Re: CMS 50F Oximeter

Post by SuperFusionX » Thu Jul 03, 2014 5:57 pm

palerider wrote:
leonlimck wrote:Is this the same pulse oximeter
http://www.amazon.com//dp/B00EGL9SC0/
that you are referring to?

Thanks for the great review by the way.
er, no, your link points to the "Acc U Rate® OLED Wrist Color Pulse Oximeter with Continuous Reading"

not the CMS50F also available cheaper from china though I bought the slightly nicer cms50i, it has a clock to timestamp your recordings, and can save multiple recordings between downloads.
??? It says "CMS 50F OLED Wrist Color Pulse Oximeter with Acc U Rate® probe" but it does seem more expensive to the one you found. Any idea if the Acc U Rate probe makes any difference?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: CMS 50F Oximeter

Post by palerider » Fri Jul 04, 2014 5:11 pm

SuperFusionX wrote:
palerider wrote:
leonlimck wrote:Is this the same pulse oximeter
http://www.amazon.com//dp/B00EGL9SC0/
that you are referring to?

??? It says "CMS 50F OLED Wrist Color Pulse Oximeter with Acc U Rate® probe" but it does seem more expensive to the one you found. Any idea if the Acc U Rate probe makes any difference?
ok, my bad, brain fart or something

yeah, that's the same one, it looks like this "acc u rate" company buys the cms stuff from contec and slaps their lablel on, while marking up the price.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Bmayzer
Posts: 2
Joined: Fri Aug 29, 2014 7:17 pm

Re: CMS 50F Oximeter

Post by Bmayzer » Fri Aug 29, 2014 7:38 pm

I just purchased a CMS50F and am having a problem with it. I get it to record, but when I import the file to the software the file is always 2k in size and a few seconds of data. Anyone had this problem and can help me?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: CMS 50F Oximeter

Post by palerider » Fri Aug 29, 2014 7:43 pm

Bmayzer wrote:I just purchased a CMS50F and am having a problem with it. I get it to record, but when I import the file to the software the file is always 2k in size and a few seconds of data. Anyone had this problem and can help me?
blue or white face? what OS?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Bmayzer
Posts: 2
Joined: Fri Aug 29, 2014 7:17 pm

Re: CMS 50F Oximeter

Post by Bmayzer » Sat Aug 30, 2014 5:34 am

Blue face Ver 3.7w

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: CMS 50F Oximeter

Post by palerider » Sat Aug 30, 2014 12:03 pm

Bmayzer wrote:Blue face Ver 3.7w
I think the blue face f is much the same as the i. mine uses spo2assistant v2.4

it creates a .spo2 file. where the actual file is depends on what version of windows (the OS) you're using.

if you can get a copy of the current testing version of sleepyhead out of jedimark, it reads the blue f directly, no other programs needed.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

StephenR
Posts: 172
Joined: Fri Jul 25, 2014 4:43 pm

Re: CMS 50F Oximeter

Post by StephenR » Wed Sep 03, 2014 4:38 pm

trying to use this device..CMS 50D+

downloaded latest software: v2.4.1

The software cant find the device, neither can Sleepyhead

Any suggestions??

Stephen

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: CMS 50F Oximeter

Post by palerider » Wed Sep 03, 2014 4:42 pm

StephenR wrote:trying to use this device..CMS 50D+

downloaded latest software: v2.4.1

The software cant find the device, neither can Sleepyhead

Any suggestions??

Stephen
first, make sure you're using the "usb" cable that came with it.

it's actually not a usb cable, it just looks like one, it's actually a usb to serial interface cable disguised as a plain cable, a regular usb to mini cable will ont work with the 50d+ or 50e. .

second, to talk to either the software, or sleepyhead, the device has to be in the menu mode, not in the normal display mode. ie, hold down the button till it changes to the menu.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

StephenR
Posts: 172
Joined: Fri Jul 25, 2014 4:43 pm

Re: CMS 50F Oximeter

Post by StephenR » Wed Sep 03, 2014 4:48 pm

palerider wrote:
StephenR wrote:trying to use this device..CMS 50D+

downloaded latest software: v2.4.1

The software cant find the device, neither can Sleepyhead

Any suggestions??

