I've been using a ResMed Airsense 11 with full face mask (AirFit F40) for about two years and have never been able to get my AHI down.
I'm currently using min pressure 11, max pressure 15, EPR 3 but still get a lot of OA events. If I try to set the pressure any higher, the air just starts blowing out of the sides of the mask or waking me up.
Here are two OSCAR charts from different days.
OSCAR data evaluation, high OA
-
silentdragon
- Posts: 1
- Joined: Wed Nov 25, 2026 10:21 pm
OSCAR data evaluation, high OA
- Attachments
-
- nov8.png (100.94 KiB) Viewed 595 times
-
- nov7.png (105.48 KiB) Viewed 595 times
-
Sleepzilla
- Posts: 74
- Joined: Fri Nov 28, 2025 3:39 pm
Re: OSCAR data evaluation, high OA
Just a few observations...
Your charts show predominantly obstructive apnea (OA) events, with a minority of additional clear airway central apnea (CA) events.
The CPAP/APAP/BiPAP machines like yours are designed to prevent OSA only. They only go to the trouble of detecting the CSA/CA events in order to cheerfully ignore them. This can be easily seen in the pressure graph, as the machine makes no changes to pressure in response to the CA events.
Your CSA component might be pre-existing, or a treatment-emergent side effect of OSA therapy pressure flushing CO2-rich air from the PAP circuit, but that's probably best saved for another discussion. I just mention it here to point out that this machine can't/won't treat the CA events.
The hypopnea (H) events can be either obstructive or central, but in the presence of obviously predominant OSA, the accompanying hypopneas are usually obstructive in nature too. The only way to categorize hypopneas definitively is via clinical in-lab PSG. However, since your machine is an OSA-only therapy device, it assumes detected hypopneas are obstructive and responds accordingly.
Your EPAP range of 11-15 is typically sufficient to pneumatically brace an obstructed airway open in order to prevent OSA and related hypopneas. But even in the presence of the highest 15 cm H20 therapy pressure, you're still experiencing apparently random periods of drastically low flow accompanied by closely grouped clusters of OA events.
These are classic symptoms of positional OSA related to changing body position and/or chin-drop during sleep. Your PAP machine can't force therapy air through an airway that's obstructed like this. The only way to correct/resolve positional OSA is by identifying the cause (i.e. sleep position) and then make the necessary changes to prevent it.
Depending on the cause of your positional apnea, some examples of corrective action include changing pillows, wearing a soft cervical collar to prevent chin drop, or even wearing something that makes it uncomfortable to sleep on your back.
TL:DR I think your OSA therapy settings look great. If you can identify and correct the positional issues, then you might even be able to eventually reduce therapy pressure, and possibly reduce the CA event frequency too in the process.
Your charts show predominantly obstructive apnea (OA) events, with a minority of additional clear airway central apnea (CA) events.
The CPAP/APAP/BiPAP machines like yours are designed to prevent OSA only. They only go to the trouble of detecting the CSA/CA events in order to cheerfully ignore them. This can be easily seen in the pressure graph, as the machine makes no changes to pressure in response to the CA events.
Your CSA component might be pre-existing, or a treatment-emergent side effect of OSA therapy pressure flushing CO2-rich air from the PAP circuit, but that's probably best saved for another discussion. I just mention it here to point out that this machine can't/won't treat the CA events.
The hypopnea (H) events can be either obstructive or central, but in the presence of obviously predominant OSA, the accompanying hypopneas are usually obstructive in nature too. The only way to categorize hypopneas definitively is via clinical in-lab PSG. However, since your machine is an OSA-only therapy device, it assumes detected hypopneas are obstructive and responds accordingly.
Your EPAP range of 11-15 is typically sufficient to pneumatically brace an obstructed airway open in order to prevent OSA and related hypopneas. But even in the presence of the highest 15 cm H20 therapy pressure, you're still experiencing apparently random periods of drastically low flow accompanied by closely grouped clusters of OA events.
These are classic symptoms of positional OSA related to changing body position and/or chin-drop during sleep. Your PAP machine can't force therapy air through an airway that's obstructed like this. The only way to correct/resolve positional OSA is by identifying the cause (i.e. sleep position) and then make the necessary changes to prevent it.
Depending on the cause of your positional apnea, some examples of corrective action include changing pillows, wearing a soft cervical collar to prevent chin drop, or even wearing something that makes it uncomfortable to sleep on your back.
TL:DR I think your OSA therapy settings look great. If you can identify and correct the positional issues, then you might even be able to eventually reduce therapy pressure, and possibly reduce the CA event frequency too in the process.
Last edited by Sleepzilla on Thu Nov 26, 2026 2:31 pm, edited 2 times in total.
Re: OSCAR data evaluation, high OA
to follow up on what sleepzilla said, much of my osa is magnified when I sleep on my back. as they say, you want to get off of your back.
what I use is a fanny pack with a thick plastic bottle inside. this is enough to remind me when I turn over to slide the fanny pack so it's in position for that particular side.
I also use a firm foam collar. this is the one i've been using for the longest time and it helps too-
https://www.amazon.com/dp/B01M1O79MM?re ... vical&th=1
what I use is a fanny pack with a thick plastic bottle inside. this is enough to remind me when I turn over to slide the fanny pack so it's in position for that particular side.
I also use a firm foam collar. this is the one i've been using for the longest time and it helps too-
https://www.amazon.com/dp/B01M1O79MM?re ... vical&th=1
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg