How can a person set the pressure on the Goodknight 420e without using the software? I just bought one from cpap auction and it didn't include the user manual.
Is there a site to download the manual?
How to adjust a Goodknight 420e and user manual
-
Bearded_One
- Posts: 597
- Joined: Fri Nov 17, 2006 9:35 am
- Location: Northern Virginia, near DC
Puritan Bennett has a relatively new knowledgebase site here:
http://solvitcenter.puritanbennett.com/afmmain.aspx
And you can find a PDF file of the users manual for the machine - though it won't teach you about setting it up.
O.
http://solvitcenter.puritanbennett.com/afmmain.aspx
And you can find a PDF file of the users manual for the machine - though it won't teach you about setting it up.
O.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
-
Bearded_One
- Posts: 597
- Joined: Fri Nov 17, 2006 9:35 am
- Location: Northern Virginia, near DC
One caveat about the 420E; you have to use a special hose that has a pressure sensing tube. The pressure sensing tube is about the same size as aquarium tubing and slips over a small nipple near the output. I understand that there is an adapter with a sensor tube that allows the use of a standard hose.
- rested gal
- Posts: 12880
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
To access the clinical menu on the Puritan Bennett 420E, press down on the letters "Go" of the word "GoodKnight" on top of your machine. While keeping that little depressed area under "Go" pushed down, press the "Information Access" button on the top of the machine (the button with the diagonal line and little hooks at each end.)
Keep both pressed down together for a few seconds then release them. You're now in the clinical menu.
Use the Info Access button to go to the next item.
Use the on/off button and ramp button (like up/down buttons) to change settings.
The last item in the menu will be "ID." When you press the Info Access button one more time, that will exit you from the menu. Any settings you changed will be automatically saved.
Note -- When going through the menu, the next-to-last item will be the "pressure sensor offset" setting. Do not change that. That item will look like this, although the setting for your machine will probably be a different number:

dunnwes, check your Private Messages. The button at upper right of this page.
Bearded_One brought up something very important. Your main air hose has to have a pressure sensor tube running down through it to attach to the 420E.
Keep both pressed down together for a few seconds then release them. You're now in the clinical menu.
Use the Info Access button to go to the next item.
Use the on/off button and ramp button (like up/down buttons) to change settings.
The last item in the menu will be "ID." When you press the Info Access button one more time, that will exit you from the menu. Any settings you changed will be automatically saved.
Note -- When going through the menu, the next-to-last item will be the "pressure sensor offset" setting. Do not change that. That item will look like this, although the setting for your machine will probably be a different number:

dunnwes, check your Private Messages. The button at upper right of this page.
Bearded_One brought up something very important. Your main air hose has to have a pressure sensor tube running down through it to attach to the 420E.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
The exact address for downloading the user's manual
Since the address I posted above requires searching for the manual, I'm adding a more precise one:
http://solvitcenter.puritanbennett.com/ ... %20(C).pdf
O.
http://solvitcenter.puritanbennett.com/ ... %20(C).pdf
O.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023