-SWS wrote:snoredog wrote:BPM=4 is 15 second breaths, 7.5:7.5 I:E
Well, I think with a max IT=3.0 seconds (as you correctly mention in the next quote) that 7.5:7.5 I:E ratio is impossible on this machine. So I take it that ratio you mentioned was a 1:1 reference of equal-distribution. Okay, I follow that part.
The above was my math
in thinking out loud, I had subtracted the 3 second maximum IT time possible from the I:E and didn't add the 4.5 seconds back to remaining to the exhale time (just the way I do math). But when you start defining what BPM will be, the only part of that breath you are defining here on this machine is the IT time from .5 seconds to 3 seconds (max).
So that 7.5:7.5 ratio shown is my bad for not completing the calculation and shifting the ratio over to the
correct result of: 3:12 for a BPM=4 setting. BPM=## defines the overall length or duration of the breath that you will be using in fixed backup settings. So BPM is a critical setting and that breath is then carved up
by the IT time used, so IT is what carves up that total breath time ratio. The I:E ratio is not a 1:1 ratio and that should not be automatically assumed.
1. You define how long the breath will be from BPM setting (60 seconds /BPM=duration of I:E)
2. You define how long Inspiration time will be (remaining time will be allocated for exhale).
3. You define how fast Rise Time will be.
You also have to remember that the person is NOT breathing when switched or entering the backup mode. They are NOT breathing because they failed to take an Inspiration breath. So that time is tagged on to the prior time allotment for the previous exhale time. So when you define BPM=xx where xx is 4 to 30, you are telling the machine
what to expect from Inspiration to Inspiration, then you are defining what the IT time should be. I then read here where someone had used BPM=4, the "minimum". When you apply this machines logic to that even though it is an allowable setting it doesn't make any sense.
-SWS wrote:
If BPM=4 and IT=3 (min and max values respectively) the unlikely exhalation time that would be spontaneously "allowed" by the machine would be a whopping 15 seconds minus 3 seconds (derived from BPM=4 minus IT=3). Actually a little more than those 11 seconds might be "allowed" if the patient spontaneously assumes an IT of less than 3 seconds (that IT=3 is only an upper limit in S/T mode, unlike PC mode where IT is a fixed value). So there are supposedly only three parameters/restrictions that come into numeric play with S/T mode here: 1) BPM setting, 2) IT setting, and 3) a restriction that IT cannot exceed half of breath rate.
I have not read where there is a "restriction" that IT cannot exceed half of breath rate. The SV has a limitation that IT time cannot exceed 3.0 seconds. I am unaware of any restrictions as of this writing on the S/T.
snoredog wrote: In BPM=Auto mode digital auto-track will find your timing settings and RR
I stand behind that statement, digital auto trak is used to detect inspiration and exhalation times in SV mode. In BPM mode those spontaneous times will be carried over to BPM=Auto mode. BPM value however will be minus -2 of what was found in Spontaneous mode.
-SWS wrote:
I agree that the autoSV algorithm will maintain an averaged BPM.
Digital auto-track is also Respironics' leak control logic. Can you believe they dedicated an entire web domain to Digital auto-trak?
http://autotrak.respironics.com/
Yes I can believe they dedicated a webpage for it, have you read it? Because this is what it says on the first page
which is all I have read in this reference (more of my selective reading I guess):
Respironics website wrote:Inside and out, the entire Flex Family of pressure relief technologies relies on two Respironics innovations: Digital Auto-Trak Sensitivity and the M Series sleep systems. Digital Auto-Trak, the brains behind all of our Flex Family pressure relief technologies, is a highly sensitive algorithm that detects the onset of inspiration
and expiration – even in the presence of mask leaks. Being able to track, react and respond to each and every breath makes it possible for C-Flex, A-Flex and Bi-Flex to deliver the right pressure relief at the right moment.
Digital Auto-Trak has the ability to recognize and compensate for unintentional and intentional leaks in the system, and to automatically adjust its variable trigger and cycle thresholds to maintain optimum performance in the presence of leaks. No manual adjustments are required.
The Digital Auto-Trak components are all active concurrently. The machine reacts to the first component that detects a change, providing maximum sensitivity.
Sorry, but I read the above to mean this machine does a very good job at finding your Inspiration and Exhale times. xFlex seems to work pretty good in finding those timings as well (not many people complain about it). So I stand behind my statement that this machine uses Digital Auto Trak to find your I:E timings in SV mode.
snoredog wrote:The idea with BPM mode is to slow your breathing down, that is what BPM=Auto does, using fixed takes it further and not only does it slow breathing down it makes it slightly uncomfortable where the body says it don't like staying there and starts breathing spontaneously again.
Snoredog stands behind the above statement too. When you enter that mode you are not breathing with a central, when you do start to breath I'm
sending you back to spontaneous mode at a slower rate than you were breathing before the central occurred or the breathing that contributed to the central event.
-SWS wrote:
I'm confused how a backup rate can force or directly drive slower breathing. I can see how a backup rate can speed up breathing: it simply shoots the patient an EPAP-to-IPAP pressure transition before the patient would have spontaneously initialized their own E-to-I transition. That's easy to do. But I'm still confused how a backup rate can force a slowdown in anybody's breath rate.
The machine either issues the EPAP-to-IPAP pressure transition, or it withholds that pressure transition. If the machine issues that pressure transition quicker than the patient's spontaneous transition, then patient gets a little "hey let's speed things up" reminder. But if the machine decides to withhold that EPAP-to-IPAP pressure transition for say, 10 more seconds.... the patient doesn't get anything from the machine whatsoever by the way of a reminder or pressure transition. At that point the spontaneous respiratory drive is simply left to it's own devices regarding that very next respiratory trigger.
