BIPAP AUTO-SV SETTINGS HELP

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: BIPAP AUTO-SV SETTINGS HELP

Post by -SWS » Fri Dec 11, 2009 9:07 am

Muffy wrote:I disagree with your disagreement. The CompSAS cycle is perpetuated by overshoot, so the key is not how it blows through the CA, but how pressure support drops back during the hyperventilatory period immediately after. That is the critical area. If this were not the case, then simple BiPAP with a backup rate would be successful in treating CompSAS and there would be no need for ASV.

Further, singular (or more specifically, isolated) central apneas are nearly always benign, and should be left alone.

Muffy
You know how I feel about these disagreements. Respironics may have to rework their wording since they clearly state their BiPAP autoSV will automatically trigger a breath for a central apneic event (singular). Respironics never mentions anything about that machine's backup rate being reserved for multiples, cycles, or sequences of central apneas. They address periodicity patterns with fluctuating PS. But they never explicitly or implicitly claim to refrain from backing up singular central apneas with their backup rate.

They even use a singular central-apnea case in their description of administering backup rate:
the BiPAP autoSV algorithm also calculates the patient's spontaneous breathing rate and will automatically trigger a breath should the patient have a central apneic event.
http://bipapautosv.respironics.com/how.aspx

So if the BiPAP autoSV refrains from administering its backup rate to singular central apnea events, I can't find that anywhere in literature. But I can find statements like the above describing "a central apneic event" (singular) receiving that algorithm's backup rate.

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: BIPAP AUTO-SV SETTINGS HELP

Post by -SWS » Fri Dec 11, 2009 9:14 am

Sorry about the side discussion, CROWPAT... I had better place your sleep hygiene answers on this page to make them easier for Muffy to work with:
CROWPAT wrote:1. You are truly a "long sleeper" and that's just how it's going to be (and if you were only sleeping 5.5 hours back when you were working, then that must have been hell)(Hey! Hell On Wheels! Wasn't that the nickname of an armored division?) 1st or 2d AD, I forget.
2. You are "oversleeping" and it's too much. Statistically, long sleepers tend to have the same morbidity/mortality as short-sleepers;
3. You have a sleep maintenance issue, and that 9 hours of DL time is really like 7 hours of sleep;
4. You have another medical issue(s) (like uncontrolled depression).

Can you get a clear copy of the graphs from 2006? I will try to get one.

Time for a sleep log:

http://www.sleepeducation.com/pdf/sleepdiary.pdf I downloaded it.

Do you take naps? Never.

See the receptionist for another appointment in 2 weeks.

Muffy

Got up, went out in 24 degrees, and walked for 30 minutes. Fog returned shortly after I got home, but no worse than usual.

User avatar
Muffy
Posts: 960
Joined: Tue Apr 21, 2009 5:28 am
Location: Schenectady, New York

Re: BIPAP AUTO-SV SETTINGS HELP

Post by Muffy » Fri Dec 11, 2009 10:53 am

-SWS wrote:So if the BiPAP autoSV refrains from administering its backup rate to singular central apnea events, I can't find that anywhere in literature. But I can find statements like the above describing "a central apneic event" (singular) receiving that algorithm's backup rate.
I didn't say it didn't, I said it shouldn't. My point is that not every central apnea needs to be bludgeoned to death.

Muffy
________________________________

Machine: Dell Dimension 8100
Mask: 3M N-95 (during flu season)
Humidifier: Avoided, tends to make me moldy
Software: XP Pro
Additional Comments: You can't find a solution when you don't know the problem

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: BIPAP AUTO-SV SETTINGS HELP

Post by -SWS » Fri Dec 11, 2009 11:44 am

Got it... and I agree that a rare or isolated (brief) central apnea is perfectly benign. I'm not sure... But here's where our paradigms might turn out to disagree a bit regarding key objectives for this ASV machine's fluctuating IPAP peak:
The CompSAS cycle is perpetuated by overshoot, so the key is not how it blows through the CA, but how pressure support drops back during the hyperventilatory period immediately after. That is the critical area. If this were not the case, then simple BiPAP with a backup rate would be successful in treating CompSAS and there would be no need for ASV.
Once in that cycle, the recurrent moments of biological undershoot are genuinely problematic as well.

This Respironics algorithm fluctuates IPAP peak between the bounds of IPAP min and IPAP max for both halves of an oscilation-based respiratory controller problem: progressive biological overshoot cyclically interspersed with progressive biological undershoot. Accordingly, the algorithm drops toward IPAP min during those moments of biological respiratory overshoot. I agree that is in part key. However, strategically it is only half-key. And IPAP peak cannot actively or instantaneously minimize or subtract that biologically-produced volume over-flow simply by dropping IPAP peak. Rather, IPAP peak simply avoids exacerbating that biological over-ventilation by stepping out of the way or minimizing mechanical ventilation.

