1st night on CPAP - disappointing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Vallan
Posts: 12
Joined: Wed Nov 25, 2009 11:30 am

1st night on CPAP - disappointing

Post by Vallan » Wed Dec 16, 2009 12:30 pm

Only 3 trips to pick up a CPAP machine and I finally got the S8 Autoset II w/ H4i humidifer I wanted. I was given the Mirage Softgel mask. Pressure set 9 - 15, no ramp, EPR on 3.

I was slightly congested, but the congestion felt worse wearing the mask. My biggest problem was exhaling through my nose. It felt like there was too much resistance to exhaling. I took my mask off in the middle of the night and my ability to breath through my nose got better. Put the mask back on 15 mins latter and back to having some problems. My machine info this morning said:
1 day 2.53 hrs
Leak .04L/s
AHI 11.7
AI 4.5
HI 7.2
Pressure 12.4

The sleep clinic study seemed easier to exhale. I realize 1 night is not much of a test, but I'm wondering how to fix the problem so I can wear the mask. I have surgery in 5 days where I know I'll get some hefty pain meds. I was hoping to get my cpap issues resolved by then.

User avatar
ozij
Posts: 10444
Joined: Fri Mar 18, 2005 11:52 pm

Re: 1st night on CPAP - disappointing

Post by ozij » Wed Dec 16, 2009 12:33 pm

Do you have a heated humidifier? Try changing it's temprature, some of us get more congested by humidity -- others need more humidity. It's trial and error.

O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

DreamOn
Posts: 1920
Joined: Fri Oct 23, 2009 10:13 am

Re: 1st night on CPAP - disappointing

Post by DreamOn » Wed Dec 16, 2009 1:33 pm

Fantastic that you were able to get the machine you wanted!!!

In addition to adjusting humidifier settings, you may want to consider using the ramp feature for a while until you become more accustomed to the machine. You could set the starting ramp pressure at 6 or so, for maybe 30 minutes. I wouldn't start the ramp pressure much lower than 6, as it can be more difficult to breathe when the pressure's too low. With the ramp feature, it will gradually increase pressure and won't get to full pressures until you've fallen asleep (hopefully), so you won't notice it so much. Over time, you'll probably want to turn the ramp off again.

It does take some time to get used to the machine. We often concentrate on our breathing too much in the beginning. Over time, you'll find what works best for you.

~ DreamOn

User avatar
BrianR4743
Posts: 83
Joined: Sun Nov 22, 2009 12:01 pm
Location: Long Beach, CA

Re: 1st night on CPAP - disappointing

Post by BrianR4743 » Wed Dec 16, 2009 1:38 pm

Hang tough there Vallan, and spend a whole lot of time drawing on the resources available here. I've been doing the CPAP thing for a month now, and have had some good days, and some days where I just felt like I tanked. Everything you're doing now is for the right reason, getting healthy, and your first night isn't too shabby. You'll most likely see and feel incremental improvements, but it'll take a bit of time.

For me, the most important part was just getting comfortable with all of the equipment and developing a nightly routine to help relax. As a guy who first hated this contraption, I now look forward to hitting the hay knowing that I'm doing something that's going to make my quality of life better, and most likely, add some years to it. Is it still uncomfortable at times? Yep. Still frustrating? Yep. Tired sometimes? Yep. Making improvements...... absolutely, and you will too.

Good luck to you! You've got tons of friends here!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Aussie Heated Hose, Pressure of 10
“Sleeplessness is a desert without vegetation or inhabitants”

User avatar
xiola
Posts: 101
Joined: Tue Oct 20, 2009 11:50 am
Location: St. Louis, Missouri

Re: 1st night on CPAP - disappointing

Post by xiola » Wed Dec 16, 2009 1:40 pm

Vallan wrote: My biggest problem was exhaling through my nose. It felt like there was too much resistance to exhaling. I took my mask off in the middle of the night and my ability to breath through my nose got better.
Hi. I'm pretty new to this too, been using it just over a month. I felt exactly the same way, in fact, one of my first posts here was me panicking because I couldn't breathe with the machine on. I felt like i just couldn't exhale.

Relax. read a book, watch tv. it DOES get better. After a day or so you won't even notice it anymore once you get used to the newness of how it feels, and adjust to breathing with it.

User avatar
bailachel
Posts: 387
Joined: Sat Oct 24, 2009 11:40 am
Location: Florida

Re: 1st night on CPAP - disappointing

Post by bailachel » Wed Dec 16, 2009 1:46 pm

Welcome! I'm another newbie who has only been using my machine for about a month. Hang in there! It will be worth it. It will get easier as you go along. I have gotten so used to mine that I woke up the other night and at first I thought that the machine was off because I couldn't feel the air until I put my hand over the vent.

This board is a great resource. In my experience, one can find answers to any question here and the support is terrific.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Alternate masks: Mirage Quattro, EasyLife Nasal mask

Guest

Re: 1st night on CPAP - disappointing

Post by Guest » Wed Dec 16, 2009 2:22 pm

Another newbie here to support and congratulate you on your success getting arguable “the” best machine, a stumbling block for most novices. Here to say if acclimation is difficult I might suggest trying straight CPAP and lowering the pressure with no or low EPR until you’re comfortable. EPR, ramp, etc. can just unnecessarily complicate matters at first. While a lower pressure will not provide for proper therapy, once you become acclimated to xPAP, you will then be able to increase your pressure and maximize therapy Efficacy
After all, if you don’t acclimate, you won’t use it, and no therapy puts you in square one again.


