Tonights the first night

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Sleepless in Indiana
Posts: 13
Joined: Mon Jun 15, 2009 11:18 am

Tonights the first night

Post by Sleepless in Indiana » Mon Jun 15, 2009 8:22 pm

Well I did get my machine today, and have listed the details down at the bottom. So I'm ready to start a new chapter of my life. Hope I can get some sleep before the sun comes up. Mask seems to seal at low pressure, but starts leaking as the pressure goes up. Have tried to loosen the fit up, and the only way I can get it to work is to tighten the head-gear up real tight. Will have to see how I do tonight. I did order a different mask (nasel) along with the Software today from the good folks here at the site.

I will shoot for 8 hours tonight, and see if I can make it through. I guess even a few hours will be better than none. The prescribing doc wanted to perscribe straight 20 right off the bat. No ramp or nothing. I understand I may end up there, but whenever the sleep-tech went above 14 I started waking up with air coming out all around the mask, and my sleep stages started to degrade again. So I think the auto and ramp will be good to help get use to things. Also if I can keep from sleeping on my back (which I don't do all the time anyways) I think I may be able to get by with a lower pressure. The software should help with getting everthing dialed in afer awhile.

Well got to go and get it all set up.

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: starting on auto at 12-20 cm h2o - Software and reader on the way

preemiern
Posts: 430
Joined: Fri Sep 01, 2006 8:06 pm
Location: Minnesota

Re: Tonights the first night

Post by preemiern » Mon Jun 15, 2009 8:32 pm

If you are needing pressures of 20, I would look at getting a bipap machine. Hopefully with the auto machine and the software, you will get everything fine tuned and find that you don't need that high of a pressure
Good luck tonight...hope you get a good night's rest!!
Cindy

_________________
MachineMask

User avatar
Gerald
Posts: 1352
Joined: Thu Sep 07, 2006 6:32 pm
Location: Central Louisiana

Re: Tonights the first night

Post by Gerald » Mon Jun 15, 2009 8:33 pm

Welcome to the tribe!

Your attitude is great.....and you'll do well. You have good equipment....and with the software, you'll quickly take charge of your own therapy.

The key to good therapy is to minimize the air leaks.......both leak areas....mask....and mouth.

You may have to change masks a bunch of times before you find "the one"....but, stay after it.

You can do it.......and there's a bunch of nice people on this forum who'll do all they can to help you.

Gerald

User avatar
BleepingBeauty
Posts: 2454
Joined: Thu Apr 02, 2009 5:30 pm
Location: Aridzona ;-)

Re: Tonights the first night

Post by BleepingBeauty » Mon Jun 15, 2009 8:36 pm

Sleepless in Indiana wrote:The prescribing doc wanted to perscribe straight 20 right off the bat. No ramp or nothing.
Yikes! Your prescribing doc is either a nitwit or a masochist. I say use the ramp. You can adjust it later (or do without it altogether) once you're used to the therapy.

Good luck tonight! Let us know how it goes. Of course, post when you have questions or problems. We're all here to help.
Veni, vidi, Velcro. I came, I saw, I stuck around.

Dx 11/07: AHI 107, central apnea, Cheyne Stokes respiration, moderate-severe O2 desats. (Simple OSA would be too easy. ;))

PR S1 ASV 950, DreamWear mask, F&P 150 humidifier, O2 @ 2L.

User avatar
ww
Posts: 523
Joined: Sun Mar 16, 2008 11:58 am

Re: Tonights the first night

Post by ww » Mon Jun 15, 2009 10:05 pm

Good luck. You have a good machine if you don't really require a pressure of 20. The ramp is kinda worthless after a while, but the afex should be set to on and set at 2 to start with and can be reduced to one. This will make breathing against your pressure seem much easier. The real problem may be the mask. You may want to look into the ResMed Ultra Mirage UMFF or the ResMed Quattro. They are a more stable mask than the one you have and stay put better at higher pressures.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear - Fit Pack (All Sizes Included)
Additional Comments: Titrated on Auto CPAP at 7/14 cm: Only licensed medical professionals can give medical advice or write prescriptions

Sleepless in Indiana
Posts: 13
Joined: Mon Jun 15, 2009 11:18 am

Re: Tonights the first night

Post by Sleepless in Indiana » Tue Jun 16, 2009 6:16 am

Well I made it through the night. And seemed to get some sleep.

Thanks to all that have replied with encourgament & help. I think I did pertty good, but this morning as the pressure went up (I think my apnea and sleep is worse the longer I sleeep) I started to have more issues with the mask leaking. But I made it through. Pressure seemed to stay around 13 early on, and was as 15 when I woke about 5:30 As I drifted in & out of sleep the rest time before I get up, the pressure increesed up to 18, and that is when I started fighting the mask leaks.

I will have to try the other mask that was suguested.

I did use the machine for 7.4 hours, but it doesn't look like I can read any of the other data yet. Guess I need to use it for a few more days, so the machine can compute the 7 day average. But I have ordered the software and reader already, which should arrive around the end of the week.

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: starting on auto at 12-20 cm h2o - Software and reader on the way

Sleepless in Indiana
Posts: 13
Joined: Mon Jun 15, 2009 11:18 am

Re: Tonights the first night

Post by Sleepless in Indiana » Tue Jun 16, 2009 6:40 am

Well I just went back in to empty the water out of the humidifier (do I need to empty this every morning?) and noticed that my numbers are now up on screen. May not have had quite as good as night as I thought, but still a vast improvement. Going to have to work on the leaks.

