Hyperinflation of lungs
-
sam thomas
Hyperinflation of lungs
Who can we ask medical questions at cpaptalk.com. For example
Will high pressure cpap cause lung hyperinflation?
Sam Thomas st2166@sccoast.net
843 756 1055
Will high pressure cpap cause lung hyperinflation?
Sam Thomas st2166@sccoast.net
843 756 1055
I would guess SAG or Dr Barry Krakow.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.
My computer says I need to upgrade my brain to be compatible with its new software.
- BarryKaraoke
- Posts: 176
- Joined: Thu Mar 06, 2008 8:03 pm
- Location: Patterson, NY
A CPAP machine does not (and will not) "inflate" your lungs...it will hardly inflate a beach ball.
The actual PSI (pounds per square inch) of a CPAP machine is very, very low and not to be confused with your pressure setting (centimeters of water). They are two very different things.
The highest pressure of most CPAP machines (20) is actually only 0.28 psi. Not very good for inflating things!
The actual PSI (pounds per square inch) of a CPAP machine is very, very low and not to be confused with your pressure setting (centimeters of water). They are two very different things.
The highest pressure of most CPAP machines (20) is actually only 0.28 psi. Not very good for inflating things!
SAG is the member called StillAnotheGuestSlinky wrote:I would guess SAG or Dr Barry Krakow.
Dr. Barry Krakow is the memeber called BarryKrakowMD. Not to be confused with the member who took on a similar name.
O.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
-
DME Daddy
-
split_city
- Posts: 465
- Joined: Mon Apr 23, 2007 2:46 am
- Location: Adelaide, Australia
I would have to disagree with this. CPAP does increase lung volume quite effectively. Heinzer et al (2006) demonstrated that 13cmH20 CPAP increased lung volume by ~770mL which is greater than an average tidal breath.BarryKaraoke wrote:A CPAP machine does not (and will not) "inflate" your lungs...it will hardly inflate a beach ball.
The actual PSI (pounds per square inch) of a CPAP machine is very, very low and not to be confused with your pressure setting (centimeters of water). They are two very different things.
The highest pressure of most CPAP machines (20) is actually only 0.28 psi. Not very good for inflating things!
Furthermore, O'Donoghue et al (2002) showed that lung volume increased by ~1.1L following the application of 10cmH20 in a group of COPD patients.
The compliance of the lungs is relatively high above resting lung volume i.e. doesn't take much pressure to increase lung volume.
Increasing lung volume helps to prevent upper airway collapse. The increase in lung volume brought upon by CPAP, is one reason why the airway remains patent. This is in addition to the direct splinting effect that CPAP has on the airway.
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): cpap machine, CPAP
-
DME Daddy
- BarryKaraoke
- Posts: 176
- Joined: Thu Mar 06, 2008 8:03 pm
- Location: Patterson, NY
Of course it increases lung volume and splints open the airway...it wouldn't be effective treatment if it didn't do this.split_city wrote:I would have to disagree with this. CPAP does increase lung volume quite effectively.
The point, is that a CPAP machine is not going to somehow "over inflate" your lungs to the point of damage/rupture, as the OP fears.
split_city wrote:I would have to disagree with this. CPAP does increase lung volume quite effectively. Heinzer et al (2006) demonstrated that 13cmH20 CPAP increased lung volume by ~770mL which is greater than an average tidal breath.BarryKaraoke wrote:A CPAP machine does not (and will not) "inflate" your lungs...it will hardly inflate a beach ball.
The actual PSI (pounds per square inch) of a CPAP machine is very, very low and not to be confused with your pressure setting (centimeters of water). They are two very different things.
The highest pressure of most CPAP machines (20) is actually only 0.28 psi. Not very good for inflating things!
Furthermore, O'Donoghue et al (2002) showed that lung volume increased by ~1.1L following the application of 10cmH20 in a group of COPD patients.
The compliance of the lungs is relatively high above resting lung volume i.e. doesn't take much pressure to increase lung volume.
Increasing lung volume helps to prevent upper airway collapse. The increase in lung volume brought upon by CPAP, is one reason why the airway remains patent. This is in addition to the direct splinting effect that CPAP has on the airway.
someday science will catch up to what I'm saying...
- BarryKaraoke
- Posts: 176
- Joined: Thu Mar 06, 2008 8:03 pm
- Location: Patterson, NY
You missed my point. The original poster fears that HIS LUNGS WILL OVER INFLATE by using a CPAP machine. I was merely pointing out that even 20 CENTIMETERS OF WATER is still a very, very low pressure. This isn't my opinion...it is a measurable fact. My reference to PSI was used as a COMPARISON as the average person does not encounter this unit of measure in there day to day lives, yet most know and understand PSI.Snoredog wrote:And I disagree with Barry's comment on pressure, it IS centimeters in water as measured with a manometer. PSI is NOT used for any measurement with these machines what so ever.
There is an interesting article on the risks of "Pneumothorax" using CPAP HERE.
From the article:
The bottom line...fear of possible lung hyper-inflation should not keep someone from continuing CPAP therapy. Agreed?"So the real question at hand is can CPAP therapy cause a pneumothorax? The aforementioned message poster which claimed this was a potentially very serious consequence of patients adjusting their own pressure settings, believed firmly that unsupervised CPAP therapy can cause a pneumothorax. After the little bit of research I've conducted on the topic it seems to me that just about anything can cause a pneumothorax, including genetic defects, deep sea diving, high altitude flying, getting hit in the chest with a baseball bat, and just sitting idly in a chair (the latter especially if you have some sort of lung disease). In fact, from what I've read, most cases of spontaneous pneumothorax don't occur during physical exertion.
If your lung is going to rupture from CPAP therapy, I'd guess it would probably rupture all by itself, even without CPAP therapy."
Last edited by BarryKaraoke on Sat Mar 29, 2008 12:27 pm, edited 1 time in total.
BarryKaraoke wrote:That was my point. The original poster fears that HIS LUNGS WILL OVER INFLATE by using a CPAP machine. I was pointing out that even 20 CENTIMETERS OF WATER is a very, very low pressure. PSI was used as a COMPARISON as the average person has no normal application for this unit of measure, yet most understand PSI.Snoredog wrote:And I disagree with Barry's comment on pressure, it IS centimeters in water as measured with a manometer. PSI is NOT used for any measurement with these machines what so ever.
SnoreDog & Split_City: Are you both then answering the original poster with a YES? That there IS a risk of "Lung Hyperinflation" with CPAP? Or merely arguing over semantics?
someday science will catch up to what I'm saying...
- BarryKaraoke
- Posts: 176
- Joined: Thu Mar 06, 2008 8:03 pm
- Location: Patterson, NY
Exactly. That was all I was getting at. A fear of lung "hyper inflation" should not keep you from using your CPAP machine.Snoredog wrote:Considering the length of time CPAP therapy has been implemented (more than 20 years) hyperinflation of the lungs does not appear to be a problem, so the answer to the original poster would be "No".