			I'm still trying (as a new user) to understand what the EP report is telling me!
I really would appreciate input from those of you that have used the software enough to describe to me what is being shown in the attached report.
Everything looks good until about the 3rd hour ... then the Leak goes up and goes nuts ... at the same time the Obstructive Apnea kicks in at a rapid rate ... then of course the pressure has gone up to the max 10 cm.
Very simply ... what is the EP report telling me?  What's happening?
The only difference from the info in my profile is that I wore a Full Face Mask for this report (had trouble breathing through my nose).  I was on Auto with A-Flex and at 6 to 10 cm.
I was awake for the last half to one hour before getting up (back pain).
Thanks ... 
Jan
The report:

 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								Julie							
- Posts: 20052
- Joined: Tue Feb 28, 2006 12:58 pm
						
						
													
							
						
									
						Post
					
								by Julie » Sat May 02, 2009 11:04 am
			
			
			
			
			Is it just possible that around the 3 hr mark, you moved, dislodged the seal a bit, created a leak and it allowed some events (apneas) to occur?  There isn't always a good, let alone rational, explanation for everything that goes on in a given night, and unless it becomes routine (the timing, etc.), I wouldn't worry a lot about it.
			
			
									
									
						 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								Wulfman							
- Posts: 12317
- Joined: Thu Jul 07, 2005 3:43 pm
- Location: Nearest fishing spot
						
						
													
							
						
									
						Post
					
								by Wulfman » Sat May 02, 2009 11:19 am
			
			
			
			
			Jan,
Now you know what makes some of us "daily downloaders".....it's like a box of chocolates.....you never know what you're going to get (see).    
And, in Auto mode (range of pressures), it can REALLY get interesting.    
Den
			
			
									
									(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm. 
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
						 
		 
				
		
		 
	 
	
				
		
		
			- 
				
																			 BeanMeScot
- Posts: 588
- Joined: Fri Aug 01, 2008 11:05 am
						
						
													
							
						
									
						Post
					
								by BeanMeScot » Sat May 02, 2009 11:44 am
			
			
			
			
			My opinion?  You need to tighten up the range on your machine.  Using APAP works best when the low number is no more than 2cm less than your 90% number.  Your machine appears to be set wide open, 4-20.  Since your 90% is at 10, I would set the bottom number to 8 and see how you do.  If this is a "weird" chart for you, check the 90% on a more normal chart and use that to set the bottom number.  Starting out at a higher pressure might also help you to get the mask adjusted for the pressures you will experience during the night so you won't get major leaks.  It should also help to reduce your AHI.
			
			
									
									
						 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								Wulfman							
- Posts: 12317
- Joined: Thu Jul 07, 2005 3:43 pm
- Location: Nearest fishing spot
						
						
													
							
						
									
						Post
					
								by Wulfman » Sat May 02, 2009 11:57 am
			
			
			
			
			janp wrote:Very simply ... what is the EP report telling me? What's happening? 
I think you should consider switching to CPAP mode.  Your initial pressure of 6 cm. seemed to be doing pretty good for you.  I'd turn off the Ramp and just set it to 6 cm. in CPAP mode for a few nights and see what the reports look like.  That will prevent the runaway pressure changes if your nasal pillows become dislodged again.
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm. 
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
						 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								janp							
- Posts: 167
- Joined: Sun Mar 15, 2009 10:07 pm
- Location: California
						
						
													
							
						
									
						Post
					
								by janp » Sat May 02, 2009 12:59 pm
			
			
			
			
			BeanMeScot wrote:My opinion?  You need to tighten up the range on your machine.  Using APAP works best when the low number is no more than 2cm less than your 90% number.  Your machine appears to be set wide open, 4-20.  Since your 90% is at 10, I would set the bottom number to 8 and see how you do.  If this is a "weird" chart for you, check the 90% on a more normal chart and use that to set the bottom number.  Starting out at a higher pressure might also help you to get the mask adjusted for the pressures you will experience during the night so you won't get major leaks.  It should also help to reduce your AHI.
BeamMeScot ...
Thanks for your response.
Actually, the Auto is set to 6cm, with 10cm max.
Except for the spikes in pressure ... I assume responding to apneas ... the pressure did stay at the 6 minimum and all was well ... until it went wild for some reason.
I'm trying to understand what that reason was.
I tend to agree with you on the tighter range, but my doctor just changed the setting to 6/10, so I'm going to try it for a week for evaluation.
Jan
 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								janp							
- Posts: 167
- Joined: Sun Mar 15, 2009 10:07 pm
- Location: California
						
						
													
							
						
									
						Post
					
								by janp » Sat May 02, 2009 1:05 pm
			
			
			
			
			Wulfman wrote:janp wrote:Very simply ... what is the EP report telling me? What's happening? 
I think you should consider switching to CPAP mode.  Your initial pressure of 6 cm. seemed to be doing pretty good for you.  I'd turn off the Ramp and just set it to 6 cm. in CPAP mode for a few nights and see what the reports look like.  That will prevent the runaway pressure changes if your nasal pillows become dislodged again.
Den
 
Den,
It was in CPAP mode at 6cm and wasn't helping (confirmed by the EP reports).
The doctor just changed it to the Auto 6 to 10 cm.  
Doesn't the first part of the pressure graph show that it's working when it spikes higher than 6?  Am I not reading the information correctly?
Thanks,
Jan
 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								Wulfman							
- Posts: 12317
- Joined: Thu Jul 07, 2005 3:43 pm
- Location: Nearest fishing spot
						
