GoodKnight 420G How to change air pressure setting

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
deadrianf
Posts: 3
Joined: Sun Sep 14, 2008 9:52 am

GoodKnight 420G How to change air pressure setting

Post by deadrianf » Sun Sep 14, 2008 9:57 am

Hi All,
I use this machine 2 months now and the pressure is set to 7 (ramp at 4) but the doc said on the last visit it should be chaged to 10 with a ramp at 8.How can I change the setting myself?

thanks

ozij
Posts: 10464
Joined: Fri Mar 18, 2005 11:52 pm

Re: GoodKnight 420G How to change air pressure setting

Post by ozij » Sun Sep 14, 2008 10:07 am

This is for the 420E - I'm sure the instructions are the same for the 420G
Setting the device
To restrict patient access to certain settings on the GoodKnight 420 Evolution, and allow complete access to the settings by the physician or home care provider, a hidden key has been provided on the device. It is located under the letters “Go” in the “GoodKnight” name and detectable to the touch as a hollow.
The parameter values may be changed using the ON/OFF (up) key and Ramp (down) key <snip>

To access the settings, press simultaneously the hidden key and then the Information Access key for one second.
The information key is the one at the very left.

O.
[Edited out a wrong comment about the information key...]

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
Last edited by ozij on Sun Sep 14, 2008 8:43 pm, edited 1 time in total.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

deadrianf
Posts: 3
Joined: Sun Sep 14, 2008 9:52 am

Re: GoodKnight 420G How to change air pressure setting

Post by deadrianf » Sun Sep 14, 2008 1:36 pm

Hi Ozij,
Thanks for help
I will try it and get back to you
Adrian

deadrianf
Posts: 3
Joined: Sun Sep 14, 2008 9:52 am

Re: GoodKnight 420G How to change air pressure setting

Post by deadrianf » Mon Sep 15, 2008 8:04 pm

Hi again,
it work beautifully
thanks again
Adrian

ozij
Posts: 10464
Joined: Fri Mar 18, 2005 11:52 pm

Re: GoodKnight 420G How to change air pressure setting

Post by ozij » Mon Sep 15, 2008 10:00 pm


Thanks for letting us know.
O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023