Based on the original post ... seems it was the hose that melted, not the power adapter. In other words, the power adapter may well be "UL" certified but that certification may not apply to the heated hose iteself.WearyOne wrote:Since I have one of these, I, too, have been wanting to know how it turned out. Don't know much about electrical stuff, so have 2 questions:
1. Is the big square heavy plug thing supposed to get hot when in use?
2. Is it okay to plug this big thing into a power strip or should it plug directly into the wall. (The lamp, cpap, and hose are currently plugged into the power strip. Mistake?)
I checked and the big heavy power supply thing has a "UL" approval on it, but there's nothing like that on the actual heated hose.
Thanks!
Pam
Hose Melted!!
- DreamStalker
- Posts: 7503
- Joined: Mon Aug 07, 2006 9:58 am
- Location: Nowhere & Everywhere At Once
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.
They already did. That's why we buy them from Australia and they don't make them here.DreamStalker wrote:I find it interesting that there is no one around to call the FDA regarding this CPAP accessory ...
Difference in the Pur Sleep case is that it was being made and marketed here.
I'm a programmer Jim, not a doctor!
blarg, I'm lost (not unusual). Can you explain? (Or are you joking.) I know there is some type of heated hose thing for some of the F&P machines (hose sold on cpap.com), but that might be different.blarg wrote:They already did. That's why we buy them from Australia and they don't make them here.DreamStalker wrote:I find it interesting that there is no one around to call the FDA regarding this CPAP accessory ...
_________________
| Machine: DreamStation 2 Auto CPAP Advanced with Humidifier |
| Mask: AirFit™ F40 System - L/STD |
| Additional Comments: Oscar Software | APAP: 11-13 | 2nd Mask: Simplus Medium |
I have 2 of them, one I purchased over a year ago and one I purchased a month or two ago. They both are still working.
However, I don't use the power supply recommended or that it came with. First one didn't come with a power supply and you had to obtain it in the states. The 2nd one I got came with the 1AMP x 12VDC power supply. It is huge.
Where I live it gets cool but not cold, so I found a 1AMP x 12VDC power supply to be too warm. So to run it cooler, I found in my junk drawer a 9VDC x 800ma power supply with same plug connector (came off an old Linksys router) and I have been using it nearly every night in the winter. I have the power supply on a digital timer, it only runs from 12AM to 6AM daily. I don't even keep track when it is on or off.
So far, my hose(s) have been working fine with the lower voltage power supply. I suspect the problem is with the new chinese made power supply, if it was getting hot, it means it couldn't handle the resistance of the hose, if it over supplied power it would melt the hose. What ever happened to the good products made in Mexico
However, I don't use the power supply recommended or that it came with. First one didn't come with a power supply and you had to obtain it in the states. The 2nd one I got came with the 1AMP x 12VDC power supply. It is huge.
Where I live it gets cool but not cold, so I found a 1AMP x 12VDC power supply to be too warm. So to run it cooler, I found in my junk drawer a 9VDC x 800ma power supply with same plug connector (came off an old Linksys router) and I have been using it nearly every night in the winter. I have the power supply on a digital timer, it only runs from 12AM to 6AM daily. I don't even keep track when it is on or off.
So far, my hose(s) have been working fine with the lower voltage power supply. I suspect the problem is with the new chinese made power supply, if it was getting hot, it means it couldn't handle the resistance of the hose, if it over supplied power it would melt the hose. What ever happened to the good products made in Mexico
someday science will catch up to what I'm saying...
All the workers came to the Us for free health care.Snoredog wrote: What ever happened to the good products made in Mexico
Mexico went on Siesta Daylight Savings Time, and now only work 2 hour days, so production has dropped.
Most of the work force is at the Y.M.C.A., taking swimming lessons. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Well, I dunno what the specific story is, but you can't straight up buy heated hoses here unless you get that F&P machine. I was more joking than anything.WearyOne wrote:blarg, I'm lost (not unusual). Can you explain? (Or are you joking.) I know there is some type of heated hose thing for some of the F&P machines (hose sold on cpap.com), but that might be different.
I'm a programmer Jim, not a doctor!
-
DreamSeeker
- Posts: 37
- Joined: Sat Sep 09, 2006 11:28 pm
- Location: Texas
Me too. Especially since I have two heated hoses.
_________________
| Machine: DreamStation Auto CPAP Machine |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Compliant since April 2003. (De-cap-itated Aura). |
Since Ellie and I (at least) had almost identical experiences, my recomendation is that you consider not using the power supply which came with the hose. I also believe that the fact that we both had the failure within a few weeks of getting the hose may mean that it is a failure which shows up in "infancy" of the power supply and if you haven't had the problem you might never have it. I don't think there is anything wrong with the hose.
