Locating acceptable mask leak rates

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
DanoJ
Posts: 8
Joined: Fri Jun 06, 2008 9:08 am
Location: Florida

Locating acceptable mask leak rates

Post by DanoJ » Sat Jun 07, 2008 11:27 am

This is a wonderful forum, particularly for those of us who are new to apnea and treatment. I swear the people on this forum seem to know more about this condition than my sleep doctor does.

Brief history: My polysomnography results reflected that I had a respiratory disturbance index of 104.3/hr. of sleep, no stage REM, 58% stage 1, 12% stage 2, and 30% stage awake. My min SaO2 was 73%. Longest apneic event was 28 seconds.

I have now had my BiPAP Auto w/ BiFlex for a week, and am using the Mirage Swift II nasal pillows. My Encore Viewer results say that my AHI is averaging between 3.7 and 5.1 with an overall average of 4.9 (but the average is dropping almost daily for some reason). This is one hell of an improvement for me!

The problem with getting my AHI lower seems to be mask leaks, if I understand all the posts on the subject. I have no large leaks, but my average 90% leak (I don't know what that means) is 45.4 and average leak is 40.2. Nowhere can I find anything that tells me what my leak numbers should be for this nasal pillow system. Also, I figured that with my total time in apnea dropped from approx. 150 minutes per night (pre-treatment) to just over 4 minutes per night, the leaks cannot be a serious issue. Am I correct, or is there still a problem with my leaks that can improve my numbers? Where can I find out what my leak numbers should be?


User avatar
GumbyCT
Posts: 5780
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Post by GumbyCT » Sat Jun 07, 2008 11:40 am

The official leak rate for each mask can be found in a little brochure that comes with the mask. There should be a hard to see graph hidden in there somewhere.

You can also check the manuf. web site or check here for the unofficial rates.

Mask Leak Rates
http://www.internetage.com/cpapinfo/leak-rates-1.html

Keep in mind the leak increases with the pressure. You should be close but I am never exact.

Good Luck,
GumbyCT

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

DanoJ
Posts: 8
Joined: Fri Jun 06, 2008 9:08 am
Location: Florida

Thanks!

Post by DanoJ » Sat Jun 07, 2008 11:45 am

GumbyCT wrote:The official leak rate for each mask can be found in a little brochure that comes with the mask. There should be a hard to see graph hidden in there somewhere.

You can also check the manuf. web site or check here for the unofficial rates.

Mask Leak Rates
http://www.internetage.com/cpapinfo/leak-rates-1.html

Keep in mind the leak increases with the pressure. You should be close but I am never exact.

Good Luck,
GumbyCT

Thanks, Gumby. That was fantastic information. The link you gave me does not include info on the Swift II, but for the Swift I at my max pressure of 14, it says leak rate should be 40.2, exactly what my average leak is! If the Swift II is about the same, something else must be causing my numbers to remain a little high...true?


ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Post by ozij » Sat Jun 07, 2008 12:07 pm

If you search Resmed's site you'll eventually find a file saying the Swift II and the Swift have the same leak rate.

You're doing great.

Congratulations!

O.


_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
GumbyCT
Posts: 5780
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Post by GumbyCT » Sat Jun 07, 2008 12:40 pm

If you are talking ahi numbers yea. It could be the pressure is too high or not high enuff. Could even be sleep position or the type of mask.

Do you have the software? oh It says you do.

How are you feeling? That is most important.


_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!