Post by dsm » Sun Dec 06, 2009 8:58 pm

CROWPAT wrote:You two are WAY above my understanding of CSDB, but I appreciate all of your comments and information.
Do either of you have experience with the oximeters at the links above? My previous experience with oximeters was not good - very uncomfortable to have them on my finger. Newer versions use a strip taped to the finger that is much less confining and far more comfortable.
My conclusions from above: Lower pressure that yields more OA and little if any CA may be good and result in better "feeling". It sounds logical and I am more than willing to try anything to get there. I sincerely doubt that supplemental oxygen is necessary in my case as O readings have been good whenever I used an oximater day or night. I am willing to pay for one to help diagnose and remedy my situation, but don't want to buy something that is so inherently uncomfortable that it will further disrupt my sleep. Comments?
I read SWS as suggesting that by going back to low pressure for gathering the data & knowing it may raise the AI score is a worthwhile test - this being based on info from a report by Gilmartin Daly & Thomas: "One approach is 'permissive flow limitation' - allowing some obstruction to persist and thus avoiding the worsening of control dysfunction. ".

The theory as I understand it is that by tolerating a level of 'permissive flow limitation' (let a little bit of obstructive flow take place) the result may well be better daytime well-being and less liklihood of 'worsening control dysfunction'. The worsening control dysfunction includes (as I am reading it) Centrals and erratic/periodic breathing (which in reality is evidence of worrisome respiratory control loss). This is when we see 'control dysfunction' in the form of increased PB + HI + AI scores and an increase in machine triggered breaths vs patient triggered breaths.

Pat, your own feedback highlights that when the machine shows high levels of machine triggered breathing cycles, your daytimes suffer. When you see high levels of patient triggered breaths, you feel better 'to a point'. My reading of your situation thus far is that as pressure is raised (even if just Ipap), you begin to exhibit a respiratory control dysfunction which includes Centrals but goes further and become erratic/periodic breathing. When pressure is lowered to far you start to lose daytime well being. Straight cpap doesn't appear to be adequate & bilevel needs careful tuning OR being implements as Servo Ventilation (where the machine only applies extra Ipap pressure if it sees you won't meet the target (on this model machine) Av peak flow.

I have noticed more & more doctors providing patients with Bipap AutoSV machines set with Epap=IpapMIN (CPAP MODE) & with SV support active(typically with IpapMAX set at 30). I am not completely sure what the logic is behind this but I can't recall seeing too many users who have had their machine set any other way of late. What I think that type of setting means is that the user is essentially on CPAP until their target Av Peak Flow is not being met in the observed 4-min window on the Bipap AutoSV (vs Peak Volume on the Vpap Adapt SV in a 3-min window). I suspect this is because the doctors are viewing a continuous bilevel mode as more likely to cause problems prior to being needed. In Cpap with SV mode active, the extra Ipap pressure is applied very quickly in up to 3 CMs bursts over a range of breaths (IIRC max is 3) . The logic being that if bilevel is applied all the time (every breath) then it may trigger problems such as centrals or periodic patterns of breathing, but if the machine is set to run in CPAP + SV mode and we know a problem is about to occur in a particular breath (target flow is not going to be met) then applying the extra pressure quickly can't make respiration any worse than it is about to get.

By testing you in cpap mode at various pressures, we are building up a set of baseline effects & have the data on RR, AHI & Flow etc:

Then SWS is expecting that from this data + your feedback, the current machine can be optimized to its best settings allowing for the balance required (perhaps by looking for 'permissive flow limitation' vs excessive & debilitating respiratory control dysfunction).

Pls keep asking questions as I am by trawling through what we understand, we can help each other better understand how some of these machines can be best set up & tuned allowing for what respiratory control complications we think are emerging.

DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

CROWPAT
Posts: 115
Joined: Sat Apr 15, 2006 7:32 pm
Contact:

Re: BIPAP AUTO-SV SETTINGS HELP

Post by CROWPAT » Sun Dec 06, 2009 9:17 pm

Got it. Understand completely. I will be going to sleep in about an hour at 12/14-22/Auto/Rise6. I have felt "normal" today and can change those settings for tonight to 12/12-20/Auto/Rise6 if I hear from someone in time. Otherwise, I will make the change tomorrow or the next night depending on what I hear back from you two. Thank you very much for the continuing education. I really hope that other BASV users are gaining some benefit as well.
Pat

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: BIPAP AUTO-SV SETTINGS HELP

Post by -SWS » Sun Dec 06, 2009 10:12 pm

CROWPAT wrote:Got it. Understand completely. I will be going to sleep in about an hour at 12/14-22/Auto/Rise6. I have felt "normal" today and can change those settings for tonight to 12/12-20/Auto/Rise6 if I hear from someone in time. Otherwise, I will make the change tomorrow or the next night depending on what I hear back from you two. Thank you very much for the continuing education. I really hope that other BASV users are gaining some benefit as well.
Actually, dsm is conveying some VERY different interpretations than I am, CROWPAT. As it turns out, you're getting lots and lots of "extra" information in your thread with a good percentage of contrasting information.

