Next step help
-
josephny
Next step help
Hello,
Can someone help me with the next step please?
I've been using a Respironics M-series Cflex Auto for about 6 months.
I download the data to Encore Pro and look at it in Encore Analyser.
I don't really know how to interpret the data.
Should I change it from Auto to a set pressure?
Should I be seeing better results?
Is there something else I should be doing?
Is there a way I can post the data file or report here?
Thank you,
Joseph
Can someone help me with the next step please?
I've been using a Respironics M-series Cflex Auto for about 6 months.
I download the data to Encore Pro and look at it in Encore Analyser.
I don't really know how to interpret the data.
Should I change it from Auto to a set pressure?
Should I be seeing better results?
Is there something else I should be doing?
Is there a way I can post the data file or report here?
Thank you,
Joseph
-
Sleepdeprived
- Posts: 325
- Joined: Mon May 14, 2007 3:07 pm
- Location: Florida
Re: Next step help
[quote="josephny"]Hello,
Can someone help me with the next step please?
I've been using a Respironics M-series Cflex Auto for about 6 months.
I download the data to Encore Pro and look at it in Encore Analyser.
I don't really know how to interpret the data.
Should I change it from Auto to a set pressure?
Should I be seeing better results?
Is there something else I should be doing?
Is there a way I can post the data file or report here?
Thank you,
Joseph
Can someone help me with the next step please?
I've been using a Respironics M-series Cflex Auto for about 6 months.
I download the data to Encore Pro and look at it in Encore Analyser.
I don't really know how to interpret the data.
Should I change it from Auto to a set pressure?
Should I be seeing better results?
Is there something else I should be doing?
Is there a way I can post the data file or report here?
Thank you,
Joseph
Hi, Joseph and welcome to the board.
You're gonna get a reply or 3 asking you to 1) register for the forum, 2) fill out your profile and include which system/humidifier/masks you've been using, 3) give us some results of your readouts, 4) read the info in the yellow lightbulb titled "Our Collective Wisdom", and, 5) in general, provide us with tons of things that will probably seem like a pain in the butt.
That said, there are some outstanding folks here who have a ton of experience with your machine and have experienced your problems and are downright eager to help folks. Unfortunately, I'm not one of those who have the same machine, etc.
But welcome to the board and hang in there!
Kajun
You're gonna get a reply or 3 asking you to 1) register for the forum, 2) fill out your profile and include which system/humidifier/masks you've been using, 3) give us some results of your readouts, 4) read the info in the yellow lightbulb titled "Our Collective Wisdom", and, 5) in general, provide us with tons of things that will probably seem like a pain in the butt.
That said, there are some outstanding folks here who have a ton of experience with your machine and have experienced your problems and are downright eager to help folks. Unfortunately, I'm not one of those who have the same machine, etc.
But welcome to the board and hang in there!
Kajun
This therapy WORKS!!!
-
Guest
Thanks guys for the quick responses.
Here's some basic info from the 9/18/07-10/2/07 report:
avg 90% 7.4
avg pressure 5.7
Avg AI 1.1
Avg HI 2.3
Avg AHI 3.3
Avg flow limitation 1.2
Avg max leak 63.8 (Breeze Nasal pillows)
Avg 90% leak 33.7
Avg leak 26.1
Avg large leak 16 seconds
Cumulative usage 105 hours
avg usage 7 hours, 2 minuts
auto-cpap mean pressure 5.6
auto-cpap peak avg pressure 9.0
avg device pressure 90% 7.3
Does this suggest I should do something different?
Thank you,
Joseph
Here's some basic info from the 9/18/07-10/2/07 report:
avg 90% 7.4
avg pressure 5.7
Avg AI 1.1
Avg HI 2.3
Avg AHI 3.3
Avg flow limitation 1.2
Avg max leak 63.8 (Breeze Nasal pillows)
Avg 90% leak 33.7
Avg leak 26.1
Avg large leak 16 seconds
Cumulative usage 105 hours
avg usage 7 hours, 2 minuts
auto-cpap mean pressure 5.6
auto-cpap peak avg pressure 9.0
avg device pressure 90% 7.3
Does this suggest I should do something different?
Thank you,
Joseph
Wow, I didn't realize these were good numbers.
I'm pretty sure I'm not mouth breathing -- no more sore throats in morning.
So as not to make you regret asking, I'll limit and summarize the answer to how do I feel
I no longer wake up feeling exhausted and with a massive headache.
But, I still have substantial afternoon fatigue (though less so than before starting the therapy).
Other health issues remain, such as difficulty breathing through the nose (frequent congestion and possible anatomical issues) and diabetes.
Here's a summary of pressure vs. AHI:
4cm for 841 minutes: 2.6 ahi
5cm for 441 minutes: 2.6
6cm for 240 minutes: 1.8
7cm for 80 minutes: 1.5
8cm for 27.5 minutes: 2.2
9cm for .5 minutes: 0
Nothing higher.
I'm pretty sure I'm not mouth breathing -- no more sore throats in morning.
So as not to make you regret asking, I'll limit and summarize the answer to how do I feel
I no longer wake up feeling exhausted and with a massive headache.
But, I still have substantial afternoon fatigue (though less so than before starting the therapy).
Other health issues remain, such as difficulty breathing through the nose (frequent congestion and possible anatomical issues) and diabetes.
Here's a summary of pressure vs. AHI:
4cm for 841 minutes: 2.6 ahi
5cm for 441 minutes: 2.6
6cm for 240 minutes: 1.8
7cm for 80 minutes: 1.5
8cm for 27.5 minutes: 2.2
9cm for .5 minutes: 0
Nothing higher.
-
Wulfman...
joseph, it looks like things are going pretty good for you. You might want to google "sleep debt" and see what info there is out there on it. In a nutshell, one theory is that we folks who have OSA have had it for years and it takes out bodies quite a while to pay back the sleep debt we've accumulated over the years. Some folks think it's hooey, some don't.
As far as bumping up the bottom pressure, there have been quite a few folks who complained that at 4 cm, they feel they can't breathe....not enough air coming in. So that suggestion may be spot on.
Good luck and keep us posted.
Kajun
As far as bumping up the bottom pressure, there have been quite a few folks who complained that at 4 cm, they feel they can't breathe....not enough air coming in. So that suggestion may be spot on.
Good luck and keep us posted.
Kajun
This therapy WORKS!!!
-
oceanpearl
- Posts: 650
- Joined: Mon Oct 31, 2005 4:21 pm
- Location: Virginia Beach, Va
- rested gal
- Posts: 12880
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Your numbers look very good.
I (I'm not a doctor) agree with Den and Snoredog about raising the minimum pressure to 7.
I (I'm not a doctor) agree with Den and Snoredog about raising the minimum pressure to 7.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
-