Stephen
first, make sure you're using the "usb" cable that came with it.

it's actually not a usb cable, it just looks like one, it's actually a usb to serial interface cable disguised as a plain cable, a regular usb to mini cable will ont work with the 50d+ or 50e. .

second, to talk to either the software, or sleepyhead, the device has to be in the menu mode, not in the normal display mode. ie, hold down the button till it changes to the menu.
I am using supplied cable..
Sleepyhead now sees the device,
CMS software still says no devices available...

Stephen

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: CMS 50F Oximeter

Post by palerider » Wed Sep 03, 2014 5:00 pm

StephenR wrote:
palerider wrote:
StephenR wrote:trying to use this device..CMS 50D+

downloaded latest software: v2.4.1
CMS software still says no devices available...
use the sw that came with it, I don't know that the 2.4.1 works with the older protocol devices like the 50d+

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65326
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: CMS 50F Oximeter

Post by Pugsy » Wed Sep 03, 2014 5:03 pm

StephenR wrote:I am using supplied cable..
Sleepyhead now sees the device,
CMS software still says no devices available...
Check out Papit's hints on using the Pulse ox stuff. 5th post down on this page in my tutorial thread
viewtopic.php?f=1&t=88983&st=0&sk=t&sd=a&start=15

I didn't use my pulse ox enough to have any clue how to help you.

You may need a different version of the CMS software...there were 2 I think for the 50 D Plus...and only one works.

Pulseoxstore has them for free download. That's where I got mine
http://www.pulseoxstore.com/Manuals-Downloads.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

StephenR
Posts: 172
Joined: Fri Jul 25, 2014 4:43 pm

Re: CMS 50F Oximeter

Post by StephenR » Wed Sep 03, 2014 5:37 pm

palerider wrote:
StephenR wrote:
palerider wrote:
StephenR wrote:trying to use this device..CMS 50D+

downloaded latest software: v2.4.1
CMS software still says no devices available...
use the sw that came with it, I don't know that the 2.4.1 works with the older protocol devices like the 50d+
You are correct,

had to use the older software...

That seems to have made everything work, it even works in SleepyHead now..

Happy Apneahead here...

Realize I may have enjoyed a slightly better Pulse Ox, but this will do for now...

Thanks for the hint....

Stephen

StephenR
Posts: 172
Joined: Fri Jul 25, 2014 4:43 pm

Re: CMS 50F Oximeter

Post by StephenR » Wed Sep 03, 2014 5:38 pm

Pugsy wrote:
StephenR wrote:I am using supplied cable..
Sleepyhead now sees the device,
CMS software still says no devices available...
Check out Papit's hints on using the Pulse ox stuff. 5th post down on this page in my tutorial thread
viewtopic.php?f=1&t=88983&st=0&sk=t&sd=a&start=15

I didn't use my pulse ox enough to have any clue how to help you.

You may need a different version of the CMS software...there were 2 I think for the 50 D Plus...and only one works.

Pulseoxstore has them for free download. That's where I got mine
http://www.pulseoxstore.com/Manuals-Downloads.html
all fixed, needed to use the 1.5 version of the software...

Thanks for your help..

Stephen

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: CMS 50F Oximeter

Post by palerider » Wed Sep 03, 2014 5:40 pm

StephenR wrote:
palerider wrote: use the sw that came with it, I don't know that the 2.4.1 works with the older protocol devices like the 50d+
You are correct,

had to use the older software...
...
Thanks for the hint....
glad to help! the newest version of SH integrates the oximeter graphs with the other graphs, it's nice, all stuff on one chart, and if you start the oximeter at the same time as the cpap, you get charts that line up nicely. good stuff.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

George Kent O'Donnal

Re: CMS 501W Oximeter

Post by George Kent O'Donnal » Thu Dec 04, 2014 4:18 pm

Mine Sucks JUst got it two days ago and only the adhesive backups work the rubber tip does not anobody answers calls or email there is no customer support location of any kind good luck on this garbage