Which is more important?
Which is more important?
Which is more important, comfort, or results?
I started off using a nasal mask, but it was fairly uncomfortable, and switched over to a full face mask (Comfortgel) which is a lot more comfortable.
I don't have the software for my machine, so I've been keeping a daily diary of my results.
The average AHI for the nasal mask is 7.2
The average AHI for the full mask is 10.9
I'm new to apnea, and don't know if these numbers are good/bad/or indifferent, but even I can see that the numbers are lower with the uncomfortable nasal mask.
IS THERE, enough difference between the two masks that I should give up the comfort of the full mask in favor of the better results of the nasal?
Another question, (if you don't mind) are my AHI numbers good or bad???
Thank's
John
I started off using a nasal mask, but it was fairly uncomfortable, and switched over to a full face mask (Comfortgel) which is a lot more comfortable.
I don't have the software for my machine, so I've been keeping a daily diary of my results.
The average AHI for the nasal mask is 7.2
The average AHI for the full mask is 10.9
I'm new to apnea, and don't know if these numbers are good/bad/or indifferent, but even I can see that the numbers are lower with the uncomfortable nasal mask.
IS THERE, enough difference between the two masks that I should give up the comfort of the full mask in favor of the better results of the nasal?
Another question, (if you don't mind) are my AHI numbers good or bad???
Thank's
John
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: ComfortGel Blue Nasal CPAP Mask with Headgear |
Re: Which is more important?
The ideal is to achieve comfort and results. Some measure of comfort is essential since discomfort over the long term will be a strong motivator to stop using the machine.
Those AHI results could be better, but don't be discouraged as they are most likely better than without your machine. With your machine you can read the results on the LCD... what does your leak rate after a night report at? Keep in mind that leak rate will be an average over the night. To see what is happening during the night with respect to leak rate you will need ResScan.
If your leak rate number is out of line then you first need to work on that... then after getting the leaks under control turn your attention to your AHI and possible pressure adjustments.
Those AHI results could be better, but don't be discouraged as they are most likely better than without your machine. With your machine you can read the results on the LCD... what does your leak rate after a night report at? Keep in mind that leak rate will be an average over the night. To see what is happening during the night with respect to leak rate you will need ResScan.
If your leak rate number is out of line then you first need to work on that... then after getting the leaks under control turn your attention to your AHI and possible pressure adjustments.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: AHI ~60 / Titrated @ 8 / Operating AutoSet in CPAP mode @ 12 |
Re: Which is more important?
John,
Resmed machines tend to log more Hypopneas per breathing pattern than do others, so, first of all, separate the AI from the HI. Do you have major differences in the AI?
Second: You won't sleep well if you're not comfotable -- and the numbers will be unreliable if you mouth leak - which happens to many of us with masks that don't cover our mouth.
It could that with a full face mask and an open mouth you need slightly more pressure, since an open mouth does contribute to more apneas.
I would try to make the more comfortable mask give me better results.
O.
Resmed machines tend to log more Hypopneas per breathing pattern than do others, so, first of all, separate the AI from the HI. Do you have major differences in the AI?
Second: You won't sleep well if you're not comfotable -- and the numbers will be unreliable if you mouth leak - which happens to many of us with masks that don't cover our mouth.
It could that with a full face mask and an open mouth you need slightly more pressure, since an open mouth does contribute to more apneas.
I would try to make the more comfortable mask give me better results.
O.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: Which is more important?
I suppose that ultimately results matter most, but one hopes to achieve optimal results + optimal comfort. Both of those AHI numbers are higher than the target max of 5.


Re: Which is more important?
Nasal leak average, 0.15
Full face average, 0.25
Both averages are actually higher than the "usual" leakage due to a couple of "out of the ordinary nights". Both masks are usually around 0.03- 0.05.
Full face average, 0.25
Both averages are actually higher than the "usual" leakage due to a couple of "out of the ordinary nights". Both masks are usually around 0.03- 0.05.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: ComfortGel Blue Nasal CPAP Mask with Headgear |
Re: Which is more important?
What are your machine settings?
What was your AI numbers in those AHI results you mentioned?