By contrast, the other fluctuating IPAP "half-key" (when IPAP peak proportionally moves toward the IPAP max limit) actively or additively assists ventilation during those moments of biological undershoot. Additional timely and proportional flow is generated by the machine to assist with biologically generated flow being under-target. So I view the IPAP peak upswings as assisting with proportional machine-generated flow, but the IPAP peak downswings as essentially stepping down to avoid mechanically exacerbating those moments of biological overshoot.

ASV's Proportional Assist Ventilation is an inherently bipolar proposition: active ventilatory assistance for increasingly deficient biological flow during IPAP peak's progressive upswings; but avoidance of biological exacerbation as IPAP peak proportionately stands down with its pressure support during progressively developing cycles of biological overshoot.

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: BIPAP AUTO-SV SETTINGS HELP

Post by dsm » Fri Dec 11, 2009 7:40 pm

Muffy wrote:
<snip

ASV technology is designed to fix a cyclical central pattern. It does not "fix" any singular central apnea per se.

<snip>
Muffy
Muffy

Excellent info on clarifying central types & their relevance - very easy to forget that aspect. But re the above point, wouldn`t the timed mode feature of an ASV address any isolated central s ? was also thinking that the flow targeting algorithm would try to smooth out a sudden dip below current flow. Both being to switch to ipap & boost ipap (reason I ask is I measured av 3 CMs boosts on my machine repeated for 3 breaths if simulating a central).

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: BIPAP AUTO-SV SETTINGS HELP

Post by -SWS » Fri Dec 11, 2009 8:28 pm

I believe Muffy's clarified point above was that the ASV machine's design intent was never to fix a few isolated benign central apneas. The original comment really had to do more with matching the right patient to the right machine. Muffy succinctly clarified later on that the comment was not to say that the ASV machine is algorithmically incapable of responding to individual central apneas.

The machine's target application:
Respironics FAQ wrote:What is the BiPAP autoSV device used for?

The intended use is: To provide non-invasive ventilatory support to treat adult patients with OSA and Respiratory Insufficiency caused by central and/or mixed apneas and periodic breathing.
http://bipapautosv.respironics.com/faq.aspx

Periodicity or cyclic central apneas are not a target requirement above---since there are cases when non-cyclic central apneas need to be addressed. The machine's backup rate will thus, by design, commence timely ventilation for central apneas either with or without cyclic characteristics.

But I agree with Muffy: a few isolated and benign central apneas are not at all what the ASV was designed to "fix".

User avatar
timbalionguy
Posts: 888
Joined: Mon Apr 27, 2009 8:31 pm
Location: Reno, NV

Re: BIPAP AUTO-SV SETTINGS HELP

Post by timbalionguy » Fri Dec 11, 2009 9:04 pm

So this begs a slightly off-topic but pertinent-to-the-discussion question: If one finds that they will have to use a servo ventilator, how does one go about making the right decision the first time? Or is this something that your clinician pretty much has to decide? These are really expensive machines, and very few can afford one of each.
Lions can and do snore....

Guest

Re: BIPAP AUTO-SV SETTINGS HELP

Post by Guest » Fri Dec 11, 2009 10:22 pm

timbalionguy wrote:So this begs a slightly off-topic but pertinent-to-the-discussion question: If one finds that they will have to use a servo ventilator, how does one go about making the right decision the first time? Or is this something that your clinician pretty much has to decide? These are really expensive machines, and very few can afford one of each.
The right decision the first time would default to the VPAP Adapt SV. As seen in this thread where the BiPAP Auto SV has run into a brick wall.

Banned

User avatar
Muffy
Posts: 960
Joined: Tue Apr 21, 2009 5:28 am
Location: Schenectady, New York

Re: BIPAP AUTO-SV SETTINGS HELP

Post by Muffy » Sat Dec 12, 2009 4:14 am

Muffy wrote:Can you get a clear copy of the graphs from 2006?
CROWPAT wrote:I will try to get one.
I meant "readable".

However, for a good chunk of the night, sleep maintenance issues are again suggested:

Image

but try to get a "clean" copy of that.

Here's a challenge for -SWS. In that pile of data you have accumulated, find ANYTHING that says a pressure < 10cmH2O is DEFINITELY ineffective.
CROWPAT wrote:Got up, went out in 24 degrees, and walked for 30 minutes. Fog returned shortly after I got home, but no worse than usual.
Great. Repeat that 300 times.

While exercise has a number of great benefits, it has little value in weight control. You're probably going to have to cut your food intake by half (your requirement would be Daily Caloric Expenditure minus about 500 Calories to lose about a pound a week, which would be an excellent goal. Have 4 equally-divided meals).

There used to be a poster who blamed all the problems of the universe on GERD. A sleep architecture such as you've shown here deserves a look at that.

