Sandman Info and Sandman Auto

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
ozij
Posts: 10527
Joined: Fri Mar 18, 2005 11:52 pm

Sandman Info and Sandman Auto

Post by ozij » Wed Jun 11, 2008 3:33 am

http://investor.covidien.com/preview/ph ... id=1162848
The Sandman Info CPAP builds on the feature-rich Sandman Intro CPAP. The Sandman Info device detects and records respiratory events while the patient is sleeping. This data can easily be downloaded to a memory card, allowing sleep clinicians and home care providers to review and provide indications of patient benefits from use of the CPAP device. The premium Sandman Auto CPAP has the Sandman Info's capabilities, plus it optimizes pressure by increasing it during obstructive events and decreasing pressure once the event has passed (while still recording all critical data).

Covidien is uniquely capable of offering comprehensive solutions to better enhance diagnosis and treatment of OSA patients. Via the global "Unite to treat Sleep Apnea" campaign, Covidien is uniting patients, physicians, Sleep Labs and Home Medical Equipment providers in an effort to simplify diagnosis and improve patient compliance with therapy. Covidien's new trio of CPAP devices is united with Covidien's sleep diagnostic system under the Sandman line.

http://www.puritanbennett.com/prod/List.aspx?S1=SPT

510(k) SUMMARY
Predicate Device Equivalence

The Puritan Bennett Sandman Info is equivalent to the Puritan Bennett GoodKnight 420S CPAP device (K020886), with the optional additive functionality of the Puritan Bennett GoodKnight H20 heated humidifier (K042184).

The Puritan Bennett Sandman Auto is equivalent to the Puritan Bennett GoodKnight 420E CPAP device (K020886), with the optional additive functionality of the Puritan Bennett GoodKnight H20 heated humidifier (K042184).

<snip>
Device Description
The Puritan Bennett Sandman Info operates only in Constant mode: the main function of the device is to deliver constant positive airway pressire to the patient at a fixed level prescribed by the practitioner.

The Puritan Bennett Sandman Auto operates in either Constant or Automatic mode. In Constant mode, the main function of each device is to delier constant positive airway pressure to the patient at a fixed level prescribed by the practitioner and between 4 and 20 cmH2O.

In Automatic mode (APAP mode), the practitioner determines and sets a maximum and minimum pressure range above and below the prescribed reference pressure and between 4 and 20 cmH20. The pressure is adjusted within the maximum and minimum limits according to the patient's respiratory pattern and the type of events detected.


O.
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): auto cpap, CPAP

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
DreamStalker
Posts: 7509
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Post by DreamStalker » Wed Jun 11, 2008 6:57 am

I thought they were supposed to add exhalation relief to their new model?

Nice looking machine ... looks more like something you would expect to see on a night stand. I guess they are trying to compete with ResMed styling appeal.

President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Post by -SWS » Wed Jun 11, 2008 7:09 am

In Automatic mode (APAP mode), the practitioner determines and sets a maximum and minimum pressure range above and below the prescribed reference pressure and between 4 and 20 cmH20. The pressure is adjusted within the maximum and minimum limits according to the patient's respiratory pattern and the type of events detected.
Thanks for the information, ozij! I wonder if Tyco will once again neglect to mention that key role of "reference pressure" in the clinician's set-up manual: that initial pressure should be set at the prescribed pressure (specifically a prescribed pressure that eliminates flow limitations). The medical community and patients should not have to dig around the patent archives for that essential information.

Another question that comes to mind is whether any of Rapoport's more recent patent refinements will go into that algorithm. For instance, will the prolific scoring or aggressive pressure response of FL change in any way?

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Post by -SWS » Wed Jun 11, 2008 7:14 am

DreamStalker wrote:I thought they were supposed to add exhalation relief to their new model?

Nice looking machine ... looks more like something you would expect to see on a night stand. I guess they are trying to compete with ResMed styling appeal.
Think they dropped that feature, once the user community managed to ace it with tape calibrations on the RJ11 jack...

.


User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Post by Slinky » Wed Jun 11, 2008 7:54 am

BIG THANKS, Ozij. I appreciate reading this type of info about the various xPAP models!!!

