Hello everyone,
First time poster here - grateful for the opportunity to share with fellow CPAP users and get tips and input. I read the stickies and hope I'm sticking to what information you need and am not adding too much - but in case it helps, a little bit of a background on my journey so far: I'm 37, male, 6'1, ~175 pounds and was diagnosed about this time last year with sleep apnea. It took about 5 months to feel really comfortable with the CPAP. I now really enjoy it and wouldn't want to sleep without it. I started with the N30i mask and didn't like it nearly as much as the mask I have now which I am quite happy with - the P30i. My nostrils are on the smaller side so it works a lot better and I feel much more... oxygenated? For lack of a better term. I also wear a chin strap and use tape to keep my mouth closed - which, in combination, work well. I also sleep slightly elevated (10 - 15 degrees or so).
My symptoms prior to diagnosis were waking up and gasping for air. I rarely would tire throughout the day, but the gasping for air was miserable and painful. So, as far as the therapy goes, most nights are good. Now that I'm using the CPAP, I feel groggy again at night or when I wake up, which in my case I think is a good thing. Untreated, I rarely if ever felt groggy or struggled to get up after sleeping (like I had a good, deep sleep) - so I'm actually quite happy with that and it's confirming to me that the treatment is largely successful. That said, if it's possible, I'd like to be around .5 AHI a night, but at least with my current setup, that has seemed elusive. I had EPR off until about 3 - 4 weeks ago, but would wake up winded due to the pressure being around 8 - 10. Setting EPR on has helped with that and is much more comfortable. However, either because of it or just my overall pressure setting perhaps being too low, it seems that my AHI is higher than it was at that time.
Last night was a good example of what I'm trying to figure out and avoid - my AHI was 3.2 and I woke up a few times. It wasn't painful, but I could tell I wasn't breathing quite right. Here are my graphs: https://imgur.com/a/w6nQFzq
It seems that around 4:40 to 5:20 in the morning the AHIs were around 11, but the pressure only hovered around 7.8'ish... And I'm curious as to why the pressure didn't increase or didn't show an increase during that time? Perhaps I should just elevate my base pressure to around 8 or 8.5? I'm careful to only adjust one thing at a time and test to see whether the results are better or worse.
Also, I'm curious as to why my apneas show up largely as "central" instead of "obstructive"? My provider mentioned that is common being on CPAP therapy and could just be a "side-effect" that occurs from time to time. Just looking to check that and maybe with my graphs something else can be seen or explained.
Overall, I'm just always aiming to improve and like to try to achieve that goal I have in mind of .5 AHIs / night. However, if that is an unreasonable expectation, let me know. Thanks in advance for any input or guidance!
First time poster here - grateful for the opportunity to share with fellow CPAP users and get tips and input. I read the stickies and hope I'm sticking to what information you need and am not adding too much - but in case it helps, a little bit of a background on my journey so far: I'm 37, male, 6'1, ~175 pounds and was diagnosed about this time last year with sleep apnea. It took about 5 months to feel really comfortable with the CPAP. I now really enjoy it and wouldn't want to sleep without it. I started with the N30i mask and didn't like it nearly as much as the mask I have now which I am quite happy with - the P30i. My nostrils are on the smaller side so it works a lot better and I feel much more... oxygenated? For lack of a better term. I also wear a chin strap and use tape to keep my mouth closed - which, in combination, work well. I also sleep slightly elevated (10 - 15 degrees or so).
My symptoms prior to diagnosis were waking up and gasping for air. I rarely would tire throughout the day, but the gasping for air was miserable and painful. So, as far as the therapy goes, most nights are good. Now that I'm using the CPAP, I feel groggy again at night or when I wake up, which in my case I think is a good thing. Untreated, I rarely if ever felt groggy or struggled to get up after sleeping (like I had a good, deep sleep) - so I'm actually quite happy with that and it's confirming to me that the treatment is largely successful. That said, if it's possible, I'd like to be around .5 AHI a night, but at least with my current setup, that has seemed elusive. I had EPR off until about 3 - 4 weeks ago, but would wake up winded due to the pressure being around 8 - 10. Setting EPR on has helped with that and is much more comfortable. However, either because of it or just my overall pressure setting perhaps being too low, it seems that my AHI is higher than it was at that time.
Last night was a good example of what I'm trying to figure out and avoid - my AHI was 3.2 and I woke up a few times. It wasn't painful, but I could tell I wasn't breathing quite right. Here are my graphs: https://imgur.com/a/w6nQFzq
It seems that around 4:40 to 5:20 in the morning the AHIs were around 11, but the pressure only hovered around 7.8'ish... And I'm curious as to why the pressure didn't increase or didn't show an increase during that time? Perhaps I should just elevate my base pressure to around 8 or 8.5? I'm careful to only adjust one thing at a time and test to see whether the results are better or worse.
Also, I'm curious as to why my apneas show up largely as "central" instead of "obstructive"? My provider mentioned that is common being on CPAP therapy and could just be a "side-effect" that occurs from time to time. Just looking to check that and maybe with my graphs something else can be seen or explained.
Overall, I'm just always aiming to improve and like to try to achieve that goal I have in mind of .5 AHIs / night. However, if that is an unreasonable expectation, let me know. Thanks in advance for any input or guidance!
- ChicagoGranny
- Posts: 15534
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Looking for input on CPAP setup / new forum member
You think feeling "groggy" is a good thing??? Define "groggy".
I hate
We can't tell much about that from an OSCAR screenshot. I'm skeptical about the central apneas. If you use a free account at SleepHQ, we can tell much more.
Welcome!
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Re: Looking for input on CPAP setup / new forum member
As far as feeling groggy goes - in my earlier years I was a heavier sleeper. I think that never changed, it's just that sleep apnea hit. Prior to CPAP treatment, I always felt very awake right away in the morning, like I hadn't been deeply asleep. Now when I wake up I feel more as though I've been deeply asleep, which I think is good... But, perhaps I'm wrong and it's not normal?ChicagoGranny wrote: ↑Sat Dec 13, 2025 1:17 pmYou think feeling "groggy" is a good thing??? Define "groggy".
I hateto see goals like this. More on that as the discussion continues.
We can't tell much about that from an OSCAR screenshot. I'm skeptical about the central apneas. If you use a free account at SleepHQ, we can tell much more.
Welcome!
Good to know if having a certain AHI goal isn't a good thing - would like to learn more there.
I think I have everything setup at SleepHQ - let me know how this looks or if I need anything different:
https://sleephq.com/public/teams/share_ ... /dashboard
- ChicagoGranny
- Posts: 15534
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Looking for input on CPAP setup / new forum member
I zoomed in on half a dozen of your CAs. Immediately preceding them, you take a deeper breath. This indicates arousals, and the CAs are called post-arousal CAs. They are not a problem. The arousals are the problem. You have a few too many arousals. Review your sleep hygiene and the side effects of any drugs you take.
How do you feel during the day? Energetic? Any excess sleepiness? At bedtime, do you feel fatigued, but not stressed?
How do you feel during the day? Energetic? Any excess sleepiness? At bedtime, do you feel fatigued, but not stressed?
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Re: Looking for input on CPAP setup / new forum member
Oh interesting, good to know. I'm not taking any medications. I believe I have pretty good sleep hygiene overall - in bed about the same time each evening. I fall asleep within 5 minutes usually of when I want to.ChicagoGranny wrote: ↑Sat Dec 13, 2025 6:45 pmI zoomed in on half a dozen of your CAs. Immediately preceding them, you take a deeper breath. This indicates arousals, and the CAs are called post-arousal CAs. They are not a problem. The arousals are the problem. You have a few too many arousals. Review your sleep hygiene and the side effects of any drugs you take.
How do you feel during the day? Energetic? Any excess sleepiness? At bedtime, do you feel fatigued, but not stressed?
I feel pretty good overall throughout the day. Not high energy, but not low either. I can't say I'm ever sleepy.
I have higher than normal anxiety in general (but as noted, no medications), so some stress I suppose, but not excessive.
Are the arousals something that CPAP therapy can resolve or is that something different entirely?
- ChicagoGranny
- Posts: 15534
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Looking for input on CPAP setup / new forum member
CPAP therapy is about breathing. It can prevent arousals caused by breathing events. But there are many reasons for arousals that are NOT related to breathing events.
So, CPAP is successfully treating your biggest problem. All I can suggest is to work on diet, moderate exercise, sleep hygiene and the ability to handle stress. Shouldn't everyone do that?
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- spitintheocean
- Posts: 193
- Joined: Thu Feb 26, 2009 10:47 am
- Location: Ottawa, Canada
Re: Looking for input on CPAP setup / new forum member
I went through something similar when I changed my mask and turned EPR on. My AHI also went up a bit during the first month until my body adjusted. What you describe with the pressure staying around 7.8 is exactly what I saw. Raising the minimum to around 8.5 helped a lot, not by much, but enough to stop those clusters from happening. Might be worth trying it for a night or two.
_________________
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Additional Comments: Settings: 17 - 19 - no EPR; CMS 50F Pulse Oximeter |
Life is something to do when you can't get to sleep.
Fran Lebowitz
Fran Lebowitz
- ChicagoGranny
- Posts: 15534
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Looking for input on CPAP setup / new forum member
At which times on his chart do you see clusters?
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Re: Looking for input on CPAP setup / new forum member
I appreciate the input from both of you!
I definitely have room to get more exercise. My diet is pretty good, I cook most nights and rarely if ever eat out. Mostly healthy meals, though they could be a little healthier depending on the night - I'm breaking into more vegetarian and vegan options. I'm thinking about trying to cut out dairy. I did that last year and it seemed to help some, but I had a lot of other changes going on at the time, so it's hard to say how impactful it was. It seems that when I do have dairy, my night is a little rougher. Anywho, lots of room on the lifestyle front and I'm going to try to work on that more.
I upped my pressure to 8, and I uploaded a few more nights of data. Last night was really awful... =( Hoping tonight will be better. Overall, upping the pressure to 8 did seem to help, though time will tell I suppose.
I'll keep providing updates as time goes on. =)
Oh, one addendum:
My sleep study showed that most of my sleep apnea events were around my deeper REM sleep that occurred during the second half of my sleep cycle(s). Which does tend to track well for what I'm still experiencing when I do have issues, for what that's worth.
I definitely have room to get more exercise. My diet is pretty good, I cook most nights and rarely if ever eat out. Mostly healthy meals, though they could be a little healthier depending on the night - I'm breaking into more vegetarian and vegan options. I'm thinking about trying to cut out dairy. I did that last year and it seemed to help some, but I had a lot of other changes going on at the time, so it's hard to say how impactful it was. It seems that when I do have dairy, my night is a little rougher. Anywho, lots of room on the lifestyle front and I'm going to try to work on that more.
I upped my pressure to 8, and I uploaded a few more nights of data. Last night was really awful... =( Hoping tonight will be better. Overall, upping the pressure to 8 did seem to help, though time will tell I suppose.
I'll keep providing updates as time goes on. =)
Oh, one addendum:
My sleep study showed that most of my sleep apnea events were around my deeper REM sleep that occurred during the second half of my sleep cycle(s). Which does tend to track well for what I'm still experiencing when I do have issues, for what that's worth.