Does anyone know how to set the pressure on a ResMed S7 Lightweight? It is obviously a CPAP.
Setting pressure on a ResMed S7 Lightweight
- LilBuckaroo
- Posts: 169
- Joined: Sat Dec 11, 2004 10:35 pm
- Location: Montgomery, AL
Setting pressure on a ResMed S7 Lightweight
Harold
Until later, stay safe, be nice, sleep well and keep on papin'
Until later, stay safe, be nice, sleep well and keep on papin'
Here is another approach
1) Turn on your machine and wait for the "Welcome "message to complete
2) Press the right button (with the red line on it - it should also have the word EXIT above it on the display)
3) Now the display should show "S7 Lightweight " or "S7 Elite" or S7 Spirit" and the word MENU should show over the left button."
4) Press both the right button and the down-arrow button at the same time and hold down for at least 3 seconds.
5) "CLINICAL MENU - WELCOME" should show in the display, then after 3 seconds the display will change again to ...
6) "CLINICAL - MENU exit"
7) Press the left button with the word MENU above it
Next you should see "SETTINGS - ENTER exit" (if SETTINGS doesn't show it may be "RESULTS" or "OPTIONS" or "SERVICING" if it is any of these other words then use the middle up/down toggle button to navigate to "SETTINGS"
9) When you have "SETTINGS" in the display, press the left button with the word "ENTER" above it (this now takes you into the SETTINGS menu)
10) Then use the up/down toggle button in the middle of the panel, to navigate to the settings you want to change - when you have the setting you want to change press the left button with the word "change" above it.
11) Use the up/down toggle button to vary the settings.
To get out just keep pressing the right button. It will have either "cancel" or "exit" written above it. Keep pressing it until back to normal ("S7 Elite" or S7 Lightweight" or "S7 Spirit").
Now that was easy, wasn't it
Cheers
DSM
2) Press the right button (with the red line on it - it should also have the word EXIT above it on the display)
3) Now the display should show "S7 Lightweight " or "S7 Elite" or S7 Spirit" and the word MENU should show over the left button."
4) Press both the right button and the down-arrow button at the same time and hold down for at least 3 seconds.
5) "CLINICAL MENU - WELCOME" should show in the display, then after 3 seconds the display will change again to ...
6) "CLINICAL - MENU exit"
7) Press the left button with the word MENU above it
Next you should see "SETTINGS - ENTER exit" (if SETTINGS doesn't show it may be "RESULTS" or "OPTIONS" or "SERVICING" if it is any of these other words then use the middle up/down toggle button to navigate to "SETTINGS"
9) When you have "SETTINGS" in the display, press the left button with the word "ENTER" above it (this now takes you into the SETTINGS menu)
10) Then use the up/down toggle button in the middle of the panel, to navigate to the settings you want to change - when you have the setting you want to change press the left button with the word "change" above it.
11) Use the up/down toggle button to vary the settings.
To get out just keep pressing the right button. It will have either "cancel" or "exit" written above it. Keep pressing it until back to normal ("S7 Elite" or S7 Lightweight" or "S7 Spirit").
Now that was easy, wasn't it
Cheers
DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)
Re: Setting pressure on a ResMed S7 Lightweight
Didnt work. Got none of the menu choices or messages listed.
Re: Setting pressure on a ResMed S7 Lightweight
Go here and request the clinical manual via email and read it. The instructions are in the manual.
https://www.apneaboard.com/adjust-cpap- ... tup-manual