M Series Heated Humidifier runs dry during night

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
brians2000
Posts: 13
Joined: Fri Jun 29, 2007 9:35 am

M Series Heated Humidifier runs dry during night

Post by brians2000 » Mon Jul 23, 2007 1:16 am

I have had my CPAP machine about two weeks. I noticed my M Series Heated Humidifier runs dry so I took it back and got another one of the same kind. The lady at the sleep clinic didnt think it should run dry so she gave me another humidifier. The new one runs dry after about six to eight hours also. She said not to allow it to run dry but if it does in the middle of the night there is nothing I can do about it. I hope it wont damage my cpap machine. What is everyones thoughts.


User avatar
kteague
Posts: 7788
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Humidfier runs dry

Post by kteague » Mon Jul 23, 2007 1:29 am

Those humidifiers are small. I don't know what yours is set on, but I have to keep mine around 2 to not run out overnight. If you need more humidification, some on here have discussed before about adding a different humidifier. They'll probably chime in with options if that's what you need.

Kathy


_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

dllfo
Posts: 882
Joined: Wed May 03, 2006 11:37 pm
Location: Sacramento, CA

Post by dllfo » Mon Jul 23, 2007 1:52 am

If your unit is like my Respironics, press down on the Heat key, holding it until the LCD display changes to a number. IF this number is a 3 or higher, that is part of why you are running out of water. I think I press the key below the heat key to reduce the numbers, then to increase you push the heat button down? Sorry, I haven't done this in a long time. I set mine on 1 most of the time.

Good luck with it...

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): respironics

Installing Software is like pushing a rope uphill.
I have Encore Pro 1.8.65 but could not find it listed
under software.

I LOVE the SV.

User avatar
Snoredog
Posts: 6397
Joined: Sun Mar 19, 2006 5:09 pm

Post by Snoredog » Mon Jul 23, 2007 2:18 am

The M series only holds about 11-12 oz. of water (about the same as the classic if you compare).

High use of water: Mouth breathing, mask leak, high pressure, turning humidifier up to high can cause high consumption.

If you are getting vapor droplets forming in the hose/mask you have it up too high, having it up too high uses more water but they are really not designed to heat the incoming air.

You may need to go to a passover humidifier with larger capacity as opposed to existing integrated.

I would check and eliminate leaks first.

someday science will catch up to what I'm saying...

User avatar
blarg
Posts: 1407
Joined: Mon Dec 11, 2006 10:21 pm
Location: Sydney, Australia

Post by blarg » Mon Jul 23, 2007 4:29 am

I had the same problem with the M series. The classic tank humidifier doesn't even come close to running dry.

You might consider getting a Fisher and Paykel HC150.

I'm a programmer Jim, not a doctor!

User avatar
MartiniLover
Posts: 364
Joined: Wed Jan 12, 2005 4:16 pm
Location: Davison Michigan

Post by MartiniLover » Mon Jul 23, 2007 6:45 am

I turned the heat off, and hooked up a passover humidifier in line. Works great. I am lucky enough to be able to stand not having the heat on, and not drying up. With both containers filled half full to start, it lasts about a week.

Works for me


_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Second favorite mask--Nasal Aire II
I am a two martini lover. Two martinis and I think I am a lover!

Bipap 13/9, 10ft Hose

mikester
Posts: 54
Joined: Mon May 14, 2007 6:05 am

Post by mikester » Mon Jul 23, 2007 8:20 am

Weird - I have an M series and run the humidifier on '2' and I usually only have to fill it up every other night.

I bet it has something to do with the relative humidity of the air in the room. Our fancy shmancy thermostat says our indoor humidity is usually around 55%. I can see how in drier areas you would use more water.


User avatar
6PtStar
Posts: 2659
Joined: Mon Jun 18, 2007 1:58 pm
Location: Texas, Is there any other place?

Post by 6PtStar » Mon Jul 23, 2007 8:44 am

I had trouble with mine and wrote Respronics about it. Below is the note I got back from them. I was concerned about damage to the unit if it ran dry.


"Dear patient,

The CPAPS were tested at a setting of 5(the highest) and a pressure of 20cm. They lasted 7 to 8 hours. There are some things which will affect that time. The first thing is a large mask leak will cause faster air flow which uses the water faster. The second is the temperature of the room (cool to cold),and the third is the humidity ( dry) in the room. These all will shorten the time which the water will last. If the water runs out before you wake, the heater will not burn up. It does not go above 145 degrees on the heating plate. That is not high enough to do any damage. If you should wake at some time, it is quite alright to add water if you wish.

Product support"

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: 11cm/H2O, Encore Pro 1.8i, Pro Analyzer, Encore Viewer1.0 - 3 Remstar Pro2's, 1 Remstar Auto
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS

User avatar
socknitster
Posts: 1740
Joined: Fri Jun 01, 2007 11:55 am
Location: Pennsylvania
Contact:

Post by socknitster » Mon Jul 23, 2007 10:07 am

My M-series is set on MAX (5) and I fill it up every night and it only comes close to running out if I sleep 10+ hours.

I would look at mouth leaks if you do nasal-only. When you leak from your mouth, machine uses more pressure and more water, I guess, not sure, but I know there is a correlation between mouth leaks or leaks in general and water loss from humidifier.

I tape my mouth, so rarely leak, unless tape is loose or something.

That would be where I would look first. I think if two machines did the same thing, it is not likely the machine.

Jen


Guest

Post by Guest » Mon Jul 23, 2007 12:38 pm

If the water runs out before you wake, the heater will not burn up. It does not go above 145 degrees on the heating plate. That is not high enough to do any damage. If you should wake at some time, it is quite alright to add water if you wish.
Ha! Yeah right. My M-Series Pro, humidifier set on 3, ran out of water exactly once (when I forgot to top it off before going to sleep that night). I awoke to a horrible, choke-inducing aroma and ripped off the mask.

Added water to tank. Still horrible smell. Cleaned tank with soap/water and refilled. Still there. Cleaned hose. Still there. Only when I thoroughly cleaned my mask too, was I able to go back to sleep without that horrible, intolerable odor.

Beware, M-series owners. Don't ever forget to fill or top off your humidifier if it needs it.


brians2000
Posts: 13
Joined: Fri Jun 29, 2007 9:35 am

Post by brians2000 » Mon Jul 23, 2007 8:45 pm

The technician at the sleep center told me the CPAP would break if the water ran out in as little as two hours. I was told I would smell it burning and what you said makes perfect sense. I didn't smell anything last night but I will make sure it is always filled and if I wake up in the middle of the night I refill it. I hate that the tank is not bigger. Does anyone know what would happen to the humidifier if I filled it above the fill mark. thanks, Brian


Daddysaur
Posts: 78
Joined: Fri Jul 13, 2007 8:01 pm
Location: Sunnyvale Ca.

Post by Daddysaur » Mon Jul 23, 2007 8:55 pm

The humidifier business is a mystery to me. You, before CPAP, and the hoseless aliens around us are fine without them. And in theory you are breathing the same amount of air with the same relative humidity as the hoseless unless you have leaks/mouth open issues. So why is the humidifier a critical part of the CPAP system? What types of issues should I watch out for without the humidifier?


User avatar
socknitster
Posts: 1740
Joined: Fri Jun 01, 2007 11:55 am
Location: Pennsylvania
Contact:

Post by socknitster » Mon Jul 23, 2007 9:22 pm

[quote="brians2000"]The technician at the sleep center told me the CPAP would break if the water ran out in as little as two hours. I was told I would smell it burning and what you said makes perfect sense. I didn't smell anything last night but I will make sure it is always filled and if I wake up in the middle of the night I refill it. I hate that the tank is not bigger. Does anyone know what would happen to the humidifier if I filled it above the fill mark. thanks, Brian


User avatar
socknitster
Posts: 1740
Joined: Fri Jun 01, 2007 11:55 am
Location: Pennsylvania
Contact:

Post by socknitster » Mon Jul 23, 2007 9:27 pm

[quote="Daddysaur"]The humidifier business is a mystery to me. You, before CPAP, and the hoseless aliens around us are fine without them. And in theory you are breathing the same amount of air with the same relative humidity as the hoseless unless you have leaks/mouth open issues. So why is the humidifier a critical part of the CPAP system? What types of issues should I watch out for without the humidifier?


User avatar
goose
Posts: 1382
Joined: Sun Mar 11, 2007 7:59 pm
Location: The left coast - CA... If you're not living on the edge, you're taking up too much space!!

Post by goose » Mon Jul 23, 2007 9:41 pm

I run my humidifier at 1 and fill it every other night.
I live in a dry climate (for the most part -- it's humid tonight), so it's nice to have a little moisture.....

cheers
goose


_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Also Use ComfortGel (s); Headrest (XL) and a PAP-Cap.
Wars arise from a failure to understand one another's humanness. Instead of summit meetings, why not have families meet for a picnic and get to know each other while the children play together?

-the Dalai Lama