Muffy
________________________________

Machine: Dell Dimension 8100
Mask: 3M N-95 (during flu season)
Humidifier: Avoided, tends to make me moldy
Software: XP Pro
Additional Comments: You can't find a solution when you don't know the problem

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: BIPAP AUTO-SV SETTINGS HELP

Post by dsm » Sat Dec 12, 2009 6:52 am

Guest wrote:
timbalionguy wrote:So this begs a slightly off-topic but pertinent-to-the-discussion question: If one finds that they will have to use a servo ventilator, how does one go about making the right decision the first time? Or is this something that your clinician pretty much has to decide? These are really expensive machines, and very few can afford one of each.
The right decision the first time would default to the VPAP Adapt SV. As seen in this thread where the BiPAP Auto SV has run into a brick wall.

Banned
Banned so often I read one of your comments & am reminded of the boy who (Emperor`s new clothes) points us to some basic home truths.

Read thee as thee will

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
Muffy
Posts: 960
Joined: Tue Apr 21, 2009 5:28 am
Location: Schenectady, New York

Nyuk, Nyuk, Nyuk!

Post by Muffy » Sat Dec 12, 2009 7:28 am

dsm wrote:
Guest wrote:
timbalionguy wrote:So this begs a slightly off-topic but pertinent-to-the-discussion question: If one finds that they will have to use a servo ventilator, how does one go about making the right decision the first time? Or is this something that your clinician pretty much has to decide? These are really expensive machines, and very few can afford one of each.
The right decision the first time would default to the VPAP Adapt SV. As seen in this thread where the BiPAP Auto SV has run into a brick wall.

Banned
Banned so often I read one of your comments & am reminded of the boy who (Emperor`s new clothes) points us to some basic home truths.

Read thee as thee will

DSM
Image

Muffy
________________________________

Machine: Dell Dimension 8100
Mask: 3M N-95 (during flu season)
Humidifier: Avoided, tends to make me moldy
Software: XP Pro
Additional Comments: You can't find a solution when you don't know the problem

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: BIPAP AUTO-SV SETTINGS HELP

Post by -SWS » Sat Dec 12, 2009 8:30 am

Muffy wrote:Here's a challenge for -SWS. In that pile of data you have accumulated, find ANYTHING that says a pressure < 10cmH2O is DEFINITELY ineffective.
You mean that well-known statistical 10cmH2O threshold in epidemiology? That very same 10cmH2O threshold beyond which significant numbers of SDB patients begin to manifest central problems?

That same well-known 10cmH2O statistical threshold beyond which unnecessarily pressure-aggressive titrations can actually create more SDB problems than solve?

CROWPAT, I bet I threw that word "statistical" in there just enough times to pique your interest in Muffy's possibility of a pressure-aggressive titration being a plausible explanation for some of your sleep issues. Muffy, I'll go back and see if I can find a shred of evidence contrary to that hypothesis.



Banned, dsm-- we really have to get some of these made up for the Resmed and Respironics buffs:

Image
Image

CROWPAT
Posts: 115
Joined: Sat Apr 15, 2006 7:32 pm
Contact:

Re: BIPAP AUTO-SV SETTINGS HELP

Post by CROWPAT » Sat Dec 12, 2009 8:45 am

No evidence at all of GERD being a problem for me.
Have never used CPAP/APAP/SV at pressures as low as 10.
Side note that I have not mentioned previously: I only eat dinner normally. Probably no more than 10 lunches a year and fewer than that number of breakfasts per year. Got into this routine from lots of field time while on active duty without time to eat the other meals. On days when I have breakfast or lunch I do not notice any change in physical feeling of tiredness.
Pat

User avatar
Banned
Posts: 602
Joined: Sun Feb 17, 2008 10:04 pm

Re: BIPAP AUTO-SV SETTINGS HELP

Post by Banned » Sat Dec 12, 2009 10:56 am

Muffy wrote: Image
Trust Us. LOL!!

Banned
AVAPS: PC AVAPS, EPAP 15, IPAP Min 19, IPAP Max 25, Vt 520ml, BPM 10, Ti 1.8sec, RT 2 (Garage)
BiPAP Auto SV: EPAP 9, IPAP Min 14, IPAP Max 25, BPM 10, Ti 2sec, RT 2 (Travel Machine)
VPAP Adapt SV: EEP 10.4, Min PS 4.4 (Every Day)
Mask: Quattro

User avatar
Patrick A
Posts: 1251
Joined: Sun Sep 18, 2005 1:00 am
Location: Yuma, Arizona
Contact:

Re: BIPAP AUTO-SV SETTINGS HELP

Post by Patrick A » Sat Dec 12, 2009 11:23 am

Banned wrote:
Muffy wrote: Image
Trust Us. LOL!!

Banned
Hey those guys look familiar

_________________
Machine
Don't Bend or Squash, My Aluminum Hat,it keeps them from knowing what I am thinking!
I need more Coffee&Old Bushmills!
"Without Truckdrivers America Stops!"
I'm not always wrong,but I'm not always right!
"Semper Fi"