Hit and Miss with SmartCard Reader-HELP

Post by atruvirgo » Sun Sep 14, 2008 10:44 am

Does anyone know why it is hit and miss with my SmartCard Reader. Several times when I have gone to plug it in to get a reading on Encore it says Reader not detected. How can this be since it has been detected in the past? How can I fix this glitch? Techy folks out there please help!!
Debbie/Atruvirgo/Armynavyx3
Proud Mom of 1 Navy & 2 Army Kids
Nana to 5 lil darlings!

Life is Grand!!

bap40
Posts: 879
Joined: Wed Feb 13, 2008 8:43 am
Location: North Dakota

Re: Hit and Miss with SmartCard Reader-HELP

Post by bap40 » Sun Sep 14, 2008 11:05 am

I can help if you are running Vista!

After you do the install of the card reader. Under accessories section on the computer go to "run" type in "services msc" Then find the reader and switch from manual to auto start. That fixed my card from disappearing. I haven't lost it since.

Good luck!
Brooke

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Hit and Miss with SmartCard Reader-HELP

Post by Wulfman » Sun Sep 14, 2008 11:15 am

atruvirgo wrote:Does anyone know why it is hit and miss with my SmartCard Reader. Several times when I have gone to plug it in to get a reading on Encore it says Reader not detected. How can this be since it has been detected in the past? How can I fix this glitch? Techy folks out there please help!!
I've had the same thing with the two (identical) computers I have Encore loaded on. Sometimes they work consistently and sometimes they're flaky. I've had to move the plug to several different ports on one of the computers until it was consistent. On the other one, I'm still looking for the "right" USB port.
I've been able to unplug and re-insert the reader plug until it works.....usually the same port.....which makes me wonder if it's the card reader plug, itself. Make sure you are not plugging it into a USB hub and that it's a USB port that's connected to the motherboard......and that the card is inserted into the reader when it's plugged in or the computer started.

Another one of the little frustrations of life.

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Georgio
Posts: 608
Joined: Thu Aug 28, 2008 5:34 pm
Location: Jacksonville Beach, Fl

Re: Hit and Miss with SmartCard Reader-HELP

Post by Georgio » Sun Sep 14, 2008 11:30 am

I found that you want to have the reader plugged into the computer and the card inserted before you start Encore.
DreamStation 2, Oscar
Resmed AirFit P30i Nasal Mask

User avatar
atruvirgo
Posts: 312
Joined: Thu May 17, 2007 8:21 pm
Location: Riverside, California

Re: Hit and Miss with SmartCard Reader-HELP

Post by atruvirgo » Sun Sep 14, 2008 11:32 am

I have 4 USB ports/hubs? on my laptop. None of them are detecting the card reader.
Debbie/Atruvirgo/Armynavyx3
Proud Mom of 1 Navy & 2 Army Kids
Nana to 5 lil darlings!

Life is Grand!!

User avatar
atruvirgo
Posts: 312
Joined: Thu May 17, 2007 8:21 pm
Location: Riverside, California

Re: Hit and Miss with SmartCard Reader-HELP

Post by atruvirgo » Sun Sep 14, 2008 11:34 am

How do i re-install the reader w/o re-installing Encore. Since when I plg in my computer does not give me the message new hardware detected.
Debbie/Atruvirgo/Armynavyx3
Proud Mom of 1 Navy & 2 Army Kids
Nana to 5 lil darlings!

Life is Grand!!

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Hit and Miss with SmartCard Reader-HELP

Post by Wulfman » Sun Sep 14, 2008 11:40 am

atruvirgo wrote:How do i re-install the reader w/o re-installing Encore. Since when I plg in my computer does not give me the message new hardware detected.
It won't hurt to re-install the driver by itself. The driver "Setup.exe" program is on the software CD in the "Smart Card Readers" directory. I would not plug in the reader plug until you run the setup program. Then re-boot the computer and plug it in. If it comes up with a message about not being approved for Windows (or something to that effect), click through it (accept/OK it).
And, do not plug it into a hub.

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

bap40
Posts: 879
Joined: Wed Feb 13, 2008 8:43 am
Location: North Dakota

Re: Hit and Miss with SmartCard Reader-HELP

Post by bap40 » Sun Sep 14, 2008 11:58 am

You will need to do a re-install and use the same USB everytime. Then try the fixes I was given from another board member and maybe that will be the cure. Mine did not recognize the hardware either.
Brooke

marshaeb
Posts: 286
Joined: Wed Jul 23, 2008 8:57 am

Re: Hit and Miss with SmartCard Reader-HELP

Post by marshaeb » Sun Sep 14, 2008 2:24 pm

bap40 wrote:You will need to do a re-install and use the same USB everytime. Then try the fixes I was given from another board member and maybe that will be the cure. Mine did not recognize the hardware either.
Did you intend to provide a link to the fixes you mentioned? Maybe I'm just not getting it, but with your statement alone, I don't understand. Thanks!!