I'm going to order another hose and I'll use the radio Shack power supply I got to replace the original one.
Just my 0.02 as an effected user (who loves his heated hose too).
TerryB
I'm going to order another hose and I'll use the radio Shack power supply I got to replace the original one.
Just my 0.02 as an effected user (who loves his heated hose too).
TerryB
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: 14 CM , C-Flex Off |
TerryB, did you get the same type of power supply at Radio Shack, or something different?
Pam
Pam
_________________
| Machine: DreamStation 2 Auto CPAP Advanced with Humidifier |
| Mask: AirFit™ F40 System - L/STD |
| Additional Comments: Oscar Software | APAP: 11-13 | 2nd Mask: Simplus Medium |
I guess you'd say it is the same type in that it is a wall plug in transformer in a single case. The output is to a cord which accepts multiple plugs to connect to the load. I took the old supply to the store to find the right size plug. Voltage and current rating are also similar if not identical.
TerryB
Remember that the original power supply overheated the heater coil wire before dying. Although the hose is damaged, it still has no appreciable leaks. I expect it to fail from fatigue where moved around the humidifier connection or where coming over the headboard. I guess I should get busy and order the replacement before I have a failure and have to put up with rainout again.
TerryB
Remember that the original power supply overheated the heater coil wire before dying. Although the hose is damaged, it still has no appreciable leaks. I expect it to fail from fatigue where moved around the humidifier connection or where coming over the headboard. I guess I should get busy and order the replacement before I have a failure and have to put up with rainout again.
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: 14 CM , C-Flex Off |
- Sleepy Dog Lover
- Posts: 221
- Joined: Thu May 04, 2006 4:27 pm
[quote="TerryB"]Since Ellie and I (at least) had almost identical experiences, my recomendation is that you consider not using the power supply which came with the hose. I also believe that the fact that we both had the failure within a few weeks of getting the hose may mean that it is a failure which shows up in "infancy" of the power supply and if you haven't had the problem you might never have it. I don't think there is anything wrong with the hose.
I'm going to order another hose and I'll use the radio Shack power supply I got to replace the original one.
Just my 0.02 as an effected user (who loves his heated hose too).
TerryB
I'm going to order another hose and I'll use the radio Shack power supply I got to replace the original one.
Just my 0.02 as an effected user (who loves his heated hose too).
TerryB
It's been over a year since the event, but my recollection is they said, more or less, Nah, can't happen, we run a hose and power supply night and day have never had one overheat. I'll look back into the emails and get some more direct quotes, but as I've said, I think the failure was in the power supply/transformer.
TerryB
TerryB
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: 14 CM , C-Flex Off |
Terry, I got my first heated hose about a year ago and it didn't come with the power supply. I had to buy it from Radio Shack.
So, are you saying that you and Ellie got your hoses about a year ago and they came with the power supply? And you think those original power supplies are the culprit?
Oops, just reread and see that Ellie has only had her's a short period of time.
So, are you saying that you and Ellie got your hoses about a year ago and they came with the power supply? And you think those original power supplies are the culprit?
Oops, just reread and see that Ellie has only had her's a short period of time.
_________________
| Machine: DreamStation Auto CPAP Machine |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Compliant since April 2003. (De-cap-itated Aura). |
Here is the exchange with Donald,
Note time frame from shipment Late April to failure late May to disccovery of coils penetrating hose late Jun.
I'm going to order a replacement hose right now.
TerryB
-----Original Message-----
>From: Donald Fischer <don.fischer@optusnet.com.au>
>Sent: Jun 20, 2006 3:04 AM
>To: TerryB <terryblauser@earthlink.net>
>Subject: Re: Sleepzone Heated CPAP tube order
>
>SleepZone wrote :
>
>G'day Terry from the land of OZ
>
>The heating coils are wound on the outside of the tube itself ..... ata resistance of 13ohms and 12v X1amp not enough heat is produced tocause any melting or problem like that .We have heated tubes on test24/7 at even high ambient temps above 30 degrees C and thereis noproblem .... The worst that can happen is an open circuit we havenever had a short circuit ... in either case the tube will simply notheat ... the product is very low power 11 watts .. and certified andlisted category safe TGA Australia which is the eqiv USFDA .