Den
What was your AI numbers in those AHI results you mentioned?
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: Which is more important?
I'm too lazy to run all the numbers right now, BUT, here's the highest AI from each mask:ozij wrote:John,
Resmed machines tend to log more Hypopneas per breathing pattern than do others, so, first of all, separate the AI from the HI. Do you have major differences in the AI?
Second: You won't sleep well if you're not comfotable -- and the numbers will be unreliable if you mouth leak - which happens to many of us with masks that don't cover our mouth.
It could that with a full face mask and an open mouth you need slightly more pressure, since an open mouth does contribute to more apneas.
I would try to make the more comfortable mask give me better results.
O.
Nasal: 0.7
Full face: 4.5
I didn't realize there was that much of a spread until you told me to check the AI results!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: ComfortGel Blue Nasal CPAP Mask with Headgear |
Re: Which is more important?
I may be off on this, if so, I'm sure someone will correct me, but with the nasal mask, if you're mouth leaking the AHI results may be meaningless because your therapy pressure is going out your mouth. With a FF it doesn't matter so much if your mouth is open as long as it's properly sealed with minimal leaks. So if the FF is more comfortable work with that one and work on the leaks. The ComfortFF never worked for my husband, he couldn't stop the leaks or sore nose bridge.
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Additional Comments: 14/8.4,PS=4, UMFF, 02@2L, |
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08
Re: Which is more important?
YIKES!!!!RipVW wrote: Both of those AHI numbers are higher than the target max of 5.
Looks like I've still got a LONG way to go.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: ComfortGel Blue Nasal CPAP Mask with Headgear |
Re: Which is more important?
I had the same problem with the sore nose bridge. It took a bit to figure out exactly what was causing it, but I found that it was the hard base of the gel pac. I trimed down a bandaid and put it over that edge, and it was fine after that.DoriC wrote: The ComfortFF never worked for my husband, he couldn't stop the leaks or sore nose bridge.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: ComfortGel Blue Nasal CPAP Mask with Headgear |
Re: Which is more important?
Pressure 14, EPR 3Wulfman wrote:What are your machine settings?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: ComfortGel Blue Nasal CPAP Mask with Headgear |
Re: Which is more important?
If your titrated/prescribed pressure is "14", then the EPR setting is dropping your "effective" pressure on exhale to "11".Komodo wrote:Pressure 14, EPR 3Wulfman wrote:What are your machine settings?
So, you need to.......either reduce your EPR setting and raise your pressure setting by that same number......or raise your pressure setting if you're going to leave your EPR setting as-is.
Unlike C-Flex on the Respironics machines, EPR actually lowers the exhale pressure by the actual (setting) number in cm.
And, since most people are (supposedly) prone to have apneas on the exhale half of the breath cycle, that's why it's important to get that pressure right.
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: Which is more important?
I took eveyones advice into consideration, but didn't make up my mind about which mask I would stick with .......... until I took a nap yesterday
After a 90 minute nap with my "comfortable" full face mask, these were my numbers:
Leak: 0.04
AHI: 20.2
AI: 3.0
HI: 17.2
THOSE numbers made my mind up for me! Forget about comfort, and go for the RESULTS!!!!
Last night I switched back to my "uncomfortable" nasal mask. I got an unbelieveable TEN HOURS of sleep, something I haven't had in well over two years!!!! Not only that, check out THESE numbers:
Leak: 0.0
AHI: 8.3
AI: 0.8
HI: 7.5
They may not be prefect numbers, but they're a HELL of a lot better than what I've been getting!
After a 90 minute nap with my "comfortable" full face mask, these were my numbers:
Leak: 0.04
AHI: 20.2
AI: 3.0
HI: 17.2
THOSE numbers made my mind up for me! Forget about comfort, and go for the RESULTS!!!!
Last night I switched back to my "uncomfortable" nasal mask. I got an unbelieveable TEN HOURS of sleep, something I haven't had in well over two years!!!! Not only that, check out THESE numbers:
Leak: 0.0
AHI: 8.3
AI: 0.8
HI: 7.5
They may not be prefect numbers, but they're a HELL of a lot better than what I've been getting!