						
													
							
						
									
						Post
					
								by Wulfman » Sat May 02, 2009 1:58 pm
			
			
			
			
			janp wrote:Den,
It was in CPAP mode at 6cm and wasn't helping (confirmed by the EP reports).
The doctor just changed it to the Auto 6 to 10 cm.  
Doesn't the first part of the pressure graph show that it's working when it spikes higher than 6?  Am I not reading the information correctly?
Thanks,
Jan
Those little pressure "bumps" or "probes" are a characteristic of the REMstar Autos.  If you'll notice, there are no "events" below or above them in that 2-hour period to warrant it changing pressures......no FL, no hypopneas, apneas, snores.....just a few little dips in your leak line.  If you look at the Events per Hour chart, you spent 118 min. at 6 and had an AHI of 1.5.  Everything above that pressure reported higher numbers of events (although that could be a cause-and-effect situation).....probably due to the leakage.  The Auto (range of pressures) didn't actually do any thing FOR you in those first 2 3/4 hours.  For all intents and purposes, you were at a constant pressure and there was only one flow limitation and one apnea (which, if you'll notice, happened during a flat spot in your leak line and there was no corresponding bump in pressure on the pressure line).  That apnea came and went without a response from the machine.
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm. 
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
						 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								Wulfman							
- Posts: 12317
- Joined: Thu Jul 07, 2005 3:43 pm
- Location: Nearest fishing spot
						
						
													
							
						
									
						Post
					
								by Wulfman » Sat May 02, 2009 2:03 pm
			
			
			
			
			Another thought or two.....would be to set your pressure range from 6 - 7 of use a straight pressure of 7 and see how that looks.
Den
			
			
									
									(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm. 
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
						 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								Wulfman							
- Posts: 12317
- Joined: Thu Jul 07, 2005 3:43 pm
- Location: Nearest fishing spot
						
						
													
							
						
									
						Post
					
								by Wulfman » Sat May 02, 2009 2:11 pm
			
			
			
			
			I went back through your previous posts to refresh my memory....
Was that the Comfortfull 2 (FFM) you were wearing last night?  
Den
			
			
									
									(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm. 
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
						 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								janp							
- Posts: 167
- Joined: Sun Mar 15, 2009 10:07 pm
- Location: California
						
						
													
							
						
									
						Post
					
								by janp » Sat May 02, 2009 2:13 pm
			
			
			
			
			Wulfman wrote:I went back through your previous posts to refresh my memory....
Was that the Comfortfull 2 (FFM) you were wearing last night?  
Den
Den,
It was the Comfortgel FFM.
Jan
 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								Wulfman							
- Posts: 12317
- Joined: Thu Jul 07, 2005 3:43 pm
- Location: Nearest fishing spot
						
						
													
							
						
									
						Post
					
								by Wulfman » Sat May 02, 2009 2:16 pm
			
			
			
			
			Oooooooo.....K.
Den
			
			
									
									(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm. 
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
						 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								Goofproof							
- Posts: 16087
- Joined: Mon Dec 05, 2005 3:16 pm
- Location: Central Indiana, USA
						
						
													
							
						
									
						Post
					
								by Goofproof » Sat May 02, 2009 2:33 pm
			
			
			
			
			You were going along well until the leak rate went over 40 LPM, after that you were no longer being treated, not enough air pressure being held in your airway to hold it open. You need to correct the leaks and as Den says a higher starting pressure without a ramp and more control for the high pressure. APAP 7cm to 8 cm, or CPAP at 7 CM with the leaks stopped.   Jim
			
			
									
									Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
						 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								janp							
- Posts: 167
- Joined: Sun Mar 15, 2009 10:07 pm
- Location: California
						
						
													
							
						
									
						Post
					
								by janp » Sat May 02, 2009 3:15 pm
			
			
			
			
			Goofproof wrote:You were going along well until the leak rate went over 40 LPM, after that you were no longer being treated, not enough air pressure being held in your airway to hold it open. You need to correct the leaks and as Den says a higher starting pressure without a ramp and more control for the high pressure. APAP 7cm to 8 cm, or CPAP at 7 CM with the leaks stopped.   Jim
  
  
Jim,  
  
Thanks ...  
  
Two questions.  
  
I was doing great on the leak rate until the last 1-1/2 hours.  Any guesses as to what changed?  In fact the leak rate seems to drop everytime the pressure went up ... I assume the pressure went up responding to an apnea event.  Am I wrong?
  
How does the ramp come into this ... since that was at the start of the night and the leak/pressure/apnea were all under control for almost 3 hours afterwards?  
  
Bear with me, I'm still very new to all of this and trying to understand it all.  
  
Jan
 
		 
				
		
		 
	 
	
				
		
		
			- 
				
								Wulfman							
- Posts: 12317
- Joined: Thu Jul 07, 2005 3:43 pm
- Location: Nearest fishing spot
						
						
													
							
						
									
						Post
					
								by Wulfman » Sat May 02, 2009 4:26 pm
			
			
			
			
			Several factors can enter into this......
If the mask (headgear tension) is adjusted for the lower pressures, when the pressure goes up too far, it can lose the seal.
Bed pillow.....if the pillow dislodges the mask (especially when one is sound asleep), the leaking will continue until the person moves to a different position or wakes up enough to rearrange it so it will stop.
Sleeping position.....if the person's airway becomes restricted, it will cause/allow the apneas to take place.
Den
			
			
									
									(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm. 
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow