Hello,
I was going through the various settings on my machine the other day and noticed the mask setting was wrong. It was set to "swift" but I use an "activa LT" . So I made the correction. But I am wondering does this have any effect on the way the CPAP machine delivers? It is set to 14cm H2O. Is this pressure not constent regardless of the mask? And if so then what's the purpose of this setting?
Thanks
Luc
Question about mask setting on unit
Re: Question about mask setting on unit
You're right, Luc, the pressure is constant regardless of mask setting.Luc wrote:Hello,
I was going through the various settings on my machine the other day and noticed the mask setting was wrong. It was set to "swift" but I use an "activa LT" . So I made the correction. But I am wondering does this have any effect on the way the CPAP machine delivers? It is set to 14cm H2O. Is this pressure not constent regardless of the mask? And if so then what's the purpose of this setting?
Thanks
Luc
Every mask has an expected leak rate -- Resmed uses the mask info you give it to tell you how good (or how bad) your leak is -- with wrong mask info you'll have wrong leak info as well.
O.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
- BleepingBeauty
- Posts: 2454
- Joined: Thu Apr 02, 2009 5:30 pm
- Location: Aridzona ;-)
Re: Question about mask setting on unit
Hi, Luc, and welcome to the forum.Luc wrote:Hello,
I was going through the various settings on my machine the other day and noticed the mask setting was wrong. It was set to "swift" but I use an "activa LT" . So I made the correction. But I am wondering does this have any effect on the way the CPAP machine delivers? It is set to 14cm H2O. Is this pressure not constent regardless of the mask? And if so then what's the purpose of this setting?
Thanks
Luc
The mask setting has no bearing on how much pressure your machine is putting out. The purpose of having the correct mask setting in the machine relates to the reported leak rate you experience. Your machine factors in what the intentional flow (or leak rate) is for the mask it's set to, at the pressure setting you're using. If the mask setting is correct on the machine, you can rely on the reported leak rate; if not, the data is skewed.
Any leak reported on a ResMed machine (provided the mask setting is correct) is over and above what it should be. The goal is to have zero leaks.
Hope that helps.
Veni, vidi, Velcro. I came, I saw, I stuck around.
Dx 11/07: AHI 107, central apnea, Cheyne Stokes respiration, moderate-severe O2 desats. (Simple OSA would be too easy.
)
PR S1 ASV 950, DreamWear mask, F&P 150 humidifier, O2 @ 2L.
Dx 11/07: AHI 107, central apnea, Cheyne Stokes respiration, moderate-severe O2 desats. (Simple OSA would be too easy.
PR S1 ASV 950, DreamWear mask, F&P 150 humidifier, O2 @ 2L.
Re: Question about mask setting on unit
I had that problem as well. My machine was set for an ultra but i have an activa. I was getting horrible leak info back. I am curious now that i changed it to activa and have also bought my chin strap gear how well this will go over the next few days.
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Additional Comments: Pressure Setting: 13 |
"Things turn out the best for the people who make the best of the way things turn out"
-- JOHN WOODEN --