I'm not sure where the 12/12-20/Auto/Rise6 setting comes from. Is that something you would like to try next? That's fine if it is. Otherwise, stick with your old standby of 12/14-22/Auto/Rise6 until you are ready to try 11/11/11 CPAP. Saying that you would rather not pursue a fixed-pressure data-collection search is perfectly fine, CROWPAT. At that point I'll resume a more passive backseat role, and offer occasional comments as your thread unfolds in any given direction or cautious line of experimentation you chose.

Also, I have that $99 CMS50D plus oximeter in both links above. The finger-clamping tension is nowhere near as tight as the probes used by most clinics. This CMS50F might be more comfortable, but at more than twice the price:
http://www.echostore.com/wrist-pulse-ox ... s-50f.html
http://www.semedicalsupply.com/cms-50f.htm

Guest

Re: BIPAP AUTO-SV SETTINGS HELP

Post by Guest » Sun Dec 06, 2009 10:47 pm

CROWPAT wrote:I will do the 3 nights and then give a try at 12/12/12.
I thought CROWPAT said he would do some nights on 12/14-22/Auto/Rise 3.

Then 12/12/12.

How were CROWPATS nights on 12/12/12?

Banned

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: BIPAP AUTO-SV SETTINGS HELP

Post by -SWS » Sun Dec 06, 2009 11:02 pm


Yeah, this seems to be yet another one of our "multi-directional" ASV online titrations. Here's my comment immediately after CROWPAT'S quoted 12/12/12 statement above:
-SWS wrote:CROWPAT, if it were me I think I'd start my downward pressure search at 11/11/11---in light of what happened at 13/13/13 (pressure toxicity).
CROWPAT'S proposed 12/12-20/Auto/Rise6 setting most likely represents either: 1) a new preference and direction of CROWPAT'S (which is just fine), or 2) confusion as CROWPAT attempts to sort out all those recent lengthy complicated ASV posts.

Guest

Re: BIPAP AUTO-SV SETTINGS HELP

Post by Guest » Sun Dec 06, 2009 11:10 pm

-SWS wrote: CROWPAT, if it were me I think I'd start my downward pressure search at 11/11/11---in light of what happened at 13/13/13 (pressure toxicity).
CROWPAT,

Your next settings will be 11/11/11. Dr. approval of course.

Stay with us,

Banned

User avatar
rested gal
Posts: 12880
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Re: BIPAP AUTO-SV SETTINGS HELP

Post by rested gal » Mon Dec 07, 2009 12:19 am

CROWPAT wrote:If SWS now has the lead that is who I will be looking to for advice.
Considering all the muddled thinking and poor settings advice that I've watched banned and DSM bring repeatedly to technical threads like this, I think you'd do well to concentrate solely on -SWS's words. That polite, diplomatic man will never push himself forward to assume "the lead", but he's definitely the one I'd pay attention to.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: BIPAP AUTO-SV SETTINGS HELP

Post by dsm » Mon Dec 07, 2009 3:46 am

rested gal wrote:
CROWPAT wrote:If SWS now has the lead that is who I will be looking to for advice.
Considering all the muddled thinking and poor settings advice that I've watched banned and DSM bring repeatedly to technical threads like this, I think you'd do well to concentrate solely on -SWS's words. That polite, diplomatic man will never push himself forward to assume "the lead", but he's definitely the one I'd pay attention to.
RG

In your own words, please do explain what was 'muddled' in my post above ?

Thanks
DSM
Last edited by dsm on Mon Dec 07, 2009 3:56 am, edited 3 times in total.
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: BIPAP AUTO-SV SETTINGS HELP

Post by dsm » Mon Dec 07, 2009 3:51 am

-SWS wrote: <snip>

Actually, dsm is conveying some VERY different interpretations than I am, CROWPAT. As it turns out, you're getting lots and lots of "extra" information in your thread with a good percentage of contrasting information.
SWS in deferrence to RG's uninformative intervention. I am really keen to look at any variance in my interpretaion of what you have said. I hope you are willing to take a pro-active response & highlight where any such misunderstanding of the points has occured.