AHI - 3.3 (I was tested at 66.7)
leak rate - 50.0 l.
90% pressure was 16 cm H2O

I presume the AHI numbers may come down as the leaks are taken care of. Am I thinking right?

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: starting on auto at 12-20 cm h2o - Software and reader on the way

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Tonights the first night

Post by Pugsy » Tue Jun 16, 2009 7:29 am

Do you have the paper work that came with your mask to tell you what the intentional leak rate is at the pressure you were at for most of the night?

The higher the pressure the higher the acceptable leak rate. You need to see what is acceptable for your particular mask. You may not be as far off on the leak as you think you are. AHI of 3.3? There are people who haven't seen that good of a number despite weeks of treatment.

I would be happy with those results for a first night. You can reset your machine to zero if you don't want to mess with the 7 day average thing which really seems confusing. That is what I did till I got my software. 7 and 30 day average doesn't give a clear picture and even more so once you actually get the software and see things.

A huge leak spike will skewer the overall average but yet the leak itself (with the exception of a spike) might be very good. Not saying that is what you had but just giving this as an example.

All in all I think you did extremely well for a first night. You should be very pleased.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
LinkC
Posts: 3154
Joined: Wed Oct 29, 2008 1:06 pm
Location: Amelia Island, FL

Re: Tonights the first night

Post by LinkC » Tue Jun 16, 2009 8:29 am

Sleepless in Indiana wrote:I presume the AHI numbers may come down as the leaks are taken care of. Am I thinking right?
They may...or not. But 3.3 AHI the first night is darn good! Anything under 5.0 is considered effective therapy. Obviously the goal is to have an AHI as low as possible. Odds are yours will get lower as you "fine tune" everything (not just leaks).

Congrats! You're well on your way...

_________________
MachineMaskHumidifier
Additional Comments: 11-14 cmH2O
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Tonights the first night

Post by Pugsy » Tue Jun 16, 2009 10:38 am

Image

If the mask you used is the one listed in your profile these numbers will give you an idea of the acceptable leak rate.

If you want to watch daily data instead of averages you will need to reset your therapy usage data to zero, then your 7 day "average" will be daily data.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Sleepless in Indiana
Posts: 13
Joined: Mon Jun 15, 2009 11:18 am

Re: Tonights the first night

Post by Sleepless in Indiana » Tue Jun 16, 2009 2:45 pm

So if you subtract the mask leakage rate, does that mean that my actual leak rate was about 6.5 l. - Is that a good rate. I'm guessing that is just last nights average of no leaks, and largee leaks as the pressure increesed.

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: starting on auto at 12-20 cm h2o - Software and reader on the way

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Tonights the first night

Post by Pugsy » Tue Jun 16, 2009 3:05 pm

Sleepless in Indiana wrote:So if you subtract the mask leakage rate, does that mean that my actual leak rate was about 6.5 l. - Is that a good rate. I'm guessing that is just last nights average of no leaks, and largee leaks as the pressure increese
In an ideal world we want the leak rate to be as close to zero as possible. It is so hard to gauge leak rates just on this one number that you have. What if you had one or two big spikes that caused the leak rate to show an elevated average numerically where in actual use you were really pretty much close to zero except for the spikes? Even if your leak was at 6 L/min the entire night at your pressures, it still isn't that bad especially for the first night.

We rarely get an ideal world and leak rates will vary from night to night. I think you did great and I wish my leak rate last night had only been 6. Try not to dwell on one number from one night. Instead look at the overall picture and factor in how you feel.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Gerald
Posts: 1352
Joined: Thu Sep 07, 2006 6:32 pm
Location: Central Louisiana

Re: Tonights the first night

Post by Gerald » Tue Jun 16, 2009 4:09 pm

Sleepless.....

Pat yourself on the back...."you done good!"

Achieving an AHI of 3.3 on your first run is great.

I respectfully disagree with LinkC concerning his opinion that anything under 5.0 is acceptable. Based on my experiments with an Oximeter, I believe that one must stay under an AHI of 2.5 to achieve effective therapy. I fully realize that my opinion flies in the face of "conventional wisdom"....but, I'm not afraid to say that I think conventional wisdom is wrong.

Congratulations for a good job on your first night.

Now....start "tweakin"

Gerald

cflame1
Posts: 3311
Joined: Sat Mar 11, 2006 6:55 am
Location: expat Canadian in Kentucky

Re: Tonights the first night

Post by cflame1 » Tue Jun 16, 2009 6:35 pm

The thing that could have happened with that AHI though, is that the leaks... could have been classed as large leaks in which the AHI won't be accurate as it stops recording.

User avatar
Gerald
Posts: 1352
Joined: Thu Sep 07, 2006 6:32 pm
Location: Central Louisiana

Re: Tonights the first night

Post by Gerald » Tue Jun 16, 2009 7:15 pm

With cFlame raising the possibility of large leaks, it's easy to see why software is vital.

Without it is like driving in the dark......no way to see we're going....and no way to see where we've been.

Sleepless....you're moving in the right direction. The picture will become more clear when you get your software up and running.

Gerald