IntelliPAP Auto 19, M Series Auto A-Flex 11

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
SleeplessInNyack
Posts: 33
Joined: Thu Mar 19, 2009 6:52 am

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by SleeplessInNyack » Thu May 21, 2009 1:14 pm

That's good info, SWS; Although the audio sounds like a 33RPM played back on an auto 16-33RPM player!

They should either redo the audio or transcribe the webinars to text.

Kitmastin
Posts: 3
Joined: Wed Nov 14, 2007 4:31 pm

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by Kitmastin » Fri May 22, 2009 4:21 pm

We also offer free CEU credits for clinicians who participate in the courses on http://www.devilbissclinicaleducation.com site.

Thank you for the feedback on the audio recordings. We are looking at alternatives to record the audio for our webinars. In addition, we will be adding additional educational courses within the next few months.

As, always - we are listening. Feel free to offer up any comments or suggestions.

SleeplessInNyack
Posts: 33
Joined: Thu Mar 19, 2009 6:52 am

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by SleeplessInNyack » Fri May 22, 2009 7:27 pm

Kitmastin wrote:We also offer free CEU credits for clinicians who participate in the courses on http://www.devilbissclinicaleducation.com site.

Thank you for the feedback on the audio recordings. We are looking at alternatives to record the audio for our webinars. In addition, we will be adding additional educational courses within the next few months.

As, always - we are listening. Feel free to offer up any comments or suggestions.
Actually, the audio on the second Webinar was not bad, just the first one is a bit funky.

I just want to say that I think it's terrific that your making these courses available.

Kitmastin
Posts: 3
Joined: Wed Nov 14, 2007 4:31 pm

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by Kitmastin » Fri May 22, 2009 9:36 pm

We're glad that you find the webinars useful. We are working on a fix for the audio issues and should have them repaired in the next month. If you have any other suggestions - courses you would like to see presented, etc...please let me know. You can always email me at Kristin.Mastin@DeVilbisshc.com.

Thanks again for the feedback.

-SWS
Posts: 5285
Joined: Tue Jan 11, 2005 7:06 pm

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by -SWS » Fri May 22, 2009 9:45 pm

Nice!

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by ozij » Fri May 22, 2009 10:49 pm

Thank you, Kit,
On my desktop,
  • the audio quality on the second was reasonable, but the sound was very weak.
    Please note that there is a major time lag between the audio and the appearance of case study charts it the second webinar. That makes the last part of the presentation almost impossible to follow - a pity because it is interesting to hear the analysis.
    The scoring for the second quiz is totally messed up - (check the first question and you will see that immediately).

Any chance of making those webinars dowloadable Powerpoint shows, in addition to video?

O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

SleeplessInNyack
Posts: 33
Joined: Thu Mar 19, 2009 6:52 am

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by SleeplessInNyack » Sat May 23, 2009 11:56 am

Kitmastin wrote:We're glad that you find the webinars useful. We are working on a fix for the audio issues and should have them repaired in the next month. If you have any other suggestions - courses you would like to see presented, etc...please let me know. You can always email me at Kristin.Mastin@DeVilbisshc.com.

Thanks again for the feedback.
One thing I'd like to see on the reports is the ability to zoom into an hour and see what the events are minute by minute. Also, I'd like to see the airflow line chart that's shown in the Webinars to see the details of the breathing pattern that's being analyzed. After all, now that we're learning how to differentiate an apnea from an hypopnea, it would be nice to see the raw data or line chart.

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by ozij » Sun May 24, 2009 8:56 pm

Kit,
According to the webinars, Devilbiss assumes a drop in flow of more than 95% is central, and it will not respond to those apneas. Can you point us to studies supporting those assumptions?
Do you have info about the percentages of false positive and false negatives when that rule is used?

Thank you
O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by ozij » Wed Jun 03, 2009 10:01 am

Bumping this up in the hope of getting a response to my week old question:
Kit,
According to the webinars, Devilbiss assumes a drop in flow of more than 95% is central, and it will not respond to those apneas. Can you point us to studies supporting those assumptions?
Do you have info about the percentages of false positive and false negatives when that rule is used?

Thank you
O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

natediggidydogg

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by natediggidydogg » Sat Mar 26, 2011 10:56 pm

yardbird wrote:Your data looks almost exactly like mine. Last night I had ZERO apneas recorded and I think that's a first for me. However I had a LOT of hypopnea events and they wake me up so I need to track them down. I also think it's important to know what a particular setting MEANS if we're going to make changes. The way I'm reading the Advanced Menu options seems to be indicating that the percentage setting is is a figure representing how far below normal flow we must fall before registering as an event.

For example (just tossing out easy numbers, not meant to be representative of actual):

Let's say our normal flow is 100

Default Hypopnea% is 50%
I read this as: Respiratory flow must drop to 50% of normal and stay at or below 50% of normal for the time specified in Hypopnea Duration. Default for that is 10 seconds.
So respiratory flow of 50 for 10 seconds registers as a Hypopnea event.

If we change Hypopnea% to 40% and change nothing else.
Respiratory flow must fall 40% from normal for 10 seconds. This means you still HAVE 60% respiratory flow. You've made the machine more sensitive because now your respiratory flow only needs to drop 40% from normal instead of having to drop to 50% below normal.

If we change Hypopnea% to 60% and change nothing else.
Respiratory flow must fall 60% from normal for 10 seconds. This means you still HAVE 40% respiratory flow. You've made the machine less sensitive because now your respiratory flow needs to drop 60% from normal before registering an event.

So the apnea% or the hypopnea% are not percentages of what you HAVE, but rather percentages of how far you have fallen off of the normal.

A hypopnea% setting of 40% would be telling the machine that if I reach a respiratory flow that is down 40% from my normal (not 40% OF my normal), start working on this as a hypopnea event. A hypopnea setting of 50% would ignore that same scenario and wait until I've reached a 50% reduction from normal before reacting.

So LOWERING that percentage number makes the machine more sensitive, and RAISING it makes the machine less sensitive to these events.

Now here's a bigger question...
If the machine pressure is set, such that I have ZERO apnea events recorded, but lots of hypopnea events. Do I want to CHASE the hypopnea events by making these small and granular adjustments or do I do something much simpler and INCREASE my low pressure setting in an effort to keep the airway splinted open and PREVENT the hypopnea events?

BE VERY CAREFUL HERE. YOUR ABOVE CALCULATIONS ARE EXACTLY THE OPPOSITE OF WHAT THEY SHOULD BE.

Changing hypoapnea to 40% makes the machine LESS sensitive to the event.
Changing hypoapnea to 60% makes the machine MORE sensitive to the event

Here's why: look at the "apnea" options. We have 0% - 20%. That means an apnea can be defined by 0% of normal breathing volume (a true apnea) to 20% of normal breathing volume. The same rules apply to hypoapneas: setting at 60% (instead of 50%) means we are adjusting the machine so that it defines hypoapneas as 60% normal breathing volume. We are making the machine more sensitive bc it is now responding to an smaller event.

User avatar
Bobby269
Posts: 803
Joined: Sat Nov 27, 2010 11:12 am
Location: San Antonio Texas

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by Bobby269 » Sun Mar 27, 2011 4:13 pm

The new model has a feature called smartflex http://www.devilbisshealthcare.com/prod ... /smartflex
It should make it better for people who like assistance in breathing when using high pressure.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Set to Apap 14.00/20. EPR 3, Climate line hose, Oscar for Mac 1.1.0. Betta.1

User avatar
DeVilbiss Marketing
Posts: 116
Joined: Sat Mar 28, 2009 9:35 am

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by DeVilbiss Marketing » Mon Mar 28, 2011 2:32 pm

I am checking with our Clinical Team for studies.

