Can I adjust pressure on a REMstar/pro2 c-flex

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
rom623drh4
Posts: 6
Joined: Thu Jul 24, 2008 11:03 am
Location: Guyandotte, WV
Contact:

Can I adjust pressure on a REMstar/pro2 c-flex

Post by rom623drh4 » Thu Jul 24, 2008 11:15 am

new here and suffocating, Can I adjust pressure on a REMstar/pro2 c-flex I about tore up mask the other night when I felt like I was suffocating and jerked it off my face It takes along time to get back into the Dr. and I can take of this myself if I had guidance. thank you

need more pressure,
suffocating


Guest

Post by Guest » Thu Jul 24, 2008 11:20 am

Yup. Go up to the Red Lightbulb at the top of the main forum screen labeled "CPAP FAQ." Scroll all the way to the bottom of the list and click on "XPAP Manuals". You'll find your manual there.

Cheers,
Babs

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Thu Jul 24, 2008 11:21 am

I don't know why not......that's also my machine of choice.

You didn't say what your existing pressure is or how long the Ramp is set to (if you're using it). A Ramp setting that starts too low or is set for too long of a period can cause that feeling.

You might also want to fill out your profile and give us more details on whatever else you can.

Welcome to the forum.

Den




REMstar Pro 2 Setup

When in the Setup Menu, the humidifier ^ and ramp v buttons operate as up and down keys to change the settings, the left/right user buttons < > allow you to go to the previous/next question or setting, and the pressure start/stop button is used to exit the Setup Menu. Holding the humidifier or ramp buttons down will cause the values to change more quickly.

To enter the Therapy Setup Menu, hold the two top user buttons < > down while plugging in the power cord.
Continue holding the buttons down until the REMstar Pro 2 with C-Flex beeps twice. (about 5 seconds)

Note: The word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu.
If you press the Pressure start/stop button, you will exit the Setup Menu.

a. Compliance hours: (recommend leaving alone)
Select next setting with >

b. Therapy mode: (CPAP or CFLE) select with ^ or v
Select next setting with >

c. CPAP pressure setting: Select with ^ or v
Select next setting with >

d. C-Flex mode (if you chose CFLE mode): Select setting 1, 2 or 3 with ^
Select next setting with >

e: Ramp time setting: Select with ^ or v
(ramp time will be turned off with a setting of 0)
Select next setting with >

f. Ramp pressure setting: Select with ^ or v
Select next setting with >

g. Patient disconnect setting: 1 = on 0 = off Select with ^ or v
Select next setting with >

h. Buttons lights setting: 1 = on 0 = off Select with ^ or v
(Last setting. Use On/Off button to exit Setup)

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

rom623drh4
Posts: 6
Joined: Thu Jul 24, 2008 11:03 am
Location: Guyandotte, WV
Contact:

suffocating

Post by rom623drh4 » Thu Jul 24, 2008 1:43 pm

thanks for the help, just got home read the replies and have the machine re-set. thanks again LORD BLESS

MikDee

Post by MikDee » Mon Jul 28, 2008 9:10 pm

I'm a newbie myself, and had this same problem! After 3 new machines I thought were defective? and recently messing with the pressure, to no avail, I got an idea to turn off the C-Flex function, and It Worked For Me! and I'm back at my starting pressure! It's been a full week now of restful sleep! i have a theory about why this works, but no time to describe it right now. Give this a try.


bryman34
Posts: 5
Joined: Sat Mar 08, 2008 11:23 am
Location: COLORADO

Post by bryman34 » Tue Jul 29, 2008 7:07 am

Anonymous wrote:Yup. Go up to the Red Lightbulb at the top of the main forum screen labeled "CPAP FAQ." Scroll all the way to the bottom of the list and click on "XPAP Manuals". You'll find your manual there.

Cheers,
Babs
lol thats actually a red ? mark not a light bulb. lol just had to give you a hard time