Marsha

Guest

Re: Hit and Miss with SmartCard Reader-HELP

Post by Guest » Sun Sep 14, 2008 3:35 pm

atruvirgo wrote:Does anyone know why it is hit and miss with my SmartCard Reader. Several times when I have gone to plug it in to get a reading on Encore it says Reader not detected. How can this be since it has been detected in the past? How can I fix this glitch? Techy folks out there please help!!
When troubleshooting computer or software problems it is always helpful to know whether it has worked before or is a new install which has never worked.

Also helpful to know the OS, Encore version and equipment being used.

User avatar
atruvirgo
Posts: 312
Joined: Thu May 17, 2007 8:21 pm
Location: Riverside, California

Re: Hit and Miss with SmartCard Reader-HELP

Post by atruvirgo » Sun Sep 14, 2008 7:36 pm

Guest wrote:
atruvirgo wrote:Does anyone know why it is hit and miss with my SmartCard Reader. Several times when I have gone to plug it in to get a reading on Encore it says Reader not detected. How can this be since it has been detected in the past? How can I fix this glitch? Techy folks out there please help!!
When troubleshooting computer or software problems it is always helpful to know whether it has worked before or is a new install which has never worked.

Also helpful to know the OS, Encore version and equipment being used.

I am using Encore PRO Version 1.8.49 I also run windows XP. I have had this installed for about 3 months. It did this once before for no reason but, I am still not sure why. I hear others have had difficulties as well. I don;t know why this is such a pain to use!

Debbie
Debbie/Atruvirgo/Armynavyx3
Proud Mom of 1 Navy & 2 Army Kids
Nana to 5 lil darlings!

Life is Grand!!

User avatar
atruvirgo
Posts: 312
Joined: Thu May 17, 2007 8:21 pm
Location: Riverside, California

Re: Hit and Miss with SmartCard Reader-HELP

Post by atruvirgo » Sun Sep 14, 2008 7:52 pm

Georgio wrote:I found that you want to have the reader plugged into the computer and the card inserted before you start Encore.

Thanks Georgio I have tried it that way also. It is a hit and miss with the card reader. No problems with Encore Pro if I could get the reader to be detected. I am so frustrated.
Debbie/Atruvirgo/Armynavyx3
Proud Mom of 1 Navy & 2 Army Kids
Nana to 5 lil darlings!

Life is Grand!!

User avatar
atruvirgo
Posts: 312
Joined: Thu May 17, 2007 8:21 pm
Location: Riverside, California

Re: Hit and Miss with SmartCard Reader-HELP

Post by atruvirgo » Sun Sep 14, 2008 7:53 pm

bap40 wrote:You will need to do a re-install and use the same USB everytime. Then try the fixes I was given from another board member and maybe that will be the cure. Mine did not recognize the hardware either.
Cany ou give me the link or the fixes that you used? I will try anything.
Debbie/Atruvirgo/Armynavyx3
Proud Mom of 1 Navy & 2 Army Kids
Nana to 5 lil darlings!

Life is Grand!!

User avatar
atruvirgo
Posts: 312
Joined: Thu May 17, 2007 8:21 pm
Location: Riverside, California

Re: Hit and Miss with SmartCard Reader-HELP

Post by atruvirgo » Sun Sep 14, 2008 7:54 pm

Wulfman wrote:
atruvirgo wrote:How do i re-install the reader w/o re-installing Encore. Since when I plg in my computer does not give me the message new hardware detected.
It won't hurt to re-install the driver by itself. The driver "Setup.exe" program is on the software CD in the "Smart Card Readers" directory. I would not plug in the reader plug until you run the setup program. Then re-boot the computer and plug it in. If it comes up with a message about not being approved for Windows (or something to that effect), click through it (accept/OK it).
And, do not plug it into a hub.

Den
Sorry donlt mean to sound dumb but how to I reinstall the driver only?
Debbie/Atruvirgo/Armynavyx3
Proud Mom of 1 Navy & 2 Army Kids
Nana to 5 lil darlings!

Life is Grand!!

User avatar
Georgio
Posts: 608
Joined: Thu Aug 28, 2008 5:34 pm
Location: Jacksonville Beach, Fl

Re: Hit and Miss with SmartCard Reader-HELP

Post by Georgio » Sun Sep 14, 2008 7:58 pm

If you purchased the software, Respironics Customer Service at 1-800-345-6443 helped me with mine.

They told me to ignore the software sent with the reader, and got me straight.