Setting pressure on Remstar
Setting pressure on Remstar
Would someone please tell me how to set the pressure on a Remstar Pro. I am away from home for an extended period and need to reset my pressure. Need your help!!!!!!!!!!!!! THANKS
Hi, I'm pretty new to this forum, but I just got a REMstar Pro w/c-flex yesterday and so have the instructions handy.
I'll type the instructions from the Setup Instructions:
"To enter the Therapy Setup menu, hold the two top user buttons down while plugging in the power cord. Continue holding the buttons down until the REMstar Pro beeps. Note: the word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu. If you press the pressure start/stop button, you will exit the Setup Menu."
[the system set up goes through several other set up actions previous to setting the pressure and the third action is setting the CPAP pressure. (range 4-20 cm in 0.2 cm increments)] <-----this is my comment.
"Set the Pressure: To change the pressure setting, press the ramp or humidifier button until the correct pressure appears. Press the right user button to go the next setting".
[In the setup menu there appears to be 8 items you can set (labeled a through h) each item is reached consecutively by pressing the right user button to go to the next setting] <------this is my comment
I hope this has helped. If there is specific info I can provide from the booklet, please reply and I'll try to provide it. Good Luck.
Connie
I'll type the instructions from the Setup Instructions:
"To enter the Therapy Setup menu, hold the two top user buttons down while plugging in the power cord. Continue holding the buttons down until the REMstar Pro beeps. Note: the word "setup" will appear on all of the screens indicating that you are in the Therapy Setup Menu. If you press the pressure start/stop button, you will exit the Setup Menu."
[the system set up goes through several other set up actions previous to setting the pressure and the third action is setting the CPAP pressure. (range 4-20 cm in 0.2 cm increments)] <-----this is my comment.
"Set the Pressure: To change the pressure setting, press the ramp or humidifier button until the correct pressure appears. Press the right user button to go the next setting".
[In the setup menu there appears to be 8 items you can set (labeled a through h) each item is reached consecutively by pressing the right user button to go to the next setting] <------this is my comment
I hope this has helped. If there is specific info I can provide from the booklet, please reply and I'll try to provide it. Good Luck.
Connie