Optilife leak rate / Respironics M Series PRO
Optilife leak rate / Respironics M Series PRO
Hello,
I am a new cpap user. I would like to know how to calculate the true leak rate. I have a Respironics M Series PRO + Optilife. I have been looking for Optilife's leak rate -or vent flow rate- I have not been able to find it.
Any tips are welcome
Thanks
I am a new cpap user. I would like to know how to calculate the true leak rate. I have a Respironics M Series PRO + Optilife. I have been looking for Optilife's leak rate -or vent flow rate- I have not been able to find it.
Any tips are welcome
Thanks
should be a Flow Chart in the paperwork that came with the mask, nearly all mask interfaces come with them.
Once you find that flow chart, you first find your pressure say its 9cm.
You look on the mfg supplied Flow Chart and find your pressure. You then look over and see the expected flow at that pressure where the two lines intersect.
This flow rate found is the "intentional" leak rate when the mask is completely sealed against your face. This intentional leak/flow is the expected amount of air escaping the exhaust holes that evacuate CO2 that you exhale at the given pressure.
So once you know what your particular mask flow rate is, you can determine from that how much "external" leak you have vs. "intentional" leak.
So if the LCD display says your leak is 48L/m and the flow chart says it should be 32L/m @9cm pressure, then it is 48 -32 or 16L/m of external leak, a bit too much.
How much external leak is acceptable? Depends on the machine and pressure, if your pressure is 18cm a 12L/m external leak may be completely acceptable, but if your pressure is only 8cm, 12L/m external leak may be too much.
12L/m is just a number picked, years ago we researched it out a bit more and it seemed that was the number you wanted to stay below with as it related to external leak, but there are many variables found within that number.
Once you find that flow chart, you first find your pressure say its 9cm.
You look on the mfg supplied Flow Chart and find your pressure. You then look over and see the expected flow at that pressure where the two lines intersect.
This flow rate found is the "intentional" leak rate when the mask is completely sealed against your face. This intentional leak/flow is the expected amount of air escaping the exhaust holes that evacuate CO2 that you exhale at the given pressure.
So once you know what your particular mask flow rate is, you can determine from that how much "external" leak you have vs. "intentional" leak.
So if the LCD display says your leak is 48L/m and the flow chart says it should be 32L/m @9cm pressure, then it is 48 -32 or 16L/m of external leak, a bit too much.
How much external leak is acceptable? Depends on the machine and pressure, if your pressure is 18cm a 12L/m external leak may be completely acceptable, but if your pressure is only 8cm, 12L/m external leak may be too much.
12L/m is just a number picked, years ago we researched it out a bit more and it seemed that was the number you wanted to stay below with as it related to external leak, but there are many variables found within that number.
someday science will catch up to what I'm saying...
Snoredog wrote:should be a Flow Chart in the paperwork that came with the mask, nearly all mask interfaces come with them.
Once you find that flow chart, you first find your pressure say its 9cm.
gmartine wrote:I do not have that flow chart My mask did not come with it
You look on the mfg supplied Flow Chart and find your pressure. You then look over and see the expected flow at that pressure where the two lines intersect.
This flow rate found is the "intentional" leak rate when the mask is completely sealed against your face. This intentional leak/flow is the expected amount of air escaping the exhaust holes that evacuate CO2 that you exhale at the given pressure.
So once you know what your particular mask flow rate is, you can determine from that how much "external" leak you have vs. "intentional" leak.
So if the LCD display says your leak is 48L/m and the flow chart says it should be 32L/m @9cm pressure, then it is 48 -32 or 16L/m of external leak, a bit too much.
How much external leak is acceptable? Depends on the machine and pressure, if your pressure is 18cm a 12L/m external leak may be completely acceptable, but if your pressure is only 8cm, 12L/m external leak may be too much.
12L/m is just a number picked, years ago we researched it out a bit more and it seemed that was the number you wanted to stay below with as it related to external leak, but there are many variables found within that number.
Here is a link to the web site PDF with the graph
http://maskseries.respironics.com/PDF/LeakRate.pdf
It has all the respironic masks
http://maskseries.respironics.com/PDF/LeakRate.pdf
It has all the respironic masks
Keep a song in your heart......it's like karaoke for the voices in your head
The Optilife mask isn't on that chart. A quick check against the leak rate chart that came with my Opti shows it to be the same as the Comfortfull2 leak chart.
Brenda
Brenda
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Love my papillow, Aussie heated hose and PAD-A-CHEEKS! Also use Optilife, UMFF(with PADACHEEK gasket), and Headrest masks Pressure; 10.5 |
I got my mask from Apria. When I got it it had a little (5"x5") booklet that says instructions for use. If you call respironics you can have them send you one, or they can e-mail it to you. The # is 1-800-345-6443.
Brenda
Brenda
_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Love my papillow, Aussie heated hose and PAD-A-CHEEKS! Also use Optilife, UMFF(with PADACHEEK gasket), and Headrest masks Pressure; 10.5 |
- feeling_better
- Posts: 803
- Joined: Sun Jun 01, 2008 7:08 pm
I am trying out the Optilife mask. My dme does not have any manual, since they are recycling this mask among trial users. I called the above number at respironics and they claimed they don't have any (obviously not very believable They do not want to deal with the end customer, even the one who is about to buy one -- companies are really funny in their goals.bdp522 wrote:I got my mask from Apria. When I got it it had a little (5"x5") booklet that says instructions for use. If you call respironics you can have them send you one, or they can e-mail it to you. The # is 1-800-345-6443.
Brenda
[/url]
I am looking for just 2 more numbers from the spec page, if they have it. Noise in dBA, and the static pressure (resistance to flow) at some specified flow rate. If anybody finds these, could you kindly post them here? Thank you very much!