Raising pressure on a RemStar Plus???

Post by Nealbopper » Sat Dec 27, 2008 12:06 pm

Purchased a RemStar Plus on line and it arrived today. I need to up the pressure from 7 H2O to 10 H2O.
Anyone have a clinical menue and a method of changing the settings? I would appreciate the help.
Thank you, Neal

_________________
MachineMask


User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Raising pressure on a RemStar Plus???

Post by Wulfman » Sat Dec 27, 2008 12:12 pm

Nealbopper wrote:Purchased a RemStar Plus on line and it arrived today. I need to up the pressure from 7 H2O to 10 H2O.
Anyone have a clinical menue and a method of changing the settings? I would appreciate the help.
Thank you, Neal
Which model is it? The M Series, the Legacy model WITHOUT C-Flex or the Legacy model WITH C-Flex?

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Nealbopper
Posts: 70
Joined: Thu Dec 18, 2008 9:41 am
Location: Oakland Twp. Michigan

Re: Raising pressure on a RemStar Plus???

Post by Nealbopper » Sat Dec 27, 2008 12:33 pm

It's the remstar plus. it doesn't say M series or C-Flex on it so i'm guessing it's the base model? Sound right?

_________________
MachineMask

User avatar
Nealbopper
Posts: 70
Joined: Thu Dec 18, 2008 9:41 am
Location: Oakland Twp. Michigan

Re: Raising pressure on a RemStar Plus???

Post by Nealbopper » Sat Dec 27, 2008 12:35 pm

Thanks jules for the links, I already have the pressure changed. Great help! Neal

_________________
MachineMask

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Raising pressure on a RemStar Plus???

Post by Wulfman » Sat Dec 27, 2008 12:41 pm

Nealbopper wrote:Thanks jules for the links, I already have the pressure changed. Great help! Neal
If it has the two gray "user" buttons < > below the LCD display, it's probably the model with C-Flex (five buttons on top of the machine), and you would have come across the C-Flex setting in the Setup screens.
The model without those two gray buttons and has a total of three buttons on top is the model without C-Flex.

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

jules
Posts: 3304
Joined: Mon Nov 27, 2006 10:51 pm

Re: Raising pressure on a RemStar Plus???

Post by jules » Sat Dec 27, 2008 12:41 pm

yvw -

that thread included links to all of it I think - m series, and the two possible tanks that Den had linked

User avatar
Nealbopper
Posts: 70
Joined: Thu Dec 18, 2008 9:41 am
Location: Oakland Twp. Michigan

Re: Raising pressure on a RemStar Plus???

Post by Nealbopper » Sat Dec 27, 2008 12:47 pm

3 buttons only, base without C-flex. What is C-flex anyhow?

_________________
MachineMask

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Raising pressure on a RemStar Plus???

Post by Wulfman » Sat Dec 27, 2008 12:52 pm

Nealbopper wrote:3 buttons only, base without C-flex. What is C-flex anyhow?
If it's the one with three buttons (no C-Flex), I'm surprised you got into the Setup menu with the other instructions.

These are the instructions for the one w/o C-Flex.

C-Flex is Respironics' version of Exhale Relief (EPR in the ResMed machines)

Den



REMstar Plus wo/C-Flex Setup

IMPORTANT!
When the device is in the Therapy Setup Menu, the humidifier and ramp buttons operate as up and down keys to change the settings and the pressure on/off button will take you to the next screen.

Therapy Setup Menu
1. Plug the pronged end of the power cord into an electrical outlet. To enter the Therapy Setup Menu hold the ramp and pressure on/off buttons down while plugging the power cord into the device. The airflow will automatically turn ON and the display screen will show the current elevation setting and the unlock symbol.

IMPORTANT!
Prescribed therapy settings can be set only using the Therapy Setup Menu.
To prevent patients from tampering with the settings, the directions to access the Therapy Setup Menu should not be revealed to the patient.

