Mask Leak Rates

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
kopoloff
Posts: 398
Joined: Mon Nov 03, 2008 5:20 am
Location: Melbourne Australia

Mask Leak Rates

Post by kopoloff » Mon Dec 01, 2008 4:25 am

Hi all

I have an Opus 360, but can't find a leak chart anywhere. Does anyone have this data, or know where I can get it?

Thanks

K

ozij
Posts: 10488
Joined: Fri Mar 18, 2005 11:52 pm

Re: Mask Leak Rates

Post by ozij » Mon Dec 01, 2008 4:28 am

Mask manual or CD? It's a detail the mask maker is supposed to report.
O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
kopoloff
Posts: 398
Joined: Mon Nov 03, 2008 5:20 am
Location: Melbourne Australia

Re: Mask Leak Rates

Post by kopoloff » Mon Dec 01, 2008 4:38 am

Thanks ozij, but I didn't get a CD, and the manual had nothing at all

Perhaps I'll contact F & P (but then they're NZ company, and they don't like Aussies at the moment because we smashed them at cricket today)

K

ozij
Posts: 10488
Joined: Fri Mar 18, 2005 11:52 pm

Re: Mask Leak Rates

Post by ozij » Mon Dec 01, 2008 5:09 am

and they don't like Aussies at the moment because we smashed them at cricket today

Can you masquerade as someone else? Or maybe - just maybe, I, I truly hesitate to suggest such sacrilege, maybe you could congratulate them on their victory before you start talking? Or would doing that doom you to a fate worse than... apnea?



O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
kopoloff
Posts: 398
Joined: Mon Nov 03, 2008 5:20 am
Location: Melbourne Australia

Re: Mask Leak Rates

Post by kopoloff » Mon Dec 01, 2008 6:00 am

I'm going to break my own rules here

wot?

User avatar
bdp522
Posts: 3378
Joined: Tue Jun 27, 2006 5:13 pm

Re: Mask Leak Rates

Post by bdp522 » Mon Dec 01, 2008 7:10 am

Here ya go;

pressure 3........leak 17
pressure 5........leak 22
pressure 7........leak 26
pressure 9........leak 30
pressure 11......leak 33
pressure 13......leak 36
pressure 15......leak 39
pressure 17......leak 42
pressure 19......leak 45
pressure 21......leak 48
pressure 23......leak 50
pressure 25......leak 52


Brenda

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Love my papillow, Aussie heated hose and PAD-A-CHEEKS! Also use Optilife, UMFF(with PADACHEEK gasket), and Headrest masks Pressure; 10.5

User avatar
Kiralynx
Posts: 2426
Joined: Wed Oct 08, 2008 5:42 am

Re: Mask Leak Rates

Post by Kiralynx » Mon Dec 01, 2008 8:12 pm

http://www.sleeprestfully.com/prodimage ... s/opus.pdf

Scroll down a bit and they have an exhaust chart at different pressures.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Additional Comments: Sleepyhead software, not listed. Currently using Dreamstation ASV, not listed
-- Kiralynx
Beastie, 2008-10-28. NEW Beastie, PRS1 960, 2014-05-14. NEWER Beastie, Dream Station ASV, 2017-10-17. PadaCheek Hosecover. Homemade Brandy Keg Chin Support. TapPap Mask.
Min PS = 4, Max PS = 8
Epap Range = 6 - 7.5

ozij
Posts: 10488
Joined: Fri Mar 18, 2005 11:52 pm

Re: Mask Leak Rates

Post by ozij » Mon Dec 01, 2008 11:21 pm

Thanks for the link, K.
O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
kopoloff
Posts: 398
Joined: Mon Nov 03, 2008 5:20 am
Location: Melbourne Australia

Re: Mask Leak Rates

Post by kopoloff » Tue Dec 02, 2008 4:04 pm

Sorry - I've been off the air for a couple of days, but thanks for the input.

Cheers

K