Especially, in this case, since I've been used to the Purtian Bennett name rather than the Sandman name. Puritan Bennett is more familiar so inspired more confidence than an unfamiliar (to me) name such as Covidien.

I was familiar w/the name of Sandman as one of the professional software used for sleep evaluation and titration.


_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

ozij
Posts: 10527
Joined: Fri Mar 18, 2005 11:52 pm

Post by ozij » Wed Jun 11, 2008 10:30 am

-SWS wrote:
In Automatic mode (APAP mode), the practitioner determines and sets a maximum and minimum pressure range above and below the prescribed reference pressure and between 4 and 20 cmH20. The pressure is adjusted within the maximum and minimum limits according to the patient's respiratory pattern and the type of events detected.
Thanks for the information, ozij! I wonder if Tyco will once again neglect to mention that key role of "reference pressure" in the clinician's set-up manual: that initial pressure should be set at the prescribed pressure (specifically a prescribed pressure that eliminates flow limitations). The medical community and patients should not have to dig around the patent archives for that essential information.

Another question that comes to mind is whether any of Rapoport's more recent patent refinements will go into that algorithm. For instance, will the prolific scoring or aggressive pressure response of FL change in any way?
Good questions, both....
Do you realize the clinicain's manual for the 420E even had a graph showing the initial pressure falling between the min. and the max? (page 19).
The problem with using the PB properly configured is that the prescribing doctors don't know the machines, and the people who set them up are not allowed to change the doctor's Rx. If a doctor prescribes a PB auto, without an initial pressure - what's the DME to do - assuming her or she knows how the machine should be set up, but the doc thinks that this info is none of a doctor's business ?

O.


_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

-SWS
Posts: 5301
Joined: Tue Jan 11, 2005 7:06 pm

Post by -SWS » Wed Jun 11, 2008 11:39 am

ozij wrote:Do you realize the clinicain's manual for the 420E even had a graph showing the initial pressure falling between the min. and the max? (page 19).
Yeah, I just took a peek at that graph on page 19. It shows that the 420e ramp function will gradually rise to either initial pressure or to CPAP pressure---depending on which modality is used.

So for good measure I text-searched the 420e clinician's manual for the word "initial". I wanted to see if PB/Tyco gave clinicians even the slightest hint that "initial pressure" should be set at a "reference pressure" or "prescribed pressure" that adequately eliminates flow limitations.

NOT A HINT IN THE MANUAL FOR CLINICIANS OR 420e APAP USERS!!!!




That omission of crucial set-up information in the clinician's manual is a HUGE problem IMHO. And I'm still amazed about it. But very glad you discovered it a few weeks back!
ozij wrote:The problem with using the PB properly configured is that the prescribing doctors don't know the machines, and the people who set them up are not allowed to change the doctor's Rx. If a doctor prescribes a PB auto, without an initial pressure - what's the DME to do - assuming her or she knows how the machine should be set up, but the doc thinks that this info is none of a doctor's business ?
Absolutely! And that's assuming the doctors, clinicians, and DME providers somehow manage to discover the well-hidden secret of how "initial pressure" needs to be set! What a procedural mess!

In the meantime cpaptalk.com members now have a 420e "leg-up" on the rest of the world!


ozij
Posts: 10527
Joined: Fri Mar 18, 2005 11:52 pm

Post by ozij » Wed Jun 11, 2008 9:33 pm

Sandman Therapy Software Clinical Kit 1.0

That's refreshing - having the software there before the machine... .

Did anyone say DeVilbiss Intellipap?

And by the way (from the first quote, above, my emphasis):
BOULDER, Colo.--(BUSINESS WIRE)--June 5, 2008--Covidien (NYSE: COV, BSX: COV), a leading global healthcare products company, today announced that two new continuous positive airway pressure (CPAP) devices, the Sandman Info(TM) and the Sandman Auto(TM), are now available in the United States
O.


_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Post by Slinky » Wed Jun 11, 2008 9:45 pm

I know the Sandman software is a popular software used by sleep clinics for their titration studies. I just didn't realize it was a Puritan Bennett software.


_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.