as the older tank & PR1 should also be excellent choices should
you locate one at a good price. Re Resmed S8 shape, it is a great travel machine if you do that.
Cheers
DSM
That kidney issue is obviously a big worry. I wish you all the best in getting that under control quickly.
I can easily see that being a major contributor to fluid retention & weight gain which will also be
affecting your xPAP therapy.
The Vpap 25 is an excellent choice & yes I guess an S9 version would be nice for the data analysis.
Don't ignore the Respironics Bipap models as the older tank & PR1 should also be excellent choices should
you locate one at a good price. Re Resmed S8 shape, it is a great travel machine if you do that.
Cheers
DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)
-
- Posts: 6
- Joined: Thu Mar 11, 2010 4:58 pm
- Location: Ontario, Canada
Re: Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
Thanks rested gal. Both yourself and DSM make good cases for the bi-level in which case I would be looking at the VPAP Auto 25. However, Resmed must have been partially correct in their research about people wanting the S9 for purely aesthetic reasons because I really like the look of it! As I mentioned to DSM in a PM I think I might pick up the S9 Auto for now and if my OSA gets any worse requiring pressures above 20 then I can pick up a bi-level next year.rested gal wrote:I would not want an S/T (spontaneous/timed) machine unless I really, really, really needed it. Just for the heck of it, I tried a borrowed bilevel S/T machine for a couple of weeks; playing around with the timed mode. Since I don't actually need a "timed" backup rate, I found any timed rate other than the very lowest BPM (breaths per minute) times to be annoying...
Since you haven't been told you need "timed", I wouldn't even consider S/T machines if I were you.
Bilevel is very comfortable for many people. I don't actually even need a bilevel machine, but I use one for the comfort when exhaling.
A bilevel auto machine is handy since it can be used in just bilevel mode or in autotitrating bilevel mode. Can also be set to work like plain CPAP by setting both the EPAP and IPAP pressures the same when bilevel mode is chosen. Very versatile.
Re: Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
The Respironic Bipap "M" models are readily available used in the US for under $500. Auto or Pro. $375 is ballpark.
The machines can vary widely in hours of use, starting with almost none. Just ask about blower hours (you may have to be ready to tell people how to find that, via the online manuals. )
The machines can vary widely in hours of use, starting with almost none. Just ask about blower hours (you may have to be ready to tell people how to find that, via the online manuals. )
VPAP ASV: BiPaP ASV: Quattro FF: Activa LT: Swift FX
- timbalionguy
- Posts: 888
- Joined: Mon Apr 27, 2009 8:31 pm
- Location: Reno, NV
Re: Confused...S9 Auto vs S8 VPAP Auto 25 vs S8 VPAP ST
The VPAP Auto 25 can also be used as an autoadjusting CPAP, if you are not ready for bileevel therapy. I sure like mine.
Lions can and do snore....