No Flow Limitations in CPAP Mode?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ca_hosehead
Posts: 150
Joined: Sat Dec 09, 2006 3:51 pm

No Flow Limitations in CPAP Mode?

Post by ca_hosehead » Sun Nov 23, 2008 3:27 am

I switched my Respironics M Series APAP over to straight CPAP mode. I notice in the reports I no longer see Flow Limitations. Does anyone know why this is?

Can I turn them back on somehow? I was thinking I could switch it back to APAP mode with top and bottom pressures set to the same number but I don't know if that will work.

ozij
Posts: 10527
Joined: Fri Mar 18, 2005 11:52 pm

Re: No Flow Limitations in CPAP Mode?

Post by ozij » Sun Nov 23, 2008 5:45 am

Switch your machine to APAP with min.=max. and you'll get fixed pressure therapy, with APAP detail reporting.

O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
ca_hosehead
Posts: 150
Joined: Sat Dec 09, 2006 3:51 pm

Re: No Flow Limitations in CPAP Mode?

Post by ca_hosehead » Sun Nov 23, 2008 1:51 pm

Lovely, thanks so much.

User avatar
Wulfman
Posts: 12316
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: No Flow Limitations in CPAP Mode?

Post by Wulfman » Sun Nov 23, 2008 6:58 pm

ca_hosehead wrote:I switched my Respironics M Series APAP over to straight CPAP mode. I notice in the reports I no longer see Flow Limitations. Does anyone know why this is?

Can I turn them back on somehow? I was thinking I could switch it back to APAP mode with top and bottom pressures set to the same number but I don't know if that will work.
The reason (my interpretation) is that since a flow limitation is a "partial hypopnea", it's also a clue to the system's algorithm that something bigger may be coming.....so....."get ready to bump up the pressure". (and CPAPs don't change pressure)

You also won't see any Non Responsive events logged in CPAP mode.