Thanks

Doug
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

CROWPAT
Posts: 115
Joined: Sat Apr 15, 2006 7:32 pm
Contact:

Re: BIPAP AUTO-SV SETTINGS HELP

Post by CROWPAT » Mon Dec 07, 2009 8:13 am

Last night at my old settings of 12/14-22, Auto, Rise3 (or 6 - doesn't seem to matter) was a good one for me with two small PB clusters (5 total), one AP, and 4 clustered H in the middle of the night. 100% PTB and good leak.
That gets me straight again. I will try the 11/11/11 tonight and post the results.
Pat

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: BIPAP AUTO-SV SETTINGS HELP

Post by -SWS » Mon Dec 07, 2009 8:18 am

DSM, my pain levels today are very high. My intended purpose in this thread was never to focus on you or your prolific ASV views. An overall pattern that I have on this message board is the inordinate amount of time I have spent having to explain why I disagree with so many of your technical interpretations---whether it was your past insistences that CPAP + SV modality has absolutely no place in treatment (versus BiLevel + SV), that SV is an effective UARS treatment (before even researching what UARS is considered by a split medical community), or your repeated past insistences that SV is the Rolls Royce of SDB treatment (along with a variety of other metaphors indicating best possible treatment).

I have spent so much time chasing tails in pseudo-technical threads with you, that in retrospect it is staggering. I won't do it in any more threads. My earnest apologies if that somehow doesn't sit well.
dsm wrote:Thanks
Welcome
__________________________________________________________________________________________________________________________


CROWPAT, the more I think about it, the more I really question whether it is feasible to run a full gamut of CPAP, BiLevel, and SV parameter variation tests on a message board over the period of weeks. I apologize. However, the more details that come out of your sleep studies, the more I think you should consider posting as much preliminary clinical information as you possibly can. If you get lucky, Muffy just may offer advice/opinions as an extremely skilled sleep professional.

Specifically, that 90% residual SpO2 baseline is what I question most at this point in your thread. Muffy, if you are reading along will you please help and comment on that 90% residual baseline? Is that as potentially problematic as I think it can be? Thanks in advance if you can help us out.

On second thought... that's probably an untreated SpO2 baseline versus residual. At least I would hope...

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: BIPAP AUTO-SV SETTINGS HELP

Post by dsm » Mon Dec 07, 2009 1:55 pm

SWS

One thing about discussion groups is that things get discussed. The greatest learning people here get is when that happens & the feedback on this point is very well stated here by many participants. We don't all get things right just as back when the Vpap AdaptSV was announce you at first didn't agree with me that the port at the side of the machine was a pressure sensing opening (seemed perfectly logical to me). None of us is god when it comes to cpap therapy or understanding it. The medical profession is grappling with it. Discussions help us all clarify our perceptions. I am sorry your pain level has driven you to feel there is nothing to gain from highlighting where what I said was at variance with what you had said.

To RestedGal dearest, any words of clarity on the 'muddle' in my post from yesterday - please open the discussion, but be specific so we can avoid any further 'muddle'.

Cheers

DSM
Last edited by dsm on Mon Dec 07, 2009 3:49 pm, edited 1 time in total.
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: BIPAP AUTO-SV SETTINGS HELP

Post by -SWS » Mon Dec 07, 2009 3:17 pm

At this point in CROWPAT's thread my own focus returns to CROWPAT of all people. And I have flagged this statement as a potential concern:
CROWPAT wrote:No PLM in either study and O saturation remained about 90% throughout the night.
If saturation really remained at 90% throughout the titration night---and still remains that low---then I have to ask why the doctor(s) did not attempt to experimentally elevate CROWPAT's O2 baseline with supplemental oxygen. And that 90% basal SpO2 adds a risk factor to home experimentation without the benefit of SpO2 monitoring.

My understanding is that CROWPAT is going to attempt to gain access to more of his clinical records. I have also asked him to see if he can obtain PSG titration pressure tables. Hoping to get Muffy to weigh in if we can get those clinical details...

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: BIPAP AUTO-SV SETTINGS HELP

Post by dsm » Mon Dec 07, 2009 3:40 pm

Steve,
Without doubt CROWPAT's needs are primary & out of respect for your significantly valuable contribution to his well being I'll leave off adding any distractions.

Doug
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

CROWPAT
Posts: 115
Joined: Sat Apr 15, 2006 7:32 pm
Contact:

Re: BIPAP AUTO-SV SETTINGS HELP

Post by CROWPAT » Mon Dec 07, 2009 5:37 pm

I was able to get copies of almost everything from 2002 to present. Also talked with sleep doctor who said to go ahead with my experiments.
2007 sleep study showed average O2 of 96.2 for the night. 90 was the lowest level recorded and highs at various pressures were generally in the 98-100% range. The 90 number I gave earlier was from memory and not correct per the actual data. I hope that this information quells concerns about oxygen saturation levels.
Pat