Filter replacement frequency

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Bookbear
Posts: 1154
Joined: Tue Apr 25, 2006 7:47 pm
Contact:

Filter replacement frequency

Post by Bookbear » Tue Nov 27, 2007 9:55 pm

I have a buddy who uses the basic RemStar cpap. He was not even given a patient manual. How often does Respironics suggest that filters be replaced? Her is having trouble with Apria saying they will replace filters every 6 months along with his mask "if needed". HE's on line now, so any quick responses would be appreciated!

Thanks!


_________________
MachineMask
Additional Comments: Avg. AHI .4
Getting old doesn't make you 'forgetful'. Having too damn many things to remember makes you 'forgetful'.

JZ
Posts: 360
Joined: Mon Feb 05, 2007 9:25 pm
Location: Texas

Post by JZ » Tue Nov 27, 2007 10:03 pm

Bookbear,

My Respironics manual says the following:

Under normal usage, you should clean the gray foam filter at least once every two weeks and replace it with a new one every six months. The white ultra-fine filter is disposable and should be replaced after 30 nights of use or sooner if it appears dirty. DO NOT clean the ultra-fine filter.

Janna


User avatar
Bookbear
Posts: 1154
Joined: Tue Apr 25, 2006 7:47 pm
Contact:

Post by Bookbear » Tue Nov 27, 2007 10:08 pm

Thank you!

He's suitably impressed with the prompt reply!

Thanks!

_________________
MachineMask
Additional Comments: Avg. AHI .4
Getting old doesn't make you 'forgetful'. Having too damn many things to remember makes you 'forgetful'.

User avatar
Wulfman
Posts: 12316
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Wed Nov 28, 2007 7:04 am

The user manuals can be found at this link:

http://global.respironics.com/manualsandliterature.asp

.....In case your friend wants to download the one for his machine.


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Bookbear
Posts: 1154
Joined: Tue Apr 25, 2006 7:47 pm
Contact:

Post by Bookbear » Wed Nov 28, 2007 10:26 am

Thanks, Den. I've passed the link on to him.

I have the feeling he's going to join the forum. He gets zero support from his doc or dme. I mean, really... one filter per every six months????


_________________
MachineMask
Additional Comments: Avg. AHI .4
Getting old doesn't make you 'forgetful'. Having too damn many things to remember makes you 'forgetful'.

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Post by ozij » Wed Nov 28, 2007 10:42 am

One gray one - the external, washable - every six months is what Resmed recommends. It's only the fine, non-washable filters that have to be changed more frequently.

O.


_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Post by Slinky » Wed Nov 28, 2007 11:29 am

The only filter my Resmed S8 Elite manual refers to it says to replace every six months. Uh, uh. Not in MY house! After 20 years of a furnace add-on woodburner sending residue thru our furnace ducts during the winter I'm having to change my filter close to once a month! I relied on the six month recommendation for my first filter change. BIG MISTAKE! There was even "dust" in and on the frame around the filter and I won't even TRY to describe what the filter looked like!!!


_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Post by ozij » Wed Nov 28, 2007 11:42 am

I don't know about Resmed, but Puritan Bennett is very specific that the external filter should washed once a week, and IIRC so is Respironics.

In any case, nobody should base their filter changing decisions on the insurance policy.

I try to remember to wash the external filter weekly - and I have to change the internal one every 10-14 days.

O.


_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

FreeLancer74
Posts: 92
Joined: Wed Nov 28, 2007 9:50 am
Location: Texas

Post by FreeLancer74 » Wed Nov 28, 2007 11:55 am

Remember that filters are not that expensive. I hope that the "insurance" factor is not what you are basing your filter changing cycle on, as I can get a 6-pack of white fine filters for the Respironics M-series for like $14. The gray foam filter 2-pack is also $14. So a whole years worth of filters for $56. If you use a flex-plan through work, you should be able to submit that for reimbursement or use your Benny card.

