Looking for comments on this data!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Hawthorne
Posts: 3972
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Looking for comments on this data!

Post by Hawthorne » Thu Feb 12, 2009 9:34 am

[img]http://i252.photobucket.com/albums ... .jpg[/img]

I am posting these 2 days of data for Ainsle14.

He never had a titration study, just a first sleep study to daignose. He has a MSeries Auto with A-Flex and is using a Hybrid full face mask.

He has only been on therapy about a week but having trouble.

I have been helping by email but want some input from others here.

Initially, his range was set a 8cm to 20 cm. Two night ago we lowered the upper pressure to 16 cm since he was jumping all over the place.

Then, last night, we set the pressure range 8.5 to 16. I am thinking we need to raise the minimum probaly quite a bit yet but mask leaks seem to be a major issue.

The first image is last night (Pressure 8.5 -16). The second image is a couple of night's ago (he forgot to put his card in the night between). In the second image his pressure setting was 8 to 16. The second image is quite typical of the previous nights on therapy.

He is using the mask 4-5 hours a night and then taking it off.

Any comments and ideas welcome!
Thanks.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap

User avatar
rested gal
Posts: 12880
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Re: Looking for comments on this data!

Post by rested gal » Thu Feb 12, 2009 9:57 am

Hawthorne wrote:I am posting these 2 days of data for Ainsle14.
Has he got untreated or undertreated GERD, by chance?
Hawthorne wrote:Initially, his range was set a 8cm to 20 cm. Two night ago we lowered the upper pressure to 16 cm since he was jumping all over the place.
Good thinking... that you lowered his max for the time being, given what all was happening in the first data results.
Hawthorne wrote:Then, last night, we set the pressure range 8.5 to 16. I am thinking we need to raise the minimum probaly quite a bit yet but mask leaks seem to be a major issue.
I think you're right. Get control of the leaks, and raise the minimum pressure. Probably to 10.

Just my layperson's opinion, of course, as I'm no doctor or anything in the health care field.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435

SharkBait
Posts: 593
Joined: Sun Jan 25, 2009 5:41 pm
Location: Texas -- the ugly part... El Paso? No, not quite THAT ugly...

Re: Looking for comments on this data!

Post by SharkBait » Thu Feb 12, 2009 10:23 am

I've been on treatment all of 1 week (minus 1 day ), but here goes... looking at that graph and seeing where they were getting good rest I'd try setting them around 10 on CPAP and see what happens. APAP was kinda messing with me. Not to that extent, but I wasn't setting my range as high so it might have done so given the chance...

Same here... not a doctor... I don't play one on television...
Encore Pro 1.8.49; Encore Pro Analyzer 0.8.9 by James Skinner
SnuggleHose - Got the 8 foot and cut it down to 6, used the rest for mask hoses.
Memory Foam Pillow - Cut my own out of my Tempur-pedic pillow. (works great!)
Hose Mgmt - Velcro Tie Strap

SharkBait
Posts: 593
Joined: Sun Jan 25, 2009 5:41 pm
Location: Texas -- the ugly part... El Paso? No, not quite THAT ugly...

Re: Looking for comments on this data!

Post by SharkBait » Thu Feb 12, 2009 10:25 am

Chicken/egg scenario -- are the leaks causing the pressure increase or is the increase in pressure causing the leaks? In a couple of cases where it went hay wire it looks like the latter...
Encore Pro 1.8.49; Encore Pro Analyzer 0.8.9 by James Skinner
SnuggleHose - Got the 8 foot and cut it down to 6, used the rest for mask hoses.
Memory Foam Pillow - Cut my own out of my Tempur-pedic pillow. (works great!)
Hose Mgmt - Velcro Tie Strap

User avatar
Gerald
Posts: 1352
Joined: Thu Sep 07, 2006 6:32 pm
Location: Central Louisiana

Re: Looking for comments on this data!

Post by Gerald » Thu Feb 12, 2009 11:58 am

Hawthorn...

I'd say....first.....get the leaks under control....that leak line needs to be as flat as possible.

Once the leak rate is pretty level (meaning fairly constant), start walking the low-end pressure up a half-point at a time.

FL - Flow limitations are "Baby Hypopneas"
Hi - Hypopneas are "Baby Apneas"
AI - Apneas are "Mature Hypopneas....that are busy trying to kill you"

Get'em before they "mature"....by bumping the pressure up....in a controlled....logical.....experimental way.

ozij
Posts: 10536
Joined: Fri Mar 18, 2005 11:52 pm

Re: Looking for comments on this data!

Post by ozij » Thu Feb 12, 2009 2:05 pm

Lood at the first graph and compare what happened at about hour 2 (intensive breathing disruption with leaks) to what happened at about hour 4.

Hour 1-2 had many leaks, may many snores, many apneas, even at the highest pressure.

Hour 4 had crazy leaks at less pressure.

When the pressure is too low, apneas can casue restlessness which can casue leak - and you can't handle them as long as the casue persists.

Based on all that, my (lay person's) suggestion would be to raise the minimum pressure to 11.5 or 12.

O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

old64mb
Posts: 216
Joined: Mon Feb 02, 2009 10:02 pm

Re: Looking for comments on this data!

Post by old64mb » Thu Feb 12, 2009 4:21 pm

Hawthorne wrote:...is using a Hybrid full face mask.
I don't know enough to comment on the data, but one other question - is there a particular reason he needs to use a FF mask?

User avatar
Hawthorne
Posts: 3972
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: Looking for comments on this data!

Post by Hawthorne » Fri Feb 13, 2009 5:45 am

old64mb - he mouth breathes so a full face mask is his choice over taping.