>
>So ... if the hose is damaged in some way it would be due to amechanical cause not heat or melting ... Does the tube leak air ... ifnot all is A OK
>
>Hope this helps
>
>Don Fischer
>
>
>
>
>
>TerryB wrote:
>Hello Donald,
> I hope things are going well down under.
>
> I have a concern about my Heated Hose and I'm hoping youwill set my mind to rest. I was rinsing out the other day and I lookedin the two ends and found heating coils visible inside the hose at oneend and not the other end. Does this mean that I have managed tooverheat the coils and melt them through the hose and that the hosewill fail soon in this area? I suppose it could also mean this is theway the coils are anchored so they dont change position. If it is asign of overheating then I am concerned about my ability to use theequipment safely as I know that some nights I decide to remove the maskfor a while and I forget that I should unplug the hose. On the otherhand could it be that the failure mode of the first transformerprovided too much current and fired this area?
>
> So you see my concerns.
>
>Thanks in advance for any information and guidance.
>
>Terry Blauser
>
>
>
>
>-----OriginalMessage-----
>From: Donald Fischer
>Sent: May 22, 2006 6:04 PM
>To: TerryB
>Subject: Re: Sleepzone Heated CPAP tube order
>
>Don Fischer wrote :
>G'day Terry what you have told us we really appreciate I will take thisup with our transformer supplier its the first prob we have had
>considering we test B4 shipping
>The heated tube apart from physical damage is petty well indestructablewe have never had a heating element problem
>
>All the best any probs we would appreciate knowing
>
>SleepZone
>
>
>
>
>TerryB wrote:
>Hello Don,
> This afternoon readings are 13.3 Ohms hose,
>1.2 ohms transformer output winding.
>Not much voltage out of the transformer. o.5 millivolts butthe good news is it is at 60 hz.
>The transformer did OK for two weeks but went in the middleof the night.
>
>Listen, don't worry, I got a free transformer that lots ofmy predesessors didn't get. I'll go to Radio Shack and get one likemost of them had to do.
>
>Best Regards,
>
>Terry
>
>
>
>
>-----OriginalMessage-----
>From: Donald Fischer
>Sent: May 21, 2006 6:43 PM
>To: TerryB
>Subject: Re: Sleepzone Heated CPAP tube order
>
>
>
>SleepZone wrote :
>
>Terry,
>havent had this problem before as we test every transformer prior toshipping. The hose should be reading 13 ohms resistance,
>and the transformer should show an infinite output impedance.
>If you have access to an voltmeter, can you please confirm the outputvoltage is 12V AC??
>The transformer we supply is AC.
>
>If you could, can you re-check the impedances and AC voltage and getback to me?
>For other simple troubleshooting see attachment
>
>Please
>
>Regards,
>Don Fischer
>
>TerryB wrote:
>Hi Don,
> I wonder if you have been having any bad reports onthe power supply you are sending with the AHH on USA shipments? I hadcondensation problems last night a first since receiving the HH, andupon checking I believe the power supply has failed. I read no voltageon the output. The hose is reading 6 ohms DC resistance. I hope youcan verify that this is correct for the hose as I will go and get aRadio Shack Power supply if the hose is not the cause of the powersupply failure. Since I see less than 1 Ohm output winding resistance,I believe the transformer secondary has shorted in the power supply.
>
>
>Best Regards,
>Terry Blauser
>
>
>-----OriginalMessage-----
>From: Donald Fischer
>Sent: Apr 27, 2006 5:33 AM
>To: terryblauser@earthlink.net
>Subject: Sleepzone Heated CPAP tube order
>
>G'Day Terry Blauser from Australia "down under"
>
>Thank you for your order, which we will ship tomorrow Friday 28th April2006
>on Mastercard Authority for US$ 109.00 # 428735
>allow around 7 to 10 days for delivery ... which is receipted.
>
>A better nights sleep awaits you... Guaranteed
>No More Cold CPAP Air... No More Annoying Rainout
>
>[]
>Sincerely Don and Carmen Fischer
>SleepZone CPAP Products
>
>
>
>
Note time frame from shipment Late April to failure late May to disccovery of coils penetrating hose late Jun.
I'm going to order a replacement hose right now.
TerryB
-----Original Message-----
>From: Donald Fischer <don.fischer@optusnet.com.au>
>Sent: Jun 20, 2006 3:04 AM
>To: TerryB <terryblauser@earthlink.net>
>Subject: Re: Sleepzone Heated CPAP tube order
>
>SleepZone wrote :
>
>G'day Terry from the land of OZ
>
>The heating coils are wound on the outside of the tube itself ..... ata resistance of 13ohms and 12v X1amp not enough heat is produced tocause any melting or problem like that .We have heated tubes on test24/7 at even high ambient temps above 30 degrees C and thereis noproblem .... The worst that can happen is an open circuit we havenever had a short circuit ... in either case the tube will simply notheat ... the product is very low power 11 watts .. and certified andlisted category safe TGA Australia which is the eqiv USFDA .
>
>So ... if the hose is damaged in some way it would be due to amechanical cause not heat or melting ... Does the tube leak air ... ifnot all is A OK
>
>Hope this helps
>
>Don Fischer
>
>
>
>
>
>TerryB wrote:
>Hello Donald,
> I hope things are going well down under.
>
> I have a concern about my Heated Hose and I'm hoping youwill set my mind to rest. I was rinsing out the other day and I lookedin the two ends and found heating coils visible inside the hose at oneend and not the other end. Does this mean that I have managed tooverheat the coils and melt them through the hose and that the hosewill fail soon in this area? I suppose it could also mean this is theway the coils are anchored so they dont change position. If it is asign of overheating then I am concerned about my ability to use theequipment safely as I know that some nights I decide to remove the maskfor a while and I forget that I should unplug the hose. On the otherhand could it be that the failure mode of the first transformerprovided too much current and fired this area?
>
> So you see my concerns.
>
>Thanks in advance for any information and guidance.
>
>Terry Blauser
>
>
>
>
>-----OriginalMessage-----
>From: Donald Fischer
>Sent: May 22, 2006 6:04 PM
>To: TerryB
>Subject: Re: Sleepzone Heated CPAP tube order
>
>Don Fischer wrote :
>G'day Terry what you have told us we really appreciate I will take thisup with our transformer supplier its the first prob we have had
>considering we test B4 shipping
>The heated tube apart from physical damage is petty well indestructablewe have never had a heating element problem
>
>All the best any probs we would appreciate knowing
>
>SleepZone
>
>
>
>
>TerryB wrote:
>Hello Don,
> This afternoon readings are 13.3 Ohms hose,
>1.2 ohms transformer output winding.
>Not much voltage out of the transformer. o.5 millivolts butthe good news is it is at 60 hz.
>The transformer did OK for two weeks but went in the middleof the night.
>
>Listen, don't worry, I got a free transformer that lots ofmy predesessors didn't get. I'll go to Radio Shack and get one likemost of them had to do.
>
>Best Regards,
>
>Terry
>
>
>
>
>-----OriginalMessage-----
>From: Donald Fischer
>Sent: May 21, 2006 6:43 PM
>To: TerryB
>Subject: Re: Sleepzone Heated CPAP tube order
>
>
>
>SleepZone wrote :
>
>Terry,
>havent had this problem before as we test every transformer prior toshipping. The hose should be reading 13 ohms resistance,
>and the transformer should show an infinite output impedance.
>If you have access to an voltmeter, can you please confirm the outputvoltage is 12V AC??
>The transformer we supply is AC.
>
>If you could, can you re-check the impedances and AC voltage and getback to me?
>For other simple troubleshooting see attachment
>
>Please
>
>Regards,
>Don Fischer
>
>TerryB wrote:
>Hi Don,
> I wonder if you have been having any bad reports onthe power supply you are sending with the AHH on USA shipments? I hadcondensation problems last night a first since receiving the HH, andupon checking I believe the power supply has failed. I read no voltageon the output. The hose is reading 6 ohms DC resistance. I hope youcan verify that this is correct for the hose as I will go and get aRadio Shack Power supply if the hose is not the cause of the powersupply failure. Since I see less than 1 Ohm output winding resistance,I believe the transformer secondary has shorted in the power supply.
>
>
>Best Regards,
>Terry Blauser
>
>
>-----OriginalMessage-----
>From: Donald Fischer
>Sent: Apr 27, 2006 5:33 AM
>To: terryblauser@earthlink.net
>Subject: Sleepzone Heated CPAP tube order
>
>G'Day Terry Blauser from Australia "down under"
>
>Thank you for your order, which we will ship tomorrow Friday 28th April2006
>on Mastercard Authority for US$ 109.00 # 428735
>allow around 7 to 10 days for delivery ... which is receipted.
>
>A better nights sleep awaits you... Guaranteed
>No More Cold CPAP Air... No More Annoying Rainout
>
>[]
>Sincerely Don and Carmen Fischer
>SleepZone CPAP Products
>
>
>
>
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: 14 CM , C-Flex Off |