User avatar
FinceMgr
Posts: 58
Joined: Thu Oct 15, 2009 2:18 pm
Location: The High Desert, CA

Re: 1st night on CPAP - disappointing

Post by FinceMgr » Wed Dec 16, 2009 2:25 pm

Guest wrote:Another newbie here to support and congratulate you on your success getting arguably “the” best machine, a stumbling block for most novices. Here to say if acclimation is difficult I might suggest trying straight CPAP and lowering the pressure with no or low EPR until you’re comfortable. EPR, ramp, etc. can just unnecessarily complicate matters at first. While a lower pressure will not provide for proper therapy, once you become acclimated to xPAP, you will then be able to increase your pressure and maximize therapy Efficacy
After all, if you don’t acclimate, you won’t use it, and no therapy puts you in square one again.
Was logged in as guest, thats me. Oooops
Last edited by FinceMgr on Wed Dec 16, 2009 5:55 pm, edited 1 time in total.

Vallan
Posts: 12
Joined: Wed Nov 25, 2009 11:30 am

Re: 1st night on CPAP - disappointing

Post by Vallan » Wed Dec 16, 2009 2:38 pm

Thanks for the many replies. I think I will try lowering the pressure to 9 (lowest pressure where I didn't snore in sleep study) and try it out on cpap mode and reduce the humidification a bit. Thank goodness for this forum and the wealth of info and folks. Doing this alone with no experience would be so much harder!

User avatar
Jaylee
Posts: 1080
Joined: Mon Nov 23, 2009 8:43 pm

Re: 1st night on CPAP - disappointing

Post by Jaylee » Wed Dec 16, 2009 6:07 pm

I agree that when my pressure was too low it was harder to breathe.

I started using a neti pot to help relieve congestion before putting my mask on for the night. Could you try that? Also toy with the humidfier settings because you might need more or less maybe. It could just take time to adjust. The first few nights I used my ramp feature and I was panicing when I put the mask on. Now it is no biggie for me and I don't use the ramp option. I think it might just take an adjustment period for some of us.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Pressure 18/13.

User avatar
SuperGeeky
Posts: 414
Joined: Mon May 19, 2008 3:55 pm
Location: MaryEtta, Gawga (Now a part of the Union)

Re: 1st night on CPAP - disappointing

Post by SuperGeeky » Wed Dec 16, 2009 6:59 pm

Vallan, pressured air sent into our sinus cavities, throat and lungs is not all that natural. I would suppose for most all of us, irritates sinuses with the occasional infection. Certainly does for me...

Neilsmed sells netipots and a squeeze bottle which works pretty well. Walmart, CVS etc..

Struggling with exhaling sounds like something to mention to the Doc, ask about a BiPap. I was fortunate to ask for a Bipap on my first titration. I liked it and the Doctor prescribed it after trying CPAP for a month or so...

Good luck and sleep well,

SG

_________________
Humidifier
Additional Comments: Cozy Hose, Hose Boss Management System, PapCap F Series

User avatar
sleepywolverine
Posts: 46
Joined: Sun Nov 15, 2009 11:18 am

Re: 1st night on CPAP - disappointing

Post by sleepywolverine » Wed Dec 16, 2009 7:55 pm

Hi I'm a newbie too so I just wanted to say welcome!! It does get better. I'm on a bipap, so it's a little easier, but I think you'll get used to it. My hubby put my mask against his face one day to feel the pressure and felt like he couldn't breathe too, but once you've had it on a few nights you won't even really notice anymore.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Machine is ResMed Vpap S, couldn't find it in options

User avatar
kteague
Posts: 7782
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: 1st night on CPAP - disappointing

Post by kteague » Thu Dec 17, 2009 1:38 am

If this congestion is only when you wear the mask, I agree, humidity could be an issue. Also, one mask I had sat on the bridge of my nose is such a way as to block my sinuses on both sides. Only got congested with that one mask. it also caused a pressure sore. One more thing to consider, soaking the mask to allow any offensive chemical or plastic smells to dissipate can help those prone to sensitivities.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

User avatar
BrianR4743
Posts: 83
Joined: Sun Nov 22, 2009 12:01 pm
Location: Long Beach, CA

Re: 1st night on CPAP - disappointing

Post by BrianR4743 » Thu Dec 17, 2009 2:15 pm

Good afternoon Vallan!

How did your 2nd night go? A bit better than night #1?

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Aussie Heated Hose, Pressure of 10
“Sleeplessness is a desert without vegetation or inhabitants”

Vallan
Posts: 12
Joined: Wed Nov 25, 2009 11:30 am

Re: 1st night on CPAP - disappointing

Post by Vallan » Thu Dec 17, 2009 5:44 pm

2nd night was much better. I'd forgot to mention I did the neti pot 90 mins before bed on night 1, but didnt help.

Last night I took 1 benedryl to help with allergies and minor help to sleep, and I used afrin an hour before bed. I know many folks are anti-afrin, but for occasional use (I never use more than 1/day, never more than 2 days in a row. 1st time I've used it in 10 years) it makes a big difference. I'll go without it tonight. Also loosened my straps to where the mask almost leaks and that seemed to help.

Day 1 was Day 2 was
Usade 2:53 hrs 6:27 hrs
Leak .04L/s 0.00L/s
AHI 11.7 11.9
AI 4.5 4.6
HI 7.2 7.3
Pressure 12.4 11.8

Numbers don't look so good. Is my AHI total or per hour? I'm still not exactly sure what my numbers are telling me other than I think we're supposed to keep AHI <5 is possible. My machine is set 9 - 15 with a setpoint on a fixed machine a 12. Should I bump up my min pressure to 10, 10.5 or ?