Its your backup rate (BR) that's getting you in trouble. Do you have someone who could observe what your resting rate of breathing is while you're watching TV or reading or the like? I'm willing to bet its going to be at 14ish, possibly even lower (mine tends to be, always has been too). With regular BiPAP ST, I was started with a too low of a backup rate originally, but then it was upped to a range of 10-12 (my sleep doc tends to script most things in ranges, letting the patient find what works best). For me, 10 was were I did best, any faster and I was fighting the machine like you're describing, but my typical resting breathing rate tends to be around 12 BPM. We were able to see a difference from the data downloaded from my Synchrony that I needed a higher BR and once we made the switch, my percentage of patient triggered breaths actually increased, because it helped stabilize my breathing even more (though not as much as the VPAP Adapt does).Nazareth wrote: My numbers are 12/8 with respiratory rate of 14 bpm's, rise of 3, inspiratory time of 1.5- the report goes on to say that at these rates, central events subsided and sustained sleep was achieved.
.... the machine just acts really wierd. the machine breaths are erratic- it Ipaps for about 1.5 seconds, then epaps for about 2 seconds, then 2 seconds Ipap, 2 second Epap, then 4 seconds Ipap, 3 second Epap, back to 1.5 repat this sequence.
If I breath out after the 4 second one, and keep breathing out while the 1.5 second machine inhale takes place, it will go right back to the 4 second machine inhale- but I have to keep breathing out while the machine does it's 1.5 Ipap.
Let's play with the numbers a little and you'll see what is happening:
BR=14, 60 seconds/14=4.28seconds per breath, and that is not from the end of exhale to the beginning of inhale, but rather each breath should take that long period.
Normal breath ratio is about 40% inhale & 60% exhale, so that makes inhale (IPAP) about 1.71 seconds, which is pretty close to the 1.5 seconds you reported. Exhale (EPAP) would be 2.6 seconds. Sound about right for what you're experiencing? You're not going to see any regular, spontaneous BiPAP time unless your BPM is over 14; as soon as it drops below 14, TIMED mode kicks in to try to keep you breathing. Something you can try is after you hook up & have been breathing for a minute or so, is to hold your breath to see what Timed mode feels like as it cycles. It should kick in pretty fast, and within 10 seconds it would cycle 2 1/2 times. I usually did it while watching a second hand just to see the timing whenever I switched machines or my BR was adjusted.
Another thing that can be helpful when you're new to BiPAP, or anytime it feels like the machine pushing you around when it shouldn't, is to breath in through your nose & out through your mouth about 10 times. It helps slow you down and get in synch- or rather let the machine get in synch with you (but I know it can feel the other way around sometimes). This probably won't help much right now, if your BR is too high, as I believe it is, but the fact you said it felt easier when you woke up & wasn't fighting you then, makes me think there is some component of anxiety coming into play too. I know for me, I did the in through the nose & out through the mouth a lot in the beginning, and now I only need to do it when my nose & sinuses are clogged- usually a few breaths that way will help clear them enough so that air flow is high enough to not affect treatment or trigger early flipping to EPAP, like DSM mention. I know if I don't do that, the feeling of the machine pushing me gets worse & worse until it makes me want to take off the mask & say screw it for the night, even with my VPAP Adapt. Its funny, now I actually look forward to going to bed and being able to breath through my nose when I'm stuffy, when it use to be that I would get so stuffy sleeping due to allergies that by morning I would be mouth breathing.
I'm appalled at your DME & doc letting you go that long without treatment well they had a supposed pissing contest with Medicare, which won't do anything until they get billed. It sounds like the DME didn't want to take a risk of not getting approved or something- who knows? The only change recently in Medicare requirements happened early in 2006. They dropped the O2 desat below 88% for 5 consecutive minutes and the addition of Complex Sleep Apnea, which is when central apneas occur consistently & persistently as a result of CPAP to treat OSA. I barely squeaked by the O2 desat originally, even with 60+ centrals per hour- centrals don't tend to cause desats as much as obstructive apneas tend to, especially if they happen very rapidly. Because there was a question if I would meet the then required desat and because BiPAP ST only reduced my centrals from 60+ per hour to 30+ per hour, my sleep doctor didn't even want to get Medicare involved at first, until he knew for certain it would actually help. So he arranged for me to use an older model BiPAP ST from the sleep labs attached DME for a month, free of charge- and considering one months rent is over $600 & they knew ahead of time I would not be using that DME for my final equipment, it was very considerate. Of course it also made it easy for my doctor to document that BiPAP ST was needed when I came back in a month telling him there was no way I would dream of sleeping without it again & had a typed list of a dozen or so improvements I had noticed (sleeping less, no naps, less sleep meds, improved cognitive function, etc). I had no problem getting my BiPAP ST rental approved, and last fall, switched to VPAP Adapt, with my DME. In fact, there was a mix up in billing my heated humidifier, so my DME wrote it off and didn't blink an eye when they had to provide me with a ResMed VPAP Adapt integrated humidifier, also at no charge. They did get more then adequate reimbursement on my BiPAP ST, as I was under the old, rental for life for several years, until April 2006, when BiPAP ST was switched to capped rental, like the rest of xPAP machines have always been.
Let me know if that helps & what your resting BPM is, if you can. It sounds like you will be needing to contact your sleep doctor to have your backup rate changed (if you do, ask if can write a range, so you & the DME don't have to bug him every time). And then you DME will have to change the BR, unless your doctor & you are comfortable with you changing your prescribed settings- some doctors are & some are not.