My Machine turned off!!!!

Post by FatiguedMe » Sat Aug 29, 2009 7:44 am

Last night I woke up to find my machine was NOT running! I sleep with nasal pillows and this freaked me out. Has anyone had this happen before? What caused it? ACKKKKK!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Pressure 11, Padacheek Hosecover, Pursleep Buckwheat pillow & Pap Cap ALL GREAT PRODUCTS!

User avatar
El Pap
Posts: 204
Joined: Thu Mar 26, 2009 12:23 pm

Re: My Machine turned off!!!!

Post by El Pap » Sat Aug 29, 2009 7:48 am

Do you have the auto-shutoff function on your machine enabled? If so, perhaps you dislodged the mask badly enough for the auto-shutoff to kick in.

_________________
Humidifier

User avatar
FatiguedMe
Posts: 242
Joined: Thu Mar 12, 2009 7:09 pm
Location: North Carolina

Re: My Machine turned off!!!!

Post by FatiguedMe » Sat Aug 29, 2009 7:50 am

Thank you El. I will check that right now.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Pressure 11, Padacheek Hosecover, Pursleep Buckwheat pillow & Pap Cap ALL GREAT PRODUCTS!

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: My Machine turned off!!!!

Post by jdm2857 » Sat Aug 29, 2009 7:51 am

I wonder what your machine does if the power is interrupted for just a few seconds.
jeff

User avatar
FatiguedMe
Posts: 242
Joined: Thu Mar 12, 2009 7:09 pm
Location: North Carolina

Re: My Machine turned off!!!!

Post by FatiguedMe » Sat Aug 29, 2009 8:01 am

I checked and my auto off is not turned on. Plus I have the mask off alarm turned on. Nothing else in the house shows an interruption of power ex: all clocks were still on. UGHH I have no idea what happened, just know I don't want this to happen again. Concerns me because of the rebreathing thru the pillows.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Pressure 11, Padacheek Hosecover, Pursleep Buckwheat pillow & Pap Cap ALL GREAT PRODUCTS!

User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: My Machine turned off!!!!

Post by GumbyCT » Sat Aug 29, 2009 8:12 am

Check to see if the power cords are loose.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
sleepycarol
Posts: 2461
Joined: Thu Aug 30, 2007 7:25 pm
Location: Show-Me State
Contact:

Re: My Machine turned off!!!!

Post by sleepycarol » Sat Aug 29, 2009 8:19 am

I agree with Gumby as this very thing happened to me last spring and when investigating I found the power supply was loose. I guess in my sleep I some how had dislodged it. I was able to restart the machine with it being in that condition, but shudder to think that if I hadn't caught it my machine probably would have had additional episodes of shutting itself off.
Start Date: 8/30/2007 Pressure 9 - 15
I am not a doctor or other health care professional. Comments reflect my own personal experiences and opinions.

User avatar
Hawthorne
Posts: 3972
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: My Machine turned off!!!!

Post by Hawthorne » Sat Aug 29, 2009 8:20 am

I am assuming you woke up right away, as I do if the power goes out and the machine turns off. I also use nasal pillows and I am awake within seconds when this happens. There is an exhalation vent on nasal pillow masks same as on other masks so there should be no big risk.

I don't know how sensitive the machines are to power outages. We have some electrical equipment that goes off if there is even a couple of seconds of power outage, while other electrical equipment stays on for brief (seconds) power outages.

Also, as suggested, check to see that all the plugs are securely in.

Have you turned on your machine since this happened? Is it running fine?

Maybe you should consider getting a "Power Failure Alarm". I see that cpap.com is selling them for $16.95.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap

User avatar
FatiguedMe
Posts: 242
Joined: Thu Mar 12, 2009 7:09 pm
Location: North Carolina

Re: My Machine turned off!!!!

Post by FatiguedMe » Sat Aug 29, 2009 8:35 am

Plugs checked and all are fine. I rolled back over when I found it off and turned it back on. It ran fine for the rest of the night. My AHI today is .09 leaks were 32 and I feel like Cr*p. I am gonna post my report from last night and see if anyone can see something I am not seeing.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Pressure 11, Padacheek Hosecover, Pursleep Buckwheat pillow & Pap Cap ALL GREAT PRODUCTS!

User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: My Machine turned off!!!!

Post by GumbyCT » Sat Aug 29, 2009 9:02 am

As mentioned, sometimes it is not a complete outage but a dip in the power that will toggle a device into the OFF state. Some will have a different threshold. Did you have rain or storms in your area? Some power Co.'s list problems on their web page.

I think more testing is required - meaning go back to sleep. LOL

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
FatiguedMe
Posts: 242
Joined: Thu Mar 12, 2009 7:09 pm
Location: North Carolina

Re: My Machine turned off!!!!

Post by FatiguedMe » Sat Aug 29, 2009 9:29 am

GumbyCT wrote:As mentioned, sometimes it is not a complete outage but a dip in the power that will toggle a device into the OFF state. Some will have a different threshold. Did you have rain or storms in your area? Some power Co.'s list problems on their web page.

I think more testing is required - meaning go back to sleep. LOL

LOL I do need to go back to sleep! Too much to do today though! WAAAAA! If we had anything it sure doesn't look like it today. Oh well. I will just keep an eye on this and see if it occurs again.
Having problems trying to upload report from last night. It keeps uploading tiny! ARGGGG! This just is NOT my day!
HOO HUMMM and fiddle de dooooo <singing>

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Pressure 11, Padacheek Hosecover, Pursleep Buckwheat pillow & Pap Cap ALL GREAT PRODUCTS!

User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: My Machine turned off!!!!

Post by GumbyCT » Sat Aug 29, 2009 10:38 am

I like to use the link labeled "Thumbnail" bc it uploads small for those who are still using dial-up it doesn't take forever to load. But you can click the pic to see a larger image, if you choose. Use the Thumbnail link.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: My Machine turned off!!!!

Post by Wulfman » Sat Aug 29, 2009 10:50 am

There have been previous reports of this happening.......and it turned out to be the power brick.

If it persists, I'd suggest seeing if you can get a different power brick and see if that fixes it.


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
FatiguedMe
Posts: 242
Joined: Thu Mar 12, 2009 7:09 pm
Location: North Carolina

Re: My Machine turned off!!!!

Post by FatiguedMe » Sat Aug 29, 2009 12:09 pm

Ok here is what I got on my report. Looks good, but I don't feel so rested today!

Image

See! I don't understand what I am doing wrong to get this to post right. GRRRRRRRRR!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Pressure 11, Padacheek Hosecover, Pursleep Buckwheat pillow & Pap Cap ALL GREAT PRODUCTS!

User avatar
Hawthorne
Posts: 3972
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: My Machine turned off!!!!

Post by Hawthorne » Sat Aug 29, 2009 1:46 pm

You were very stressed last night when the machine went off and probably stressed the rest of the night even though you turned it back on. At least that's the way I would be!

I am not surprised that you don't feel very rested today!

I can't really see your data clearly but it does look quite good.

Hopefully it was a "one time" glitch!

I remember, early on, when I first got this machine, it started to beep during the night out of nowhere for no reason that I could see. I had an older machine and it never happened with it (same brand different model). I unplugged it and plugged it in again and everything was fine for the rest of the night.

I was stressed about what problem it might be for the rest of the night and for the next night! It never happened again! That was a year and a half ago.

Hope it is the same for you!

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap