A-Flex anomaly

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
JimIllinois
Posts: 104
Joined: Sat Feb 07, 2009 3:36 pm
Location: Illinois, USA

A-Flex anomaly

Post by JimIllinois » Tue Mar 10, 2009 10:40 am

Hello, everyone.

In the process of adjusting the low side of my auto settings, I decided to experiment with A-Flex. The doc had specified C-Flex in the prescription, but my machine supports both. I immediately noticed something I thought was a little strange, and I ended up going back to C-Flex.

During an inhale, at about 80% through, I would get a stutter in the pressure. It seemed to drop to near-zero for a very short period. Maybe 1/10th of a second.

This occurred with every breath, and caused a "can't breath!" reaction in me. After the stutter, everything else (the remainder of the inhale, and the entire exhale) was normal. I had a very hard time going to sleep, with this going on. Maybe my breathing was too deep at this point, but wow, was it an uncomfortable sensation.

Is this a known artifact of the A-Flex routine? Anyone else experience this?

Thanks!

BeanMeScot
Posts: 588
Joined: Fri Aug 01, 2008 11:05 am

Re: A-Flex anomaly

Post by BeanMeScot » Tue Mar 10, 2009 10:56 am

I tried A-Flex on my machine and I could never quite identify what I didn't like about it. It made me feel like my lungs were empty was as close as I could describe it. I tried it for 2 nights and it still bothered me so I went back to C-Flex. It felt very comfortable to breath with the A-Flex but it still bothered me.

User avatar
Georgio
Posts: 608
Joined: Thu Aug 28, 2008 5:34 pm
Location: Jacksonville Beach, Fl

Re: A-Flex anomaly

Post by Georgio » Tue Mar 10, 2009 10:58 am

Yes, both my mother and I both experience the "stutter" or momentary drop in pressure, that makes it difficult to inhale for a fraction of a second every night. I don't recall ever having heard this subject discussed here. In our case, it only occurs when we first lay down and take 3 or 4 very deep breaths before settling into a more normal, and not so deep breathing pattern. In other words, it only does it when we take very deep breaths, so we only experience it for 3 or 4 breaths. It never occurs during our "normal" breathing. I speculated that our lung capacity is larger than the norm, and that the A-Flex algorithm does not "fit" us perfectly at that time. Not a big deal, since in our opinion the A-Flex feature makes breathing so much more normal feeling, we are able to overlook that small flaw.

I hope that more will contribute to this discussion, and that Resperonics gets the message.

Thanks for bringing it up!

Georgio
DreamStation 2, Oscar
Resmed AirFit P30i Nasal Mask

User avatar
OldLincoln
Posts: 779
Joined: Wed Mar 26, 2008 7:01 pm
Location: West Coast

Re: A-Flex anomaly

Post by OldLincoln » Tue Mar 10, 2009 11:23 am

I also was uncomfortable with A-Flex initially. It seemed the machine was in a hurry for me to take the next breath. That was when it was set on 3. I lowered it to 2 and it was in sync with my normal pattern. Now I honestly can't tell if my machine is running a lot of the time and often tug on it to hear the leak.

So, before giving up on A-Flex, you might try different settings. BTW, I had the EPR on my ResMed CPAP and I find A-Flex much better.
ResMed AirSense 10 AutoSet / F&P Simplex / DME: VA
It's going to be okay in the end; if it's not okay, it's not the end.

User avatar
Georgio
Posts: 608
Joined: Thu Aug 28, 2008 5:34 pm
Location: Jacksonville Beach, Fl

Re: A-Flex anomaly

Post by Georgio » Tue Mar 10, 2009 11:28 am

Yes, Old Lincoln, I believe you had the same sensation, as though the machine's algorithm was telling you, "OK that breath was long enough, take another one!" It is possible to continue with that particular deep breath, after you get past the momentary resistance. I would think this could be an easy fix for Resperonics.
Last edited by Georgio on Tue Mar 10, 2009 11:52 am, edited 1 time in total.
DreamStation 2, Oscar
Resmed AirFit P30i Nasal Mask

JimIllinois
Posts: 104
Joined: Sat Feb 07, 2009 3:36 pm
Location: Illinois, USA

Re: A-Flex anomaly

Post by JimIllinois » Tue Mar 10, 2009 11:49 am

Thanks for the comments, it's good to know I wasn't dreaming it. And I will experiment with different A-Flex settings.

I don't have a clue what their algorithm looks for and how it is supposed to respond, but it seems like it could wait 5 minutes to see how deep the breathing is, and adjust accordingly. But, what do I know...

User avatar
Babette
Posts: 4231
Joined: Mon Apr 30, 2007 5:25 pm

Re: A-Flex anomaly

Post by Babette » Tue Mar 10, 2009 1:59 pm

I'm also an Aflex hater. I don't remember the "stutter" but there was definitely something in there I didn't like. I slept horribly and felt like I was back in the initial stages of therapy. I lasted two weeks and went back to cflex.

Some people really love it. Others hate it. Glad I've got friends.
B.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Started XPAP 04/20/07. APAP currently wide open 10-20. Consistent AHI 2.1. No flex. HH 3. Deluxe Chinstrap.
I currently have a stash of Nasal Aire II cannulas in Small or Extra Small. Please PM me if you would like them. I'm interested in bartering for something strange and wonderful that I don't currently own. Or a Large size NAII cannula. :)

User avatar
roster
Posts: 8153
Joined: Wed Jan 18, 2006 8:02 pm
Location: Chapel Hill, NC

Re: A-Flex anomaly

Post by roster » Tue Mar 10, 2009 4:21 pm

Glad to hear it is working for some folks. In my experience, the problem with tracking my breathing was a mental one. When I was awake, I felt like I had to keep up with the machine. But I assume when I fell asleep, I just breathed naturally and the machine follow me closely.

I don't use the feature because of these two problems:

1. The noise of the machine cycling up and down tended to keep me awake longer. It is much quieter with the Flex features off.

2. I needed to set the pressure 2 to 3 cm higher with A-Flex on to achieve the same AHI. If I remember correctly the machine drops the pressure about 2 cm on exhale. So I had to set it up by 2 cm wiping out this part of the relief and now the inhale pressure was 2 cm higher. Counterproductive.

If I ever see a large number of people here saying the like the A-Flex feature, I might give it one more go.

Glad to be able to read what you are saying about A-Flex,

Hose_Head
Posts: 804
Joined: Wed Mar 12, 2008 6:43 pm
Location: Ontario Canada

Re: A-Flex anomaly

Post by Hose_Head » Tue Mar 10, 2009 5:54 pm

JimIllinois wrote:Hello, everyone.

In the process of adjusting the low side of my auto settings, I decided to experiment with A-Flex. The doc had specified C-Flex in the prescription, but my machine supports both. I immediately noticed something I thought was a little strange, and I ended up going back to C-Flex.

During an inhale, at about 80% through, I would get a stutter in the pressure. It seemed to drop to near-zero for a very short period. Maybe 1/10th of a second.

This occurred with every breath, and caused a "can't breath!" reaction in me. After the stutter, everything else (the remainder of the inhale, and the entire exhale) was normal. I had a very hard time going to sleep, with this going on. Maybe my breathing was too deep at this point, but wow, was it an uncomfortable sensation.

Is this a known artifact of the A-Flex routine? Anyone else experience this?

Thanks!

I use Aflex setting of 2. My pressure is 7 - 12. I can't say that I've ever noticed any of these symptoms. I've used Cflex before. Can't say I noticed much of a difference between Aflex and Cflex. I guess that makes me an indiscriminate breather!

Personally, I love them both! I too sometimes need to lift my mask to make sure it's still working. To me the breathing pattern is no natural that I cannot tell that I'm breathing against pressure.
I'm workin' on it.

User avatar
Georgio
Posts: 608
Joined: Thu Aug 28, 2008 5:34 pm
Location: Jacksonville Beach, Fl

Re: A-Flex anomaly

Post by Georgio » Tue Mar 10, 2009 6:33 pm

Hose Head, when you go to bed, put your mask on, lay down and take 3 or 4 verydeep long breaths. Inhale for 2 seconds or so. See if you feel a "hiccup" in the middle of your breath. That is what we are experiencing. (This occurs at our ramp pressure of 7).

Georgio
DreamStation 2, Oscar
Resmed AirFit P30i Nasal Mask

danw61
Posts: 124
Joined: Wed Aug 29, 2007 5:17 pm
Location: Southern California
Contact:

Re: A-Flex anomaly

Post by danw61 » Tue Mar 10, 2009 7:37 pm

Georgio wrote:Hose Head, when you go to bed, put your mask on, lay down and take 3 or 4 verydeep long breaths. Inhale for 2 seconds or so. See if you feel a "hiccup" in the middle of your breath. That is what we are experiencing. (This occurs at our ramp pressure of 7).

Georgio
I feel the same thing since I recently switched to A-Flex to try it out. It's kind of irritating. Feels like the machine is trying to rush my breathing. I may switch back to C-Flex.

Dan

User avatar
elader
Posts: 529
Joined: Wed Oct 08, 2008 7:45 am
Location: Maryland

Re: A-Flex anomaly

Post by elader » Tue Mar 10, 2009 7:58 pm

I had the same issue and switched back to c flex

_________________
MaskHumidifier

User avatar
tdm5032c
Posts: 89
Joined: Sat Nov 29, 2008 3:15 pm
Location: Richland, WA

Re: A-Flex anomaly

Post by tdm5032c » Tue Mar 10, 2009 8:22 pm

I have been using the A-flex since I started on CPAP on Thanksgiving of this last year. I have tried C-flex and to me, it felt totally unnatural and was hard to breathe with it. I have tried switching back and forth several times and always end up back with the A-flex at 2. Just works the best for me and seems completely natural. I haven't noticed any stuttering in the machine even when I do breathe deeply. I guess we all breathe a little bit different, and different modes and settings work better for some than others. Probably why they make different settings

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: Encore Pro Software
You can't have everything....where would you put it?

P.S. I LOVE/HATE my Swift LT!!

User avatar
Georgio
Posts: 608
Joined: Thu Aug 28, 2008 5:34 pm
Location: Jacksonville Beach, Fl

Re: A-Flex anomaly

Post by Georgio » Tue Mar 10, 2009 9:58 pm

Well, we're on A-Flex 2 also and experience the flow resistance.....maybe we have a series of Resperonics a-Flex machines that have a "defective code" or something that is causing the "hiccup" when we are trying to breathe. Don't get me wrong, because I love my Resperonics machine and it has done wonders for me...but they will be hearing more about the "hiccup" phenomenon.
DreamStation 2, Oscar
Resmed AirFit P30i Nasal Mask

User avatar
Captain_Midnight
Posts: 761
Joined: Sat Jan 28, 2006 8:10 pm
Location: The Great State of Idaho

Re: A-Flex anomaly

Post by Captain_Midnight » Tue Mar 10, 2009 10:32 pm

Jim writes...During an inhale, at about 80% through, I would get a stutter in the pressure. It seemed to drop to near-zero for a very short period. Maybe 1/10th of a second.

Hmmm, interesting. I agree with above posters that ask what AFLEX setting you are using, Jim.

I've been using AFLEX (@ AFLEX = 2) for about a year and a half (after 2 yrs of CFLEX). I was going to experiment w other AFLEX settings, but I liked 2 so much, I've never looked back. As Babs says, some folks don't resonate w AFLEX, and that's fine. But this is the first I've heard of a "stutter". I would definitely consider experimenting w AFLEX 2 and 1, if you happen to be on 3.

The biggest problem w AFLEX that I can ascertain (from reading posts here and elsewhere) is not having the lower pressure of the range set high enough. For many (including moi) it needs to be at least 1 cm higher than the titrated pressure, otherwise there might be apneic occlusions just after exhalation. (No science to back this up, so please feel free to challenge.) Maybe the pressure delta at the end of inhalation and beginning of exhalation is too great for you at your AFLEX setting.

Good luck finding a mode and setting that works best for you.

.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: ComfortGel Blue Nasal CPAP Mask with Headgear
Additional Comments: APAP range = 10 - 12.5 In H20