Elevation Setting
a. The elevation setting and symbol will be displayed.
The patient also has access to this setting in the Patient Setup Menu.
Lo = less than 2,500 ft. (<762 m.)
M = 2,500 to 5,000 ft. (762 m. to 1524 m.)
Hi = 5,000 to 7,500 ft. (1525 m. to 2286 m.)

Note: Altitudes of 7,500 ft. (2,286 m.) may affect the accuracy of the pressure. Verify the pressure settings with a water column manometer.

To change the elevation setting, press and release the ramp or humidifier button until the correct setting is displayed.
Press and release the pressure on/off button to go to the next setting.

Pressure Setting
b. The CPAP pressure setting will appear.
To change the setting, press and release the ramp or humidifier button.
Range: 4 to 20 cm H2O (in 1 cm. H2O increments)
Press and release the pressure on/off button to go to the next setting.

Fine Adjustment Setting
c. The CPAP fine adjustment setting will appear. This setting allows you to calibrate the device so that the pressure setting can be verified with a manometer.

If you DO NOT want to calibrate the device, press and release the pressure on/off button to go to the next setting.

If you DO want to calibrate the device, follow the directions below:
1. Connect the patient circuit to the REMstar Plus. Make sure there is an exhalation leak in the circuit.

2. Zero the manometer and connect the manometer at the patient mask. Make sure the pressure has stabilized (let the manometer sit for at least 60 seconds).

3. If the pressure setting is not the same as the manometer reading, press and release the ramp or humidifier button to change the setting.
Range: -0.9 to 0.9 H2O (in 0.1 cm H2O increments)
Press and release the pressure on/off buton to go to the next setting.

Ramp Time Setting
d. The Ramp Time symbol and setting will appear. The patient also has access to this setting in the Patient Setup Menu.
To change the setting, press and release the ramp or humidifier button.
Range: 0 to 45 minutes (in 5 minute increments)
Press and release the pressure on/off button to go to the next setting.
Note: If the ramp time is set to "0" minutes, or if the CPAP pressure setting is 4 cm. H2O, the following ramp setting will not be displayed. Go to Step f.

Ramp Starting Pressure Setting
e. The Ramp Starting Pressure symbol and setting will appear. To change the setting, press and release the ramp or humidifier button.
Range: 4 to Prescription Setting (in 1 cm. H2O increments)
Press and release the pressure on/off button to go to the next setting.

>4 Setting
f. The number of times this device has been used for more than four consecutive hours (>4) will appear. To erase the total and go back to zero, press and release the ramp or humidifier button.
An "X" will appear in the upper left corner and the total will change to "0000". (Pressing the button again will restore the original count)
Press and release the pressure on/off button to go to the next setting.

LED Setting
g. The LED on/off symbol and setting will appear. This setting allows you to have the lights behind the button turned on or off.
The patient also has access to this setting in the Patient Setup Menu.
1 = ON
0 = OFF
To change the setting, press and release the ramp or humidifier button.

Exit Settings
h. To go back to the beginning of the settings, press and release the pressure on/off button.

To exit the Therapy Setup Menu, hold the pressure on/off button down and press the ramp button one time. Release the on/off button. The airflow will turn off.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Guest

Re: Raising pressure on a RemStar Plus???

Post by Guest » Tue Aug 17, 2010 7:34 pm

Nealbopper wrote:It's the remstar plus. it doesn't say M series or C-Flex on it so i'm guessing it's the base model? Sound right?

User avatar
LoQ
Posts: 1475
Joined: Sun Apr 06, 2008 1:59 pm
Location: America

Re: Raising pressure on a RemStar Plus???

Post by LoQ » Tue Aug 17, 2010 8:04 pm

Does anyone see anything more than the quote in the previous post by Guest? I'm seeing these again, where the post contains only a quote. Occasionally there is stuff written INSIDE THE QUOTE that's really hard to see, but I see lots that don't seem to have that.


This problem (part of the post below the quote is not displayed) happened once before but they fixed it. I'm just wondering if it's happening again.