Remstar Pro

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
abburk1

Remstar Pro

Post by abburk1 » Thu Oct 29, 2009 6:05 am

My husband just got a Remstar Pro w/ Cflex. It was a used one, since our insurance will not pay for it. We are having a problem. It comes up with an error 15. Does anyone know what this means? Also does anyone have a clinical maunal that they could share with us?

Thanks,

Barb

User avatar
Pugsy
Posts: 65198
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Remstar Pro

Post by Pugsy » Thu Oct 29, 2009 6:53 am

I believe this is for the Pro2 and it should be similar to the plain Pro. Wulfman should be along later to confirm or tell us what else it might be. Doesn't look good though.

E-15 ERR_ARM_UNDEFINED_INSTRUCTION_EXCEPTION •Defective FLASH chip •Replace Main PCA

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Remstar Pro

Post by Wulfman » Thu Oct 29, 2009 9:28 am

The E-15 error code for the older (plain) REMstar Pro w/C-Flex is this:

15 * Rotor Locked - The motor could not be started * Replace blower assembly If error persists, replace Main PCA

I hope you didn't pay a lot of money for it.
There are lots of inexpensive machines out there now.....CPAPAuction.com and Craig's List.


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05