Question about Split Night, 140, 180, 220

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
josef
Posts: 63
Joined: Sun Dec 13, 2009 1:03 pm
Location: South Texas

Question about Split Night, 140, 180, 220

Post by josef » Sat Feb 06, 2010 8:08 am

I have recently bought a Respironics Remstar auto m series a flex on CPAP Auction with no problems. In the setup menu, one of the questions is, Split Night with the 3 options of 140, 180, 220. I know the answer is so simple that they didn't even bother to put it in the 55 page Clinician's Manual. If they did, they did a pretty good job of hiding it. Thanks.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: I recently bought a CMS-50E OLED Fingertip Pulse Oximeter. It does miniture sleep studies.
Josef

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Question about Split Night, 140, 180, 220

Post by Wulfman » Sat Feb 06, 2010 8:32 am

josef wrote:I have recently bought a Respironics Remstar auto m series a flex on CPAP Auction with no problems. In the setup menu, one of the questions is, Split Night with the 3 options of 140, 180, 220. I know the answer is so simple that they didn't even bother to put it in the 55 page Clinician's Manual. If they did, they did a pretty good job of hiding it. Thanks.
There are actually FOUR options:
-Split Night Time = Off (off, 120, 180, 240)

You want to choose "Off".
The other numbers represent minutes of time (2, 3 and 4 hrs.) that the machine will stay at a certain pressure setting before switching to another setting or range. It's for diagnostic purposes.


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
josef
Posts: 63
Joined: Sun Dec 13, 2009 1:03 pm
Location: South Texas

Re: Question about Split Night, 140, 180, 220

Post by josef » Sat Feb 06, 2010 8:56 am

Thanks very much, I deffinitely would not have guesed that one.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: I recently bought a CMS-50E OLED Fingertip Pulse Oximeter. It does miniture sleep studies.
Josef

User avatar
rested gal
Posts: 12881
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Re: Question about Split Night, 140, 180, 220

Post by rested gal » Sat Feb 06, 2010 9:14 am

I agree with Den. You want to keep the "Split Night" setting "off."
As Den said, "Split Night" is a diagnostic tool. It's not a treatment setting.

Here's an example of what would happen if you chose a Split Night time:

Say your M auto is set to operate in auto mode and set for a pressure range of 5 - 15 cm H2O.

If you chose the Split Night time of 180 (3 hours), and started your machine that night, the machine would use only the minimum pressure (5 cm) set for your range, for the first three hours of the night.

It would record the usual data but would not change the pressure at all even if you were being slammed with apneas, hypopneas, snoring, etc. It would continue to use your minimum pressure of 5 cm for the entire first three hours and simply gather data about the airflow.

When three hours was finished, the machine would automatically begin behaving as an autopap does...responding to flow limitations, snores, etc., and varying the pressure as needed within your range of 5 - 15. And continue to record data, as usual.

Each night, Split Night would continue to do its thing until you go back into the therapy setup menu and turn it off.

For regular treatment purposes, the Split Night setting should always be "off."