Just keep in mind that the filters keep the air clean that is being forced into your body by your xPAP. You want to make sure as many irritants are cleared as possible and that dirty filters can make you sick or reduce your xPAPs ability to flow and may make it wear out more quickly.

Thank You,
FreeL

"He hates these cans. STAY AWAY FROM THE CANS!"

"DIE GAS PUMPER!"

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Resmed filters are different:

Post by ozij » Wed Nov 28, 2007 12:09 pm

And it is different for Resmed:

http://www.resmed.com/en-au/products/se ... ide_en.pdf
REPLACING THE AIR FILTER
To replace the air filter:
1. Remove the air filter cover at the back of the S8 AUTOSET SPIRIT.
2. Remove and discard the old air filter.
3. Insert a new filter with the blue tinted side facing out from the device.
4. Replace the air filter cover.
Inspect the air filter every month to check if it is blocked by dirt or contains holes. With
normal use of an S8 AUTOSET SPIRIT, the air filter needs to be replaced every six months
(or more often if your device is in a dusty environment).
!
WARNING
• Do not wash the air filter. The air filter is not washable or reusable.
• The air filter cover protects the device in the event of accidental liquid spillage
onto the device. Ensure the air filter and air filter cover are fitted at all times
My emphasis.
O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Post by Slinky » Wed Nov 28, 2007 12:17 pm

Duh, Ozij!!! Trying to make me look bad?? I must have missed that "check every month" when I went thru the manual. Color me red-faced. I sure do check it every month now! And replace it too. My poor little ole beloved S8. I feel so guilty about abusing it that way. It isn't just the woodburner dust, we have long haired dogs and cats too. Heck, I even name my dust bunnies!!!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Post by ozij » Wed Nov 28, 2007 12:28 pm

Slinky wrote:Duh, Ozij!!! Trying to make me look bad?? I must have missed that "check every month" when I went thru the manual. Color me red-faced. I sure do check it every month now! And replace it too. My poor little ole beloved S8. I feel so guilty about abusing it that way. It isn't just the woodburner dust, we have long haired dogs and cats too. Heck, I even name my dust bunnies!!!

Oh well, Slinky, we can always say that when you first went through the manual you were still suffering from the effects of sleep apnea....


O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Post by Slinky » Wed Nov 28, 2007 12:35 pm

Yeah, Ozij! That's it! I was. Thanks for helping me save face. That's my story and I am sticking to it!!!!

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.

User avatar
Bookbear
Posts: 1154
Joined: Tue Apr 25, 2006 7:47 pm
Contact:

Re: Resmed filters are different:

Post by Bookbear » Wed Nov 28, 2007 2:59 pm

ozij wrote:And it is different for Resmed:
REPLACING THE AIR FILTER
To replace the air filter:
1. Remove the air filter cover at the back of the S8 AUTOSET SPIRIT.
2. Remove and discard the old air filter.
3. Insert a new filter with the blue tinted side facing out from the device.
4. Replace the air filter cover.
Inspect the air filter every month to check if it is blocked by dirt or contains holes. With
normal use of an S8 AUTOSET SPIRIT, the air filter needs to be replaced every six months
(or more often if your device is in a dusty environment).
!
WARNING
• Do not wash the air filter. The air filter is not washable or reusable.
• The air filter cover protects the device in the event of accidental liquid spillage
onto the device. Ensure the air filter and air filter cover are fitted at all times
The wording in my Vantage manual is the same (inspect monthly, change at 6 months, more often if needed).

My buddy's machine is a Reprironics (basic tank). A monthly change for any machine makes sense to me, cheap insurance that your machine (and lungs) will continue to function optimally.

_________________
MachineMask
Additional Comments: Avg. AHI .4
Getting old doesn't make you 'forgetful'. Having too damn many things to remember makes you 'forgetful'.

User avatar
Slinky
Posts: 11372
Joined: Wed Nov 01, 2006 3:43 pm
Location: Mid-Michigan

Post by Slinky » Wed Nov 28, 2007 4:21 pm

Go ahead, BookBear, rub it in! 20 lashes w/a wet noodle for you my bookish friend!