So I'm not certain how an S/T machine that has no purely Timed mode can slow a patient's breathing down? If the autoSV had PC mode (also called Timed or T mode), then it could do what you describe (although S mode would never be resumed by the patient). According to all the Respironics marketing brochures and setup guides the autoSV just doesn't offer any PC or purely Timed mode (T). Rather, the autoSV only offers a patient either spontaneous mode (S) or backup modem (S/T). The BiPAP Synchrony and AVAPs machines, on the other hand, do offer a PC or purely Timed mode (T).
I haven't totally dissected exactly what the S/T is doing or not doing, so I cannot comment on it, so my comments are with what the Adapt SV is doing or not doing for time being.
On the SV, when the patient gets sent to backup mode from system detecting a central, they are NOT breathing. IF BPM=Fixed settings is input the amount of time elapsed for next IPAP Min puff will be as defined by the BPM=## value used just as I explained above and elsewhere. The BPM defined will be intentionally defined with a slower value than seen in Spontaneous mode. It will do the SAME when BPM=Auto. Spontaneous minus -2. Any Inspiration "duration" time will be defined by IT=value. Rise time will define how fast that Inspiration will be delivered. With those fixed settings input, the machine establishes when it will cycle from IPAP to EPAP. Person lands there, 3 seconds is exceeded, machine is going to cycle to EPAP, once the time for that has elapsed IPAP Min pressure will be delivered. From Inspiration to Inspiration will be the time alloted by BPM. By using a ridiculously low BPM=4 when they do start spontaneous breathing again, it will be like hitting a brick wall. When they go back to SV side they will be all screwed up again with breathing.
Think of how central dysregulation begins, it starts by fast hyperventilation breathing or very slow and/or deep breathing. Just as the SV side targets Peak volumes, the backup mode is the same but you define that by the IT time and BPM used.
How can I say this nicely? As you know that is hard for me at times, so all I can say is: Don't assume you have the BPM mode totally nailed down as to its functional understanding of it, because that is where we continue to have a misunderstanding. Having that understanding is critical in the way you go about setting this machine up or correcting a particular event. There is your understanding of how it works, dsm's understanding of how it works and SAG's understanding of how it works. To be honest (and again with all due respect), neither one of you guys have gotten it right so far. If you set BPM=Off the patient better not have any central dysregulation period. Sure if you set BPM=Off, the Patient Triggered Breathing will go to 100%. Now you might think that is working much better but that is only because you have completely disabled the backup mode and your ability to see what is happening. Setting BPM=Off you have NO indicator you are going into Central Dysregulation as you can no longer see it.
That also means you are positive the SV side is going to correct breathing which contributed to the central dysregulation. Now if the SV was truly a panacea for controlling Central Dysregulation flawlessly there would be no need at all for a BPM mode. They put it there for a reason. And assuming that is simply not the case. While I agree the SV side can correct that condition through control of breathing, if the patient is continually going to central dysregulation as a result of SV side settings, you have removed any fail-safe mode to correct it. And that is ALL BPM mode is,
a fail-safe way to correct central dysregulation. When a patient lands in the BPM mode they should only be there for a very short time, if you can use that short period of time in BPM to put them in a more favorable state for when they do return to spontaneous breathing you are ahead of the game. If you do nothing in response to a central it will resolve itself. Even in the case of Bev, where they were seen as high as 84 seconds even those resolve themselves. But if a person is going 84 seconds with a central, and you induce Insp for 1.2 seconds of one pressure then 4.8 seconds of EPAP pressure they will eventually start following that cycle of pressure. IF the goal behind BPM was to speed up a patients breathing they wouldn't suggest you use a value of minus -2 of spontaneous.
So under no circumstances is this machine's logic for BPM designed to speed-up a person's breathing. So if your understanding is the patient is going to BPM mode because they are breathing too slow, then well you have that backwards too.
As you know when you shift that I:E center line ratio left or right of center (what I meant by 7.5:7.5 above) you are controlling Hypercapnia:Hypocapnia dysfunction. So by looking at a Encore report and its indicators which does a person have?
And with all due respect, I don't mean to be argumentative here about this, but for some time now there has been a clear misconception by both you and SAG (and dsm) on how this mode works. Now I realize that you guys are never wrong and try to get funny about it when someone calls you on it, but I don't blame dsm one bit for NOT understanding it either, he is only following the suggestions made by you two and well neither one of you have gotten it right to date. Sorry to be so blunt, but that is the way I see it.
Now I'll be happy to discuss the function of the BPM mode for a better understanding of it, go through any patent you want and basically discuss it. But the rules are you cannot automatically assume this machine has a function or a technology in use just because some other machine this manufacturer happens to make has it. Bringing into this discussion on what a Resmed machine does also is not a valid explanation/excuse on how this machine functions either. A lot of people that read these boards and rely on what you say as the "gold standard" on how a machine functions. This is not about who's right or wrong this is about who's right, because there is no room for being wrong.
I'm only for getting the functions of the machine nailed down accurately and that is where I am coming from. If dsm is interested in experimenting with PB and eliminating his while at the same time getting better sleep, I'm all for helping out where I can because it is my belief you can virtually eliminate the conditions of PB and Central Dysregulation with this machine. If you are unable to control those two phenotypes then well you don't have the SV settings set to allow that. My end goals are the same as yours, knowing exactly how this machine responds and what to expect when a change is made.
This machine is finely tuned and working correctly when there are
NO therapy flags present, other indicator is
100% Patient Triggered Breathing while still
having an available BPM mode set up for fail-safe. While this may be near
impossible to achieve, striving for the lowest values is all you can expect.
someday science will catch up to what I'm saying...