User avatar
DeVilbiss Marketing
Posts: 116
Joined: Sat Mar 28, 2009 9:35 am

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by DeVilbiss Marketing » Tue Mar 29, 2011 6:22 pm

Ozij - Dr Scharf conducted a study of the performance of our device compared to classical hand scoring. The study concluded that the AutoAdjust Algorithm accurately detected respiratory events, including "mixed" (central) events, when compared to a hand scored record during an 8.5 hour evaluation of which 237 events where scored. Manual respiratory event scoring versus device event detection scoring and event durantion correlated extremely well. A second method of statistical analysis revealed an event detection sensitivity and specificity of 86.8% and 98.4% respectively.

Unfortunately, this study is not published.

Evette Toepels

...

Post by Evette Toepels » Wed Mar 30, 2011 12:22 am

Hello. And Bye.

marylandcpap
Posts: 7
Joined: Mon Jan 04, 2010 6:14 pm
Location: Hagerstown, MD
Contact:

Re: IntelliPAP Auto 19, M Series Auto A-Flex 11

Post by marylandcpap » Sun Apr 17, 2011 3:46 pm

natediggidydogg wrote:
yardbird wrote:Your data looks almost exactly like mine. Last night I had ZERO apneas recorded and I think that's a first for me. However I had a LOT of hypopnea events and they wake me up so I need to track them down. I also think it's important to know what a particular setting MEANS if we're going to make changes. The way I'm reading the Advanced Menu options seems to be indicating that the percentage setting is is a figure representing how far below normal flow we must fall before registering as an event.

For example (just tossing out easy numbers, not meant to be representative of actual):

Let's say our normal flow is 100

Default Hypopnea% is 50%
I read this as: Respiratory flow must drop to 50% of normal and stay at or below 50% of normal for the time specified in Hypopnea Duration. Default for that is 10 seconds.
So respiratory flow of 50 for 10 seconds registers as a Hypopnea event.

If we change Hypopnea% to 40% and change nothing else.
Respiratory flow must fall 40% from normal for 10 seconds. This means you still HAVE 60% respiratory flow. You've made the machine more sensitive because now your respiratory flow only needs to drop 40% from normal instead of having to drop to 50% below normal.

If we change Hypopnea% to 60% and change nothing else.
Respiratory flow must fall 60% from normal for 10 seconds. This means you still HAVE 40% respiratory flow. You've made the machine less sensitive because now your respiratory flow needs to drop 60% from normal before registering an event.

So the apnea% or the hypopnea% are not percentages of what you HAVE, but rather percentages of how far you have fallen off of the normal.

A hypopnea% setting of 40% would be telling the machine that if I reach a respiratory flow that is down 40% from my normal (not 40% OF my normal), start working on this as a hypopnea event. A hypopnea setting of 50% would ignore that same scenario and wait until I've reached a 50% reduction from normal before reacting.

So LOWERING that percentage number makes the machine more sensitive, and RAISING it makes the machine less sensitive to these events.

Now here's a bigger question...
If the machine pressure is set, such that I have ZERO apnea events recorded, but lots of hypopnea events. Do I want to CHASE the hypopnea events by making these small and granular adjustments or do I do something much simpler and INCREASE my low pressure setting in an effort to keep the airway splinted open and PREVENT the hypopnea events?

BE VERY CAREFUL HERE. YOUR ABOVE CALCULATIONS ARE EXACTLY THE OPPOSITE OF WHAT THEY SHOULD BE.

Changing hypoapnea to 40% makes the machine LESS sensitive to the event.
Changing hypoapnea to 60% makes the machine MORE sensitive to the event

Here's why: look at the "apnea" options. We have 0% - 20%. That means an apnea can be defined by 0% of normal breathing volume (a true apnea) to 20% of normal breathing volume. The same rules apply to hypoapneas: setting at 60% (instead of 50%) means we are adjusting the machine so that it defines hypoapneas as 60% normal breathing volume. We are making the machine more sensitive bc it is now responding to an smaller event.
WHO IS RIGHT HERE? CAN SOMEONE PLEASE TELL ME ?

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Additional Comments: Diagnostic PSG, AHI 89 O2 60% nadir