Detailed data for S8 Elite II and Resscan 3.7

Post by ChrisC » Wed Apr 08, 2009 5:29 am

I'm having some issues with downloading detailed data. When I put in my card and download, it only shows me 'summary' data. Is there somewhere on the machine I have to set to make it record detailed data? Also, it is showing me summary results for AHI/AI/HI, but the data for leak rate is absolutely nil.. as if it's not recording it at all. When I access the Elite II's LCD, it shows me daily leak/sec with no problems.
"Human beings, who are almost unique in having the ability to learn from the experience of others, are also remarkable for their apparent disinclination to do so."
-- Douglas Adams

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Re: Detailed data for S8 Elite II and Resscan 3.7

Post by Slinky » Wed Apr 08, 2009 5:50 am

I have no experience w/the S8 Elite II but the S8 Elite was not capable of providing the Detailed Data w/o the assistance of the S8 ResLink which is an additional expensive device.
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

ScottyT
Posts: 43
Joined: Thu Oct 30, 2008 1:33 pm
Location: Toronto, Ontario

Re: Detailed data for S8 Elite II and Resscan 3.7

Post by ScottyT » Wed Apr 08, 2009 6:18 am

The Elite II is capable of detailed data, though not quite as much as you'd get with Reslink. Reslink would let you see every breath you took and your snoring and so forth. Anyway, on the download screen in Resscan, you'll see the data section. It'll say something like "summary data only." Click the button to the right and select all available data. After that, you should be good to go.

User avatar
ChrisC
Posts: 176
Joined: Tue Nov 25, 2008 11:39 am
Location: Ontario, Canada

Re: Detailed data for S8 Elite II and Resscan 3.7

Post by ChrisC » Wed Apr 08, 2009 10:02 am

Ahh... perfect, Scott! Thank-you very much. It only had it for the last few days, but at least I got some of the detailed data!!
"Human beings, who are almost unique in having the ability to learn from the experience of others, are also remarkable for their apparent disinclination to do so."
-- Douglas Adams