I am considering buying one for myself, however I have a question for the experienced users out there.
I read or remember that the 420E measures pressure at the mask and has an additional tube to do this. Does this mean you have to use the 420e 6' tubing? Does it also require a coupling at the mask? I can't quite picture how this works. Are certain masks a problem and won't work?

Curtis
curtcurt46

Bearded_One
Posts: 597
Joined: Fri Nov 17, 2006 9:35 am
Location: Northern Virginia, near DC

Post by Bearded_One » Sun Apr 27, 2008 4:32 pm

The connector on the end of the hose that attaches to the CPAP machine has a small hole punched in it and a piece of tubing is passed through the hole. A short piece of the sensor tubing hanging outside of the CPAP hose and is connected to a nipple on th CPAP machine; the other end of the tubing runs inside the hose to the mask end. The tubing ends and the pressure is measured at the input of the mask. cpap.com sells an 8 foot hose for the 420E and you could fabricate a longer hose; the tubing is standard oxygen tubing.

Image


User avatar
NightHawkeye
Posts: 2431
Joined: Thu Dec 29, 2005 11:55 am
Location: Iowa - The Hawkeye State

Re: PB 420E

Post by NightHawkeye » Sun Apr 27, 2008 4:48 pm

curtcurt46 wrote:I am considering buying one for myself, however I have a question for the experienced users out there.
I read or remember that the 420E measures pressure at the mask and has an additional tube to do this. Does this mean you have to use the 420e 6' tubing? Does it also require a coupling at the mask? I can't quite picture how this works. Are certain masks a problem and won't work?
Nearly all masks work just fine with the GK420E. However, any mask which comes with its own hose, like the ComfortCurve, will need special consideration. The ComfortCurve comes with an extra lightweight hose and uses a special connector at the mask which, while not entirely compatible, nevertheless can still be press-fit into the 420E hose. At least that works satisfactorily for me.

Bill


User avatar
curtcurt46
Posts: 262
Joined: Wed Sep 27, 2006 12:35 pm
Location: Retired US Army

Post by curtcurt46 » Sun Apr 27, 2008 6:20 pm

Thanks for the replies. I believe I can picture it better. Many Thanks.
Curtis
curtcurt46

User avatar
Snoredog
Posts: 6397
Joined: Sun Mar 19, 2006 5:09 pm

Post by Snoredog » Sun Apr 27, 2008 11:15 pm

the 1/8" ID separate hose enter the center of the main hose by coming in the side of the rubber/silicone end on the regular CPAP hose, it then runs down the center to the opposite end.

While you can adjust this hose it doesn't need to be extended into your mask for it to function, it simply measures the air pressure at the mask for a more accurate reading.

That hose connects to the pressure transducer on the inside of the machine. Most other mfgs just run the hose from the transducer to the main air outlet and compensate for the hose length internally.

but you can use it with any mask without anything special even the leader hoses some masks come with, nothing special is needed to use those masks.

someday science will catch up to what I'm saying...

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Post by dsm » Mon Apr 28, 2008 4:00 am

Here are some pics - go to panel 3 using the nav arrows & click on the last 3 pics to enlarge them & see the hose ends.

Cheers

DSM

http://www.internetage.com/cpapinfo/pb4 ... index.html

xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

User avatar
barb_z
Posts: 53
Joined: Sat Sep 29, 2007 5:27 pm
Location: Treasure Coast of Florida

Re: the inside!

Post by barb_z » Mon Apr 28, 2008 5:04 am

Wow... I always love seeing how things look inside! Thanks...
Barb

_________________
Mask: Oracle HC452 Oral CPAP Mask
Peace begins with each of us, as individuals

User avatar
tangents
Posts: 750
Joined: Thu Mar 08, 2007 11:03 am
Location: Cleveland, Ohio

Post by tangents » Mon Apr 28, 2008 8:49 am

Good decision, if you decide to buy one. I really like mine. Small, quiet, great